Question about Lincare and doctor's orders

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Country4ever
Posts: 1373
Joined: Wed Oct 31, 2007 6:22 pm

Question about Lincare and doctor's orders

Post by Country4ever » Wed Feb 12, 2014 10:49 am

I took my prescription for a new cpap machine to Lincare. I had decided that would be the cheapest way to go.......but not the easiest.

Here's my question.........Do they ever "fix" the machine so you can't accessess the clinical stuff? My doc wrote for a Resmed 9 Autoset, and then wrote an arbitrary 8-12cm pressure setting, knowing that I was on straight cpap. But he said he had to do it that way, since it was an apap.

So........if I call the doc's office and ask them to tell Lincare that I'm on a straight setting of 9, will they say I can't have an apap machine then?
If I go ahead and don't call the doc's office and then just let them set it for apap settings, can I easily change it when I get home?
Of course, the Lincare person told me I couldn't change things like that........but I can, if they don't remove that ability, right?
Which course should I take, to cover my arse? They'll be upset if they later see that I'm on straight cpap............and they never received an order from the doc.

Damn..........I get so irritated at these people treating me like I haven't got a brain.
Thanks for your input.

_________________
MachineMask
Additional Comments: Resmed HumidAir 11 for humidifier

User avatar
ChicagoGranny
Posts: 14471
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Question about Lincare and doctor's orders

Post by ChicagoGranny » Wed Feb 12, 2014 10:53 am

Be quiet and pickup the machine. Someone here will tell you how to change the pressure. It's easy.

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question about Lincare and doctor's orders

Post by Pugsy » Wed Feb 12, 2014 11:26 am

Most DMEs never even notice you changed the pressure..all they care about is that you use it so they get paid.
All they can do is verbally slap your hands if they do notice and you can tell them the doctor changed his mind or whatever.

No, they cannot "fix" it so you can't change the mode. All the machines use the same little secret code for getting to the clinical menu to make changes or just look or whatever.

http://www.apneaboard.com/resmed-s9-cpap-setup

Just smile sweetly and pickup the machine and don't let on to them that you know how to do anything...and go home and do what your want.
Or you can do like I did...I just smiled sweetly and said "it's amazing what you can find on the internet" and left it at that.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Country4ever
Posts: 1373
Joined: Wed Oct 31, 2007 6:22 pm

Re: Question about Lincare and doctor's orders

Post by Country4ever » Wed Feb 12, 2014 2:10 pm

Thanks ChicagoGranny and Pugsy!

_________________
MachineMask
Additional Comments: Resmed HumidAir 11 for humidifier

User avatar
kaiasgram
Posts: 3568
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Question about Lincare and doctor's orders

Post by kaiasgram » Wed Feb 12, 2014 2:17 pm

Had to post this one more time :
Image

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Mask: Aloha Nasal Pillow CPAP Mask with Headgear
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
glarnold
Posts: 84
Joined: Wed Sep 18, 2013 12:55 pm
Location: Bellevue, NE USA

Re: Question about Lincare and doctor's orders

Post by glarnold » Wed Feb 12, 2014 2:49 pm

Love it...

_________________
Mask: Quattro™ Air Full Face Mask with Headgear
Additional Comments: SleepyHead Software, PR System One BiPap Auto DS750 Machine Package With Bi-Flex (Discontinued)

Guest

Re: Question about Lincare and doctor's orders

Post by Guest » Wed Jul 23, 2014 10:18 pm

Really people shouldn't be changing the pressure without guidance from a respirologist, however, it's your body and if you want to take your therapy into your own hands there are plenty of online resources to help you do that. The APAP machines usually won't set at a specific pressure (S9 Escape Auto might be able to do CPAP only) so a range must be set. Also, government guidelines for funding your APAP may specify that the range must be +/- a certain pressure for APAP. A DME will get paid for the machine but they might be trying to help you out with qualifying for funding - not necessarily being greedy. My question would be the same as a previous post - why give an APAP when a CPAP would suffice? If the DME offers compliance downloads and the respirologist is worth his/her pay they should see if the pressure setting is effective and change it for the customer.

