Did you forget and leave the SD card out of the machine last night during the blank time? Or maybe it didn't get stuck in the slot quite correctly?
We lose the detailed graphs if the SD card isn't in the S9 while in use.
I had that happen last Mon and Tues and the SD card was in the slot and I thought it was in correctly but I have no graphs at all for those 2 nights.
The event pattern looks suspicious for REM sleep stage pattern. If you wish to continue with the same EPR setting you likely will need just a little more minimum APAP pressure. Maybe 11 cm and I think the OAs and hyponeas will reduce a little more.
The centrals don't appear to be problematic with these pressures. I think we can put them on the back burner and just keep one eye on them.