Re: Help -- Auto BIPAP Settings
Posted: Sun Aug 24, 2014 7:54 pm
If you changed PS minimum to 0 the machine would function like an APAP and not a bipap. You wouldn't like that after using 5 PS for a while at your pressures.
Here's an example. My min EPAP is 10 cm....PS was 4 min and 5 max I think but it rarely did 5 and max IPAP was 20. I rarely hit max IPAP. The little saw tooth looking spikes that you see are the auto mode pressure probes...little increases just to test things a little. It's a normal function of both the Respironics auto adjusting pressures...be it bipap or apap.

That would be correct.Phil587 wrote:So, if I read you correctly this time, my settings as stated with PS min/max set at 5 should give me the results I desire -- a machine that will range from 12 to 17 on EPAP and 17 to 22 on IPAP and never separate by more than 5 CM3 and do that automatically as required. That's what I am after! If I were to change the min PS to zero, my machine might possibly run the same pressure IPAP and EPAP. GAK!
Here's an example. My min EPAP is 10 cm....PS was 4 min and 5 max I think but it rarely did 5 and max IPAP was 20. I rarely hit max IPAP. The little saw tooth looking spikes that you see are the auto mode pressure probes...little increases just to test things a little. It's a normal function of both the Respironics auto adjusting pressures...be it bipap or apap.