big_dave
Posts: 202
Joined: Thu May 17, 2012 11:17 am
Location: Eugene, OR

Re: Question about Lincare and doctor's orders

Post by big_dave » Wed Jul 23, 2014 11:43 pm

Guest wrote:If the DME offers compliance downloads and the respirologist is worth his/her pay they should see if the pressure setting is effective and change it for the customer.
True--but only in a perfect world. Too many of us have had to deal with arrogant, incompetent, or unresponsive doctors or DME staff, billing or insurance problems, delays of weeks or months for appointments, having equipment or supplies break while traveling, or myriad other problems.

In this imperfect world, it's best to keep your doctor involved whenever possible, but learn enough to fend for yourself when the system lets you down. That requires a machine with full efficacy data capability, such as an Autoset.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR DS-560 System One; former S8 Autoset II user; Sleepyhead (Mac); I sometimes use the Mirage Activa

User avatar
Sclark08
Posts: 475
Joined: Mon Feb 18, 2013 10:13 am

Re: Question about Lincare and doctor's orders

Post by Sclark08 » Thu Jul 24, 2014 1:44 am

We all know how to change our own pressures. Pick up the machine and tune back in here. Someone will help

Janknitz
Posts: 8413
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Question about Lincare and doctor's orders

Post by Janknitz » Thu Jul 24, 2014 2:28 am

Guest wrote:Really people shouldn't be changing the pressure without guidance from a respirologist, however, it's your body and if you want to take your therapy into your own hands there are plenty of online resources to help you do that. The APAP machines usually won't set at a specific pressure (S9 Escape Auto might be able to do CPAP only) so a range must be set. Also, government guidelines for funding your APAP may specify that the range must be +/- a certain pressure for APAP. A DME will get paid for the machine but they might be trying to help you out with qualifying for funding - not necessarily being greedy. My question would be the same as a previous post - why give an APAP when a CPAP would suffice? If the DME offers compliance downloads and the respirologist is worth his/her pay they should see if the pressure setting is effective and change it for the customer.
Any APAP can be set for a fixed pressure to be used as a CPAP, but having an APAP allows you to titrate when needed, and some people do better with an APAP even if that wasn't picked up during the titration study.

The "government" and your insurer don't care if you get an APAP or a CPAP. Same billing code, same reimbursement rate. Your DME makes more money on CPAPs, so they care because of that. DMEs are not trying to be helpful to anyone but their own bottom line by choosing CPAP over APAP for you without giving you the option.

Sleep doctors (often a pulmonogist but not always) prescribes the pressure for a patient. A DME tech cannot change the pressure setting without a doctor's order. You seem to be using the term "respirologist" to refer to a DME sleep
Tech. Many of us are educated sufficiently to adjust our pressure according to our needs, but you should not change your pressure without a good understanding of what you are doing.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
wm_hess
Posts: 334
Joined: Sat Oct 16, 2010 2:39 am

Re: Question about Lincare and doctor's orders

Post by wm_hess » Thu Jul 24, 2014 2:45 am

Guest wrote:Really people shouldn't be changing the pressure without guidance from a respirologist, however, it's your body and if you want to take your therapy into your own hands there are plenty of online resources to help you do that. The APAP machines usually won't set at a specific pressure (S9 Escape Auto might be able to do CPAP only) so a range must be set. Also, government guidelines for funding your APAP may specify that the range must be +/- a certain pressure for APAP. A DME will get paid for the machine but they might be trying to help you out with qualifying for funding - not necessarily being greedy. My question would be the same as a previous post - why give an APAP when a CPAP would suffice? If the DME offers compliance downloads and the respirologist is worth his/her pay they should see if the pressure setting is effective and change it for the customer.
If you set the minimum to say 9 and the maximum to