Mask carbon dioxide levels - toxicity
- chunkyfrog
- Posts: 34394
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Mask carbon dioxide levels - toxicity
I commend your enthusiasm, and I feel that your curious mind will help you with maintaining therapy;
even if it adds little new knowledge to the general pool--but who knows? You could hit on something important.
This just probably isn't it.
even if it adds little new knowledge to the general pool--but who knows? You could hit on something important.
This just probably isn't it.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Mask carbon dioxide levels - toxicity
Where is the manufacturer and the FDA hiding the bodies of the xpap users tgey kill with poor CO2 venting. For me if I used a pressure of 5cm, I'd be dead, I might as will put a plastic bag over my head. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Mask carbon dioxide levels - toxicity
BTW I think this is the user manual for the CO2 measuring device being ueds.
http://www.co2meters.com/Documentation/ ... Sensor.pdf
It appears to be a very interesting device.
http://www.co2meters.com/Documentation/ ... Sensor.pdf
It appears to be a very interesting device.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Mask carbon dioxide levels - toxicity
miltf,
I would suggest you increase your pressure from 5 to 6 & repeat your CO2 test. I would feel suffocated at 5cm and such a low pressure could be part of why you feel bad. I also don't think the CO2 is the real problem, if this were the case, as mentioned the mask manufacturing companies would have lawsuits coming out their ears by now.
Why not post some of your data here & see if our forum members can identify why you are feeling bad. Maybe you have undetected leaks, some untreated OSA due to such a low pressure or other anomaly our data experts might be able to help you with.
You have come to the right place for help, it just doesn't seem your theory is plausible. That said, I will try to keep an open mind. Also, welcome to the forum.
Jen
I would suggest you increase your pressure from 5 to 6 & repeat your CO2 test. I would feel suffocated at 5cm and such a low pressure could be part of why you feel bad. I also don't think the CO2 is the real problem, if this were the case, as mentioned the mask manufacturing companies would have lawsuits coming out their ears by now.
Why not post some of your data here & see if our forum members can identify why you are feeling bad. Maybe you have undetected leaks, some untreated OSA due to such a low pressure or other anomaly our data experts might be able to help you with.
You have come to the right place for help, it just doesn't seem your theory is plausible. That said, I will try to keep an open mind. Also, welcome to the forum.
Jen
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Pressure Auto 12-20cm CPAP compliant since 2000 |
Other masks I've tried: *=liked, #= no way
Piliaro, SleepWeaver Elan*, Swift FX w&w/o Bella Loops#, OpitLife#,Simplicity*, Mirage Vista*, Go Life for Her#, IQ (original hg only)*, Quattro FX (barely)###, Wisp*, Nuance#, Swift LT for her**
Piliaro, SleepWeaver Elan*, Swift FX w&w/o Bella Loops#, OpitLife#,Simplicity*, Mirage Vista*, Go Life for Her#, IQ (original hg only)*, Quattro FX (barely)###, Wisp*, Nuance#, Swift LT for her**
- TangledHose
- Posts: 862
- Joined: Wed Mar 13, 2013 11:13 am
- Location: Centennial, CO
Re: Mask carbon dioxide levels - toxicity
that's a very mature and productive post Todzo.......especially considering that in essence I was making a very similar point as you to the OP's post.Todzo wrote:It is interesting to see how if issues which deal with air hunger and carbon dioxide maintenance come up here the shills, trolls, and their sock puppets also appear. Sometimes you can tell them because they tend to use very gross avatars.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Simplus Full Face CPAP Mask with Headgear |
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited |
Link to Sleepyhead:
http://sleepfiles.com/SH2/
http://sleepfiles.com/SH2/
Re: Mask carbon dioxide levels - toxicity
miltf wrote: The S9 is set at 5 cm constant pressure.
I share the concern of a couple others about this low pressure. 5 cm would make me feel choked for air even before I fall asleep.
Do you have software with your CPAP and do you keep up with the detail reports about events and leak?
I am not a medical professional and I have no medical training.
Re: Mask carbon dioxide levels - toxicity
From the user manual for the Profile Lite:miltf wrote:Hi,
This is my first post here. I am a relatively new CPAP user (6 months). Initially, I found that the silicone in most masks gave me nasal congestion, so I switched to Sleepweaver masks which are cloth. No congestion, but I sometimes had a shortness of breath feeling. I thought maybe the fixed vent rate on the mask was too low as it felt like my expired breath was not leaving the mask before I took a new breath. Well, long story short, the company and I had an email exchange which remains unresolved for me and I went ahead and measured the CO2 in the mask using a very sensitive and rapid sensor called a SprintIR (details upon request).
The measured CO2 in the mask air space ranges from 2-3.3% during use (air is less than 0.1%, exhaled breath is 3-5%). The US OSHA sets a limit on CO2 of 1% (time weighted average) and says that higher levels can lead to discomfort (!), headache, lethargy, etc. Levels above 3% are not allowed for any length of time. I made some holes in the mask wall to increase the vent rate and decrease the CO2 and, sure enough, my nights are far more comfortable. My ResMed S9 Autoset seems to adjust very well to this change and for the most part, this problem is solved for me.
However, I am still concerned about other users of CPAP. I have measured the two other masks I happened to have, a Respironics Profile Lite (2-3%) and a ResMed Swift FX (1.5-2.5%) and both of these have high CO2 levels. My guess is that most masks are not designed to achieve adequate venting of CO2, partly because of inadequate vent rate but mostly because of vent placement. I think this must contribute to the low compliance rate for CPAP among those who are prescribed this technology. After all, if you are feeling discomfort, headache and lethargy, you might as well just stick with OSA.
