saturation C02 level's drop while trying to use
- Indi's Gmomma
- Posts: 29
- Joined: Fri Mar 21, 2014 7:46 pm
saturation C02 level's drop while trying to use
Im hoping someone here can help me, I went back to see the drs that got me the trilogy, I have not yet had a sleep study because my need is mainly I'm a c02 retainer or I have neromuscular pulmanory failure and am hypercapnic, the dr increased my trilogy from 12/8 to now 18/10 rr15.
I use C02 24/7 @ 2.5lpm and my C02 during sleep is now bumped to 4 LPG.
My big problem is my saturation drops below 70 as soon as I start to fall asleep, I thought it was all because of leaks but I think I have those all fixed for now. My husband sat up and watched my saturation level and each time we tried I quickly dropped below 70 so we aborted what am I doing wrong? I feel like I'm not getting my last breath really out before my next one is being forced down me, and does it make any difference where the oxygen line enters? We have tried the back of the trilogy machine and also on the tubing itself, at both ends. Has anyone had low saturation levels while using there non- invasive vent? Any input is appreciated, I hoping Madalot will see this.
TIA all
I use C02 24/7 @ 2.5lpm and my C02 during sleep is now bumped to 4 LPG.
My big problem is my saturation drops below 70 as soon as I start to fall asleep, I thought it was all because of leaks but I think I have those all fixed for now. My husband sat up and watched my saturation level and each time we tried I quickly dropped below 70 so we aborted what am I doing wrong? I feel like I'm not getting my last breath really out before my next one is being forced down me, and does it make any difference where the oxygen line enters? We have tried the back of the trilogy machine and also on the tubing itself, at both ends. Has anyone had low saturation levels while using there non- invasive vent? Any input is appreciated, I hoping Madalot will see this.
TIA all
Last edited by Indi's Gmomma on Fri Apr 25, 2014 12:08 pm, edited 1 time in total.
- Indi's Gmomma
- Posts: 29
- Joined: Fri Mar 21, 2014 7:46 pm
Re: saturation C02 level's drop while trying to use trilogy
Anyone here use oxygen with there cpap? Or bipap? I'm interested in talking about any success?
Re: saturation C02 level's drop while trying to use
Yes. I use 2.5 LPM into an adapter at the mask. This generally keeps my nadir above 88 and my average around 92 or 93.
Without the O2 I would drop to the low 80s and occasionally the high 70s.
I am confused by your reference to CO2 rather than O2.
Without the O2 I would drop to the low 80s and occasionally the high 70s.
I am confused by your reference to CO2 rather than O2.
_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear |
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i |
- BlackSpinner
- Posts: 9745
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: saturation C02 level's drop while trying to use
You may want to get in touch (PM) the member Madalot she uses both O2 and a trilogy.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear |
Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
- Indi's Gmomma
- Posts: 29
- Joined: Fri Mar 21, 2014 7:46 pm
Re: saturation C02 level's drop while trying to use
I guess I'm not very well versed in all the lingo I'm new to all this I was only referring to my oxygen, thanks for the correction, I learn something new daily my problem is when I try to go to sleep my saturation level drops fast and we stop and I go back on my cannula at 2.5lp.bwexler wrote:Yes. I use 2.5 LPM into an adapter at the mask. This generally keeps my nadir above 88 and my average around 92 or 93.
Without the O2 I would drop to the low 80s and occasionally the high 70s.
I am confused by your reference to CO2 rather than O2.
My dr just changed my setting and also told me to bump my 02 to 4 lpm when hooked up! Clearly something is wrong, my setting on my bipap were 12/8 and auto for something else but this trilogy the dr increased my presure setting to 18/10 and a rr15 I feal like I'm not getting my last breath out before my next one comes and it prevents the fresh 02 from making it to my lungs if this makes sense?
- Indi's Gmomma
- Posts: 29
- Joined: Fri Mar 21, 2014 7:46 pm
Re: saturation C02 level's drop while trying to use
Thanks I will pm Madalot also.BlackSpinner wrote:You may want to get in touch (PM) the member Madalot she uses both O2 and a trilogy.
-
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: saturation C02 level's drop while trying to use
I have to say that this doesn't make any sense at all and you are justified to be concerned.
Looking at some possibilities...
When you go to sleep you stop breathing? If not what is your respiration rate? and what is a "normal" respiration rate?
Your mask set up may be allowing you to re-breath some of your air? I don't think this is going on but it needs to be eliminated. Are your mask vents open and working?
When you go to sleep you loose circulation to your hands? This would render the oximeter results from a finger mounted unit inaccurate. Perhaps an earlobe unit would tell a different story?
