Phillips Respronics PR System one Question
My Phillips System One machine is running in the A-Flex mode with the "Auto On" enabled, "Auto Off" disabled.
About 6 times a month when I first put my mask on it will start when I breathe, then shut down for 1 or 2 seconds and restart 2 or 3 times, then run normal for the rest of the night. My machine is 3 years old and I just started doing this.
Things I've tried: Switching masks; changing hose; Trying new humidifier tank.
Has anyone else had this problem?
Rich
Phillips System one PR Machine
Mirage Quattro Mask
Max pressure 13
Min pressure 10
About 6 times a month when I first put my mask on it will start when I breathe, then shut down for 1 or 2 seconds and restart 2 or 3 times, then run normal for the rest of the night. My machine is 3 years old and I just started doing this.
Things I've tried: Switching masks; changing hose; Trying new humidifier tank.
Has anyone else had this problem?
Rich
Phillips System one PR Machine
Mirage Quattro Mask
Max pressure 13
Min pressure 10
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Additional Comments: A-Flex Mode: 10cm/H2O min. 13cm/H2O max. pressure |
Re: Phillips Respronics PR System one Question
I have never had that problem, but, I think my machine is just about to be 3 years old.
- Bama Rambler
- Posts: 189
- Joined: Wed Feb 19, 2014 10:25 am
Re: Phillips Respronics PR System one Question
What happens if you manually start the machine?
What happens if you turn off "Auto Off"?
What happens if you turn off "Auto Off"?
Re: Phillips Respronics PR System one Question
I'm going to try turning off "Auto On" and manually starting the machine although you loose the auto start after a power interruption ability.
The Auto Off is always off.
The Auto Off is always off.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Additional Comments: A-Flex Mode: 10cm/H2O min. 13cm/H2O max. pressure |
Re: Phillips Respronics PR System one Question
Sounds like an equipment malfunction on the breath measuring circuit or sensor.
Do you have a backup machine? If so, you might consider sending this one to the ACCBIO folks for repair. They do a good job and are quite reasonable in pricing. With the machine being 3 years old, all warranty time has expired.
I would not, under any circumstances, consider turning the Auto On function off just because of power glitches, etc. A momentary power fluctuation could turn your machine off and then it would stay off.
Do you have a backup machine? If so, you might consider sending this one to the ACCBIO folks for repair. They do a good job and are quite reasonable in pricing. With the machine being 3 years old, all warranty time has expired.
I would not, under any circumstances, consider turning the Auto On function off just because of power glitches, etc. A momentary power fluctuation could turn your machine off and then it would stay off.
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Phillips Respronics PR System one Question
have you tried a different ac outlet
and checked the wire connections at the dc power adapter
make sure they are seated well
have been known to cause similar problems when loose
and checked the wire connections at the dc power adapter
make sure they are seated well
have been known to cause similar problems when loose
Re: Phillips Respronics PR System one Question
I've started having similar issues with my PRS1, also 3 years old. It turns itself on and off for no apparent reason. I don't have auto-off enabled. I recently changed outlets and so far so good. My humidifier isn't working, either. I got a new one but it seems to be an issue with the blower unit as the new humidifier didn't help. Time to fight with Crapria over the service.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Phillips Respronics PR System one Question
I tried all the suggestions... Thanks
Finally ended up removing the cover and found a loose connector going from the PC board to below the blower unit.
I repair industrial electronics for a living so I wasn't too scared
So far so good.
Finally ended up removing the cover and found a loose connector going from the PC board to below the blower unit.
I repair industrial electronics for a living so I wasn't too scared
So far so good.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Additional Comments: A-Flex Mode: 10cm/H2O min. 13cm/H2O max. pressure |
Re: Phillips Respronics PR System one Question
I had same issue for about 1 yr. It seems tha problem is eliminated temporarily after removing machine cover and unplug the electrical connectors and reinsert again. But it will come back after a few weeks. Is this a fiwmware issue - does is programed to create this fault and could only clear by company?
Re: Phillips Respronics PR System one Question
This is probably a silly question but what is the purpose of those functions? Is it just convenience or is there some therapeutic value?
When I had my old machine Kaiser set it up to use auto on and off. I thought it was cool but that was about it. Now my System One rental has both those functions disabled (not with Kaiser anymore).
Thanks!
When I had my old machine Kaiser set it up to use auto on and off. I thought it was cool but that was about it. Now my System One rental has both those functions disabled (not with Kaiser anymore).
Thanks!
Re: Phillips Respronics PR System one Question
For me it is partially therapeutic and partially convenient. If there is a short power interruption, the machine will go off. With Auto On enabled, it will restart. I don't have Auto Off enabled but it would allow for just taking the mask off if you had to get up during the night and the machine would shut itself off.jaye8898 wrote:This is probably a silly question but what is the purpose of those functions? Is it just convenience or is there some therapeutic value?
_________________
Mask: Oracle HC452 Oral CPAP Mask |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Phillips Respronics PR System one Question
The system will restart after a power loss as long as the power loss is less than a few minutes regardless of the auto on being turned on or not.