New User... Cheyne Stokes reading has me concerned

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Dozer_Ont
Posts: 5
Joined: Tue Aug 11, 2015 8:17 am

New User... Cheyne Stokes reading has me concerned

Post by Dozer_Ont » Tue Aug 11, 2015 8:32 am

Hi There,

Been on the APAP (Respironics) for just under a week now. I've been using Sleepyhead to analyze my results, though will not "wing it" with the pressure settings without doctor approval. I also know it's early in the process, so the numbers haven't had a chance to average out... though my sleep study showed an AHI of 85.1, and last night I had my lowest yet (AHI 3.95, 6 day average = 5.83). I also know that AHI is an imperfect measurement (kinda like calories, but hey gotta measure something).

My concern is that the machine is showing some Cheyne-Stokes respirations. Dr. Google has me a little worried with the whole... this is how you breath, just before you die stuff.

While I am not looking for device adjustment advice, I would appreciate some insight into these readings... Thanks!

BTW, my average time in CSR for the 6 days of using the APAP is 0.37%. That seems low to me, but I only have a medical degree from Google U.


THANKS!

--Dozer_Ont

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: New User... Cheyne Stokes reading has me concerned

Post by icipher » Tue Aug 11, 2015 8:49 am

Dozer_Ont wrote:Hi There,

Been on the APAP (Respironics) for just under a week now. I've been using Sleepyhead to analyze my results, though will not "wing it" with the pressure settings without doctor approval. I also know it's early in the process, so the numbers haven't had a chance to average out... though my sleep study showed an AHI of 85.1, and last night I had my lowest yet (AHI 3.95, 6 day average = 5.83). I also know that AHI is an imperfect measurement (kinda like calories, but hey gotta measure something).

My concern is that the machine is showing some Cheyne-Stokes respirations. Dr. Google has me a little worried with the whole... this is how you breath, just before you die stuff.

While I am not looking for device adjustment advice, I would appreciate some insight into these readings... Thanks!

BTW, my average time in CSR for the 6 days of using the APAP is 0.37%. That seems low to me, but I only have a medical degree from Google U.


THANKS!

--Dozer_Ont
Are you using sleepyhead software to track your data? If so, I wouldn't be too worried. The sleepyhead designer/developer decided to use "CSR" for all forms of periodic breathing. CSR is only one form of periodic breathing, and even if you were having 0.36% per night, I don't know that I would be concerned with that myself.

If you want to provide some zoomed in images of your "CSR" graph, we can further examine it and hopefully provide you some relief.

Dozer_Ont
Posts: 5
Joined: Tue Aug 11, 2015 8:17 am

Re: New User... Cheyne Stokes reading has me concerned

Post by Dozer_Ont » Tue Aug 11, 2015 9:08 am

Thanks for the quick reply and the reassurance. It didn't seem like a high number, and the events are few and far between. There's even been a couple nights without a single CSR. Some of them also coincide with times just after I had a bathroom break and was falling back asleep. Also noticed the bulk of VS and VS2 readings are from the times that I am falling asleep.

As for the graph, which one would it help to see? The zoomed in Event Flags?

icipher
Posts: 304
Joined: Thu Oct 10, 2013 11:09 am

Re: New User... Cheyne Stokes reading has me concerned

Post by icipher » Tue Aug 11, 2015 9:23 am

Dozer_Ont wrote:Thanks for the quick reply and the reassurance. It didn't seem like a high number, and the events are few and far between. There's even been a couple nights without a single CSR. Some of them also coincide with times just after I had a bathroom break and was falling back asleep. Also noticed the bulk of VS and VS2 readings are from the times that I am falling asleep.

As for the graph, which one would it help to see? The zoomed in Event Flags?

Yes, if you are using sleephead, you can zoom in on the "CSR" events.

Should look something like this...