Your oxygen supply is corrupt and your oxygen metering equipment is malfunctioning? This is also highly unlikely but should be checked.
At any rate you need to contact your doctor and report that you are seeing your oxygen levels drop to 70% even at higher O2 flow rates. Tell them you think 70% may be a "little on the low side", and you feel like crap. That should get their attention.
Looking at some possibilities...
When you go to sleep you stop breathing? If not what is your respiration rate? and what is a "normal" respiration rate?
Your mask set up may be allowing you to re-breath some of your air? I don't think this is going on but it needs to be eliminated. Are your mask vents open and working?
When you go to sleep you loose circulation to your hands? This would render the oximeter results from a finger mounted unit inaccurate. Perhaps an earlobe unit would tell a different story?
Your oxygen supply is corrupt and your oxygen metering equipment is malfunctioning? This is also highly unlikely but should be checked.
At any rate you need to contact your doctor and report that you are seeing your oxygen levels drop to 70% even at higher O2 flow rates. Tell them you think 70% may be a "little on the low side", and you feel like crap. That should get their attention.
_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...
Re: saturation C02 level's drop while trying to use
How did your husband watch your saturation level? What was he watching?Indi's Gmomma wrote:My husband sat up and watched my saturation level and each time we tried I quickly dropped below 70 so we aborted what am I doing wrong? I feel like I'm not getting my last breath really out before my next one is being forced down me, and does it make any difference where the oxygen line enters? We have tried the back of the trilogy machine and also on the tubing itself, at both ends. Has anyone had low saturation levels while using there non- invasive vent? Any input is appreciated, I hoping Madalot will see this.
TIA all
_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7 |
- Indi's Gmomma
- Posts: 29
- Joined: Fri Mar 21, 2014 7:46 pm
Re: saturation C02 level's drop while trying to use
Madalot wrote:How did your husband watch your saturation level? What was he watching?Indi's Gmomma wrote:My husband sat up and watched my saturation level and each time we tried I quickly dropped below 70 so we aborted what am I doing wrong? I feel like I'm not getting my last breath really out before my next one is being forced down me, and does it make any difference where the oxygen line enters? We have tried the back of the trilogy machine and also on the tubing itself, at both ends. Has anyone had low saturation levels while using there non- invasive vent? Any input is appreciated, I hoping Madalot will see this.
TIA all
Yes hubby stayed up and watched my larger Drive Oximeter (model #18715) with the alarm turned off so I could fall asleep. I do have neuropathy in my hands and feet from Gullian Berré but Hosp confirmed my low sat with abg's (arterial blood gass)
- The Latinist
- Posts: 465
- Joined: Sat Apr 19, 2014 10:00 pm
Re: saturation C02 level's drop while trying to use
Have you seen a pulmonologist? This strikes me as an issue that's well above our pay grade, so to speak.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: APAP 12-16 cmH2O, EPR 1. Untreated AHI: 96; treated AHI 2.3. |
Re: saturation C02 level's drop while trying to use
You have told me via PM that your Trilogy is set up in S/T Mode with AVAPS (you read it off the screen).
I use the same mode with AVAPS as well.
I would be interested in your specific AVAPS settings:
Min IPAP
Max IPAP
EPAP
Tidal Volume
If anyone with more experience comes in, they'll want to know what these settings are.
I use the same mode with AVAPS as well.
I would be interested in your specific AVAPS settings:
Min IPAP
Max IPAP
EPAP
Tidal Volume
If anyone with more experience comes in, they'll want to know what these settings are.
_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7 |
- Indi's Gmomma
- Posts: 29
- Joined: Fri Mar 21, 2014 7:46 pm
Re: saturation C02 level's drop while trying to use
I just saw a Pulmoligist but basically only to get this trilogy going, at this point I have not been tested for sleep apnea or for a sleep study although the Pulm ordered it (to be scheduled)
Dr marked under bi-level pressure settings 18/10 RR15 oxygen 4 lpm I copied this from the form I gave my rt.
Before I the script said I??? 12 EPAP 8 Eauto? Oxygen 2 auto timing m mode b? pap (?=can't make out)
What does the above all stand for?
Thanks to everyone that helps
Dr marked under bi-level pressure settings 18/10 RR15 oxygen 4 lpm I copied this from the form I gave my rt.
Before I the script said I??? 12 EPAP 8 Eauto? Oxygen 2 auto timing m mode b? pap (?=can't make out)
What does the above all stand for?