I'm putting this information out on this forum because I don't know what else to do with it. I'll be interested to hear your thoughts.
Milt
Warning:
This warning applies to most CPAP devices. At low CPAP pressures, the air flow through the exhalation port
may not be enough to clear all of the exhaled gas (CO2) from the mask. You may breathe in some of the air that
you have exhaled. Do not block or otherwise try to seal the vent holes on the exhalation port.
From the user manual for the Swift FX:
Warning:
As with all masks, some rebreathing may occur at low CPAP pressures.
I think the definition of "LOW" is conditional upon each individual's breathing. But, I think 4 - 6 is definitely in the "low" category for nearly everyone.
Den
.
Re: Mask carbon dioxide levels - toxicity
Start with a pressure of 10 like I do every night then see if you have any CO2 in the mask. I doubt you will find any.
No Fate But what we Make...
Sleep Apnea Sucks...
Sleep Apnea Sucks...
Re: Mask carbon dioxide levels - toxicity
So anyone who disagrees with you is a shill, troll, or sock puppet? Hmm.Todzo wrote:It is interesting to see how if issues which deal with air hunger and carbon dioxide maintenance come up here the shills, trolls, and their sock puppets also appear. Sometimes you can tell them because they tend to use very gross avatars.
And claiming that folks who use gross avatars fit the above description makes about as much sense as my claiming that I can paint as well as Monet.
49er
_________________
Mask: SleepWeaver Elan™ Soft Cloth Nasal CPAP Mask - Starter Kit |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Use SleepyHead |
Re: Mask carbon dioxide levels - toxicity
And wasn't Tuesday the day that he wasn't supposed to be an asshole?49er wrote:So anyone who disagrees with you is a shill, troll, or sock puppet? Hmm.Todzo wrote:It is interesting to see how if issues which deal with air hunger and carbon dioxide maintenance come up here the shills, trolls, and their sock puppets also appear. Sometimes you can tell them because they tend to use very gross avatars.
And claiming that folks who use gross avatars fit the above description makes about as much sense as my claiming that I can paint as well as Monet.
49er
Is just one day too much to ask?
You Kids Have Fun!!
Re: Mask carbon dioxide levels - toxicity
Oh right, back in day:Wulfman... wrote:Another discussion I was also part of about 7 years ago had to do with certain masks also not sufficiently clearing CO2.
viewtopic.php?f=1&t=11458&p=174261&hili ... ro#p174261
http://erj.ersjournals.com/content/23/1 ... e90504ad98
And before that, back in the frequie days, when the fad was trying to increase FICO2 levels.
I tell ya, if you stand in one spot long enough, what goes around...
You Kids Have Fun!!
- Sheriff Buford
- Posts: 4082
- Joined: Mon Aug 09, 2010 8:01 am
- Location: Kingwood, Texas
Re: Mask carbon dioxide levels - toxicity
Todzo wrote:Sometimes you can tell them because they tend to use very gross avatars.
Hey... hey.... look at my mug.... tell me its gross....
_________________
Machine: AirSense 11 Autoset |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Re: Mask carbon dioxide levels - toxicity
Ain't that the truth:Todzo wrote:Sometimes you can tell them because they tend to use very gross avatars.
You Kids Have Fun!!
Re: Mask carbon dioxide levels - toxicity
i would call the picture a lie but then it would be changed
much of what shills trolls and sockpuppets do is to manipulate
gross avatars are simply a part of that
much of what shills trolls and sockpuppets do is to manipulate
gross avatars are simply a part of that
May any shills trolls sockpuppets or astroturfers at cpaptalk.com be like chaff before the wind!
Re: Mask carbon dioxide levels - toxicity
Carbon dioxide maintenance is a very real issue with CPAP use however. It is simply that the volume of the mask is too small to matter.
Todzo wrote:The actual volumes inside the mask are low so you will be rebreathing very little additional air (keep in mind that we do not totally purge our systems and do rebreath some air as part of normal healthy breathing). And indeed a more common problem with CPAP use is carbon dioxide maintenance so some have come up with ways to help this which include enhancing the rebreathing space[1] or adding carbon dioxide at critical phases in the breathing process[2].
If the mask vent is working properly I would not worrry about it.
[1]: Gilmartin G, McGeehan B, Vigneault K, Daly RW, Manento M, Weiss JW, Thomas RJ.
Treatment of positive airway pressure treatment-associated respiratory instability with enhanced expiratory rebreathing space (EERS).
Source: J Clin Sleep Med. 2010 Dec 15;6(6):529-38. Division of Pulmonary, Critical Care and Sleep Medicine, Beth Israel Deaconess Medical Center, Boston, MA, USA.
Link: http://www.ncbi.nlm.nih.gov/pubmed/21206741
[2]: Dynamic CO2 therapy in periodic breathing: a modeling study to determine optimal timing and dosage regimes
Yoseph Mebrate, Keith Willson, Charlotte H. Manisty, Resham Baruah, Jamil Mayet, Alun D. Hughes, Kim H. Parker and Darrel P. Francis
J Appl Physiol 107:696-706, 2009. First published 23 July 2009; doi: 10.1152/japplphysiol.90308.2008
Link: http://www.ncbi.nlm.nih.gov/pubmed/19628721