just a 'how high have I turned it up to".... unfortunately, it does not go to 11CzechCzar wrote:What does that scale measure? There were a couple scales with my new machine.
Would somebody be able to confirm that I should gradually increase the maximum pressure from my prescription of 6? From reading online, the machine should only go up as high as it needs to. If I set the maximum to 20, it will only go up that high if it needs to, correct?
In other words, am I losing out on potentially useful data by limiting my maximum pressure in this early stage?
yes, the machine will only go up as high as it finds it needs to in order to overcome snores, flow limitations, obstructive apneas and hypopneas (the latter two it increases pressure afterwards, because it can't do anything while they're happening, it simply can't generate enough pressure)
in general, the only time you have to worry about limiting the upper pressure is if you start having air swallowing (aerophagia) or other discomfort caused by too high pressure.
the lower number is the one that's more important for good results, if it's too low, you'll have events until the pressure gets up to where it needs to be, then the machine will start trying to lower the pressure back to your minimum set point, because that's where you told it to start.... so it's best to have the minimum pressure set close to where most of the events get treated... you'll figure that out over time by looking at the data .