Image

User avatar
deerslayer
Posts: 1193
Joined: Tue Dec 11, 2007 10:46 am
Location: Mid Tn

Re: New User... Cheyne Stokes reading has me concerned

Post by deerslayer » Tue Aug 11, 2015 9:30 am

Welcome Dozer !
When you get a chance list your mask,machine,etc.

tim

_________________
MachineMask
Additional Comments: straight cpap 12.20 AHI avg.0.79
RESMED AUTOSET 10 / P 10 NASAL PILLOWS /straight cpap 12.20 / AHI 0.79 avg /
thanks to Nick sleep HQ
Feeling Blessed & firmly believe in The Holy Trinity 🙏

Dozer_Ont
Posts: 5
Joined: Tue Aug 11, 2015 8:17 am

Re: New User... Cheyne Stokes reading has me concerned

Post by Dozer_Ont » Tue Aug 11, 2015 9:53 am

This image is from last night:

Image



Philips Respironics (v3.06) with humidifier (heated)
Wisp Nasal Mask


Thanks for your opinion, much appreciated!

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New User... Cheyne Stokes reading has me concerned

Post by palerider » Tue Aug 11, 2015 10:08 am

Dozer_Ont wrote:This image is from last night:

Image
Philips Respironics (v3.06)
several things,
1) the v3.06 is the software version and doesn't tell which of the many many machines it is, and many behave differently, look on the bottom for a REF number.
2) your machine *DOES NOT REPORT* CSR. it reports periodic breathing, you can go in and change the label in the menu options in sleepyhead,
3) https://sleep.tnet.com/resources/sleepyhead/shorganize read these *THREE* pages which will show you how to arrange your sleepyhead charts best, take a proper screenshot, AND post it to imgur so that it works properly. links at the bottom of each page to go to the next.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 64178
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New User... Cheyne Stokes reading has me concerned

Post by Pugsy » Tue Aug 11, 2015 10:18 am

Respironics machines report Periodic Breathing which is just a general waxing and waning of the breathing pattern. Cheyne Stokes Respiration is one of many forms of Periodic Breathing..not the only form of PB.
I suggest that you do what PR says...rename the CSR events to PB or just ignore unless you see large amounts of PB that really looks like CSR.

Your breathing pattern looks nothing like CSR...it looks like PB but even if it was CSRish looking..that small of an amount wouldn't really be all that alarming.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Dozer_Ont
Posts: 5
Joined: Tue Aug 11, 2015 8:17 am

Re: New User... Cheyne Stokes reading has me concerned

Post by Dozer_Ont » Tue Aug 11, 2015 11:22 am

Thanks, both of you!

I'll upload the zoomed in shot as directed once I get home tonight. My workplace blocks the imgur site. I'll follow the instructions you've provided.


Glad to hear that it's not likely a concern. I'm due to return to have the machine recalibrated next week and will ask more about the PB. Based on the AHI reduction from 85.1 (sleep study) to 3.95 (last night), I am cautiously optimistic. And sitting at my desk 30 minutes after lunch and not drifting off to sleep, that's gotta be a good sign!

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New User... Cheyne Stokes reading has me concerned

Post by palerider » Tue Aug 11, 2015 1:11 pm

Dozer_Ont wrote:Glad to hear that it's not likely a concern. I'm due to return to have the machine recalibrated next week and will ask more about the PB.
oh, one more thing, the amount you reported < 1% is completely insignificant.

now, on the other hand, if your chart looked like THIS:

Image

then the answer would be RUN (or at least hobble) as fast as possible to a cardiologist.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jdr999
Posts: 175
Joined: Wed Mar 06, 2013 12:59 pm
Location: New Jersey

Re: New User... Cheyne Stokes reading has me concerned

Post by jdr999 » Tue Aug 11, 2015 10:44 pm

Dozer_Ont wrote:Dr. Google has me a little worried...
I just had to post this. Nothing to do with your original question, just a bit of levity.

Don't stop asking questions or finding answers -- just keep this image in the back of your mind and smile

Image

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV, EPAP 8-15 / PS 5-10, Airfit P10, Sleepyhead MAC

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: New User... Cheyne Stokes reading has me concerned

Post by JohnBFisher » Wed Aug 12, 2015 11:37 am

Shoot! Googling your symptoms even when you feel well is a good way to convince yourself that death is stalking you!!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński