Re: Bipap not working. working as cpap...HELP
Posted: Fri Apr 24, 2015 7:45 am
by Pugsy
yaconsult wrote:I'm starting to wonder if I should check that wire on my low hour 550P as opposed to waiting for it to fail? Do we have any idea how common this problem is? If rare, I wouldn't bother. If it happens 50% of the time or more, it would probably be worth it.
While not extremely rare (unless you are one of those it happens to and then it doesn't seem so rare )...I don't think it comes anywhere near the 50% category.
I would be of the opinion to go by the "if it ain't broke don't fix it" rule.
Re: Bipap not working. working as cpap...HELP
Posted: Fri Apr 24, 2015 8:52 am
by palerider
yaconsult wrote:Nice job figuring it out! So the machine wasn't able to sense what was happening properly - makes sense.
I always advise when doing jobs like this - grab your phone and take lots of pictures as you take things apart. If you don't need them, they are easy to delete. If you do need them, they can be invaluable!
I'm starting to wonder if I should check that wire on my low hour 550P as opposed to waiting for it to fail? Do we have any idea how common this problem is? If rare, I wouldn't bother. If it happens 50% of the time or more, it would probably be worth it. I have lots of experience with electronics so I'm comfortable working on stuff.
it's so easy to pop the cover of the sysone machines, I say, just do it... long shank t20, two screws, gently pull off one knob, and you're in. I'd unplug and replug all the connectors, maybe even take a pencil eraser to the posts to clean them...
call it 'preventative maintenance'.
there's no telling how prevalent this issue is, but it keeps getting repeated.... how many people simply replace their units without ever finding out that's what's wrong?
Re: Bipap not working. working as cpap...HELP
Posted: Fri Apr 24, 2015 7:49 pm
by mubiksski
My Machine was about 4.5 years old. Used most every night. But why fix it if it aint' broken? It is an easy repair thanks to this forum. I would wait. and pictures, which I do as well wouldn't have helped here. It was nothing you could see. I am just glad I got it working. I can't really afford a new one at this point.
Re: Bipap not working. working as cpap...HELP
Posted: Sat Apr 25, 2015 6:42 am
by JDS74
SGearhart wrote:Quick question. You stated you have a P-R 750, I thought the BiPAP model was a 760. With model is the BiPAP?
In case this hasn't been answered before:
The 750 is the 50 series Auto BiPap
The 760 is the 60 series Auto BiPap
They are very similar machines. The 60 series can accept a heated tube heated humidifier while the 50 series cannot.
There are likely other minor differences between the machines but that is the big one.
Re: Bipap not working/ fixed it/ Thx forum
Posted: Sat Apr 25, 2015 9:25 am
by Sir NoddinOff
Glad things worked out. Just a cautionary note on soldering ANYTHING in close proximity or adjacent to a printed circuit board - especially the board connector to various power sources (as in this case): First be sure to clamp an alligator clip (see link below), a hemostat (roach holder, for those with MJ experience) OR one or two of small needle-nose pliers (w/ a sturdy rubber band around the handles) around the metal area to be soldered. Also make them as close to the area soldered as reasonably possible. This trick will provide a heat sink and keep the heat from the soldering iron from going down into the circuit board and damaging the delicate printed wiring.
http://www.instructables.com/id/Save-yo ... heat-sink/