As I understand it, the Airsense 10 uses some "gentle" puffing to detect whether the airway is open or not, so it knows how to score an event.
I notice this toward the end of my sleep period. It always happens just as I finish an exhale (not every exhale) - it's kind of disturbing. Does it make sense that it would be checking then, since I'm actually breathing at that point?
It also seems to happen if EPR is set to 2 or 3, but not 1. Then again, we're relying on my perception here, and not a lot of samples at EPR 1, at that - but I'm pretty certain. The only other possibility is that I actually exhale like that - which I don't think is the case.
Any ideas what this might be telling us? Does it suggest that I should be trying to use EPR=1 (or less) if I can stand it at the effective pressure (which can be a challenge)?
Thanks.
Airsense 10 "Puffing"?
Airsense 10 "Puffing"?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: EPAP 17, Max IPAP 25, PS 4, Sleepyhead software |
Re: Airsense 10 "Puffing"?
There are several possible reasons for this.
1) The machine tries to control the pressure at the mask, but its sensors are at the output of the blower and have to go through the humidifier, hose, and mask. I think sometimes, as you're switching between inhale and exhale, it's trying to be "smart" and may make some unintended variation at the other end.
2) If you have EPR turned on, it reduces the pressure as you exhale and ramps it back up when you inhale. You may feel this, and it may occasionally happen at the wrong time. You can turn off or reduce EPR if you want to.
You shouldn't feel the FOT (Forced Oscillation Technique) pulses unless you've stopped breathing for several seconds. FOT is what ResMed uses to detect a clear airway.
1) The machine tries to control the pressure at the mask, but its sensors are at the output of the blower and have to go through the humidifier, hose, and mask. I think sometimes, as you're switching between inhale and exhale, it's trying to be "smart" and may make some unintended variation at the other end.
2) If you have EPR turned on, it reduces the pressure as you exhale and ramps it back up when you inhale. You may feel this, and it may occasionally happen at the wrong time. You can turn off or reduce EPR if you want to.
You shouldn't feel the FOT (Forced Oscillation Technique) pulses unless you've stopped breathing for several seconds. FOT is what ResMed uses to detect a clear airway.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Airsense 10 "Puffing"?
resmeds us a technique called Forced Oscillation Technique (FOT) to determine if your airway is open or not, it does that by varying the pressure up and down by 1cm 4 times a second, after 6 seconds of not sensing any breathing (variation in airflow).cby1 wrote:As I understand it, the Airsense 10 uses some "gentle" puffing to detect whether the airway is open or not, so it knows how to score an event.
cby1 wrote:I notice this toward the end of my sleep period. It always happens just as I finish an exhale (not every exhale) - it's kind of disturbing. Does it make sense that it would be checking then, since I'm actually breathing at that point?
if you're still exhaling, and not pausing, then either you're not feeling FOT, or your machine is defective.
you can see FOT in the mask pressure trace on sleepyhead/resscan if you zoom in on it, it's pretty obvious.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Airsense 10 "Puffing"?
Sure sounds like the FOT you guys describe. I'll have to look at SH. It seems to me I have a "long" exhale - maybe it's not long, maybe it's stopped in these cases.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: EPAP 17, Max IPAP 25, PS 4, Sleepyhead software |
- SleepyBobR
- Posts: 312
- Joined: Tue Jan 19, 2010 4:42 pm
- Location: Toronto, Canada
Re: Airsense 10 "Puffing"?
I was concerned about this which was why I borrowed the AirSense for a week before buying it. I didn't want to find out I was bothered by the FOT after committing to the machine. I did not notice anything unusual on my first night but that's not much to go on. I'm not looking for a problem with the AirSense, just want to be sure there isn't one. So far so good.
_________________
Mask: Simplus Full Face CPAP Mask with Headgear |
Additional Comments: AirSense 10 AutoSet For Her @ 7-10 |
Re: Airsense 10 "Puffing"?
go in SH, zoom in on the flow and mask pressure to where you can see the individual breaths. if it's doing FOT, then you'll see it in the traces.cby1 wrote:Sure sounds like the FOT you guys describe. I'll have to look at SH. It seems to me I have a "long" exhale - maybe it's not long, maybe it's stopped in these cases.
fot kicks in six seconds after no flow variations from inhalation (flow trace above zero line) or exhalation (flow trace below zero line).