Thanks to everyone that helps
Re: saturation C02 level's drop while trying to use
I am concerned that AVAPS is enabled but you haven't been told what those settings are. I hope whoever set it up knew what they were doing with it.Indi's Gmomma wrote:I just saw a Pulmoligist but basically only to get this trilogy going, at this point I have not been tested for sleep apnea or for a sleep study although the Pulm ordered it (to be scheduled)
Dr marked under bi-level pressure settings 18/10 RR15 oxygen 4 lpm I copied this from the form I gave my rt.
Before I the script said I??? 12 EPAP 8 Eauto? Oxygen 2 auto timing m mode b? pap (?=can't make out)
What does the above all stand for?
Thanks to everyone that helps
It ALMOST looks like they ordered Auto EPAP, but according to what you told me, you are NOT using AVAPS-AE (just plain AVAPS like I am).
_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7 |
Re: saturation C02 level's drop while trying to use
These settings mean:Indi's Gmomma wrote: Dr marked under bi-level pressure settings 18/10 RR15 oxygen 4 lpm I copied this from the form I gave my rt.
IPAP = 18 cm H20; the pressure during the inhalations is set to 18 cm H20, which is sort of high, but not exceptionally high.
EPAP = 10 cm H20; the pressure during the exhalations is set to 10 cm H20.
RR = 15; the back up respiratory rate is set to a fixed 15 breaths per minute. If your RR drops below that the machine kicks in and switches from EPAP to IPAP in an attempt to trigger an inhalation.
O2 = 4 L/min; the oxygen is supposed to be bled into the system at a rate of 4 Liters per minute. That's on the high side, probably because you're still dealing with some serious O2 desats.
This script has the settings set at:Before I the script said I??? 12 EPAP 8 Eauto? Oxygen 2 auto timing m mode b? pap (?=can't make out)
IPAP = 12 cm H20; the pressure during the inhalations is set to 12 cm H20
EPAP = 8 cm H20; the pressure during the exhalations is set to 8 cm H20.
RR = appears to be set to Auto; the back up respiratory rate determined on the fly by the machine based on the last several minutes of "normal" breathing. If your RR drops below that automatically determined RR, then the machine kicks in and switches from EPAP to IPAP in an attempt to trigger an inhalation.
O2 = 2 L/min; the oxygen is supposed to be bled into the system at a rate of 2 Liters per minute.
****************
The two scripts are VERY different from each other. You write:
Any idea on what the doc used to determine the new settings?Im hoping someone here can help me, I went back to see the drs that got me the trilogy, I have not yet had a sleep study because my need is mainly I'm a c02 retainer or I have neromuscular pulmanory failure and am hypercapnic, the dr increased my trilogy from 12/8 to now 18/10 rr15.
You also write:
and[With the new settings] I feel like I'm not getting my last breath really out before my next one is being forced down me
It sounds like the RR = 15 breaths per minute is set high enough where you are simply not "keeping up" with the machine's desire to trigger an inhalation every 4 seconds. You need to inform the doctor about how you feel like you are not able to fully exhale with these settings. It's also worth asking the doc why the RR is set to 15 and whether you might be more comfortable if the RR was set a wee bit lower.[With the new settings] I feal like I'm not getting my last breath out before my next one comes
Finally you write:
It is critically important to have the O2 fed into the system in the correct place. The durable equipment provider (DME) that provided your Trilogy should have set this up for you and should have taught you and your hubby how and where to attach the oxygen line.and does it make any difference where the oxygen line enters? We have tried the back of the trilogy machine and also on the tubing itself, at both ends.
Maybe part of your on-going low O2 saturation problem is due to the O2 line not being attached to the system correctly.
I'd call the DME and tell them that you and your hubby are confused about how and where to attach the O2 line into the Trilogy system and that you need help with that immediately. If they can't tell you or send someone out to help you out, then you need to call the doc's office and tell them that you don't know where the O2 should be attached to the Trilogy system AND that the DME can't seem to help you either.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: saturation C02 level's drop while trying to use
Robysue --
Excellent input and suggestions. I know in my situation, if they had my backup rate to 15, I'd be hyperventilating. Anything above 7 for me is rushed. My last sleep study, the tech bumped it up to 8 and I knew immediately he had done it because it woke me up.
And there is a port on the back of the Trilogy, lower right hand side (as you're looking at the back) for the oxygen to be attached.
Thanks for chiming in on this, Robysue! I'm glad you're onboard to help the OP.
Excellent input and suggestions. I know in my situation, if they had my backup rate to 15, I'd be hyperventilating. Anything above 7 for me is rushed. My last sleep study, the tech bumped it up to 8 and I knew immediately he had done it because it woke me up.
And there is a port on the back of the Trilogy, lower right hand side (as you're looking at the back) for the oxygen to be attached.
Thanks for chiming in on this, Robysue! I'm glad you're onboard to help the OP.
_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7 |