So my 2 month old System one humidifier broke last night. The light is flashing and it's not heating up. I tried the unplug and plug back in thing and it didn't work.
I called my DME and the receptionist confirmed it was under warranty, but I had to schedule an appointment with the RT to get it looked at. She can't see me until Wednesday! I guess I have a dry throat and compromised therapy until then. Ridiculous. I should be able to go up there, give them the broken one and take a new one home today.
Humidifier broke
Humidifier broke
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Simplus Full Face CPAP Mask with Headgear |
Re: Humidifier broke
Put some water in the tank anyway. With the air moving over the water it will pick up some moisture even if it isn't heating up.
Not a lot but better than nothing.
Just another reason why we have spare equipment. DMEs often don't go out of their way to help you out...sad but true.
Not a lot but better than nothing.
Just another reason why we have spare equipment. DMEs often don't go out of their way to help you out...sad but true.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Humidifier broke
So I'm finding. I bought my own machine because of trouble with the DME, but I'm still using the humidifier they gave me. Since I paid for it, and it's under warranty I expect them to replace it.
They're more trouble then they worth.
They're more trouble then they worth.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Simplus Full Face CPAP Mask with Headgear |
- StuUnderPressure
- Posts: 1377
- Joined: Mon Jun 18, 2012 10:34 am
- Location: USA
Re: Humidifier broke
Your post should come as a Wake Up Call for those who think that if anything goes wrong with their machine, their DME will replace it overnight.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Cleanable Water Tub & Respironics Premium Chinstrap |
In Windows 10 Professional 64 bit Version 22H2 - ResScan Version 7.0.1.67 - ResScan Clinician's Manual dtd 2021-02
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
SD Card Formatter 5.0.2 https://www.sdcard.org/downloads/format ... index.html
Re: Humidifier broke
Did you actually disconnect the humidifier from the XPAP, or did you just unplug the machine from the power?alpha752 wrote:So my 2 month old System one humidifier broke last night. The light is flashing and it's not heating up. I tried the unplug and plug back in thing and it didn't work.
I called my DME and the receptionist confirmed it was under warranty, but I had to schedule an appointment with the RT to get it looked at. She can't see me until Wednesday! I guess I have a dry throat and compromised therapy until then. Ridiculous. I should be able to go up there, give them the broken one and take a new one home today.
Are there any error messages in the LCD screen?
Den
.
Re: Humidifier broke
I did both, disconnected it from the machine, and unplugged the machine, several times. I tried to run the preheat with out the chamber in, and the plate doesn't warm up, the light just flashes. There are no messages on the screen, just the light flashing. I tried to turn off system one in the settings, and get the same results.Wulfman... wrote:Did you actually disconnect the humidifier from the XPAP, or did you just unplug the machine from the power?alpha752 wrote:So my 2 month old System one humidifier broke last night. The light is flashing and it's not heating up. I tried the unplug and plug back in thing and it didn't work.
I called my DME and the receptionist confirmed it was under warranty, but I had to schedule an appointment with the RT to get it looked at. She can't see me until Wednesday! I guess I have a dry throat and compromised therapy until then. Ridiculous. I should be able to go up there, give them the broken one and take a new one home today.
Are there any error messages in the LCD screen?
Den
.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Simplus Full Face CPAP Mask with Headgear |
Re: Humidifier broke
How about if you completely disconnect the HH unit from the CPAP? Any errors or flashing lights?
I don't have a service manual for that machine, so I'm not much help in this regard.
Den
.
I don't have a service manual for that machine, so I'm not much help in this regard.
Den
.
Re: Humidifier broke
If the humidifier icon lights are blinking, More than likely it's the connection between the pap and the humidifier, not the humidifier. Make sure the dme switches out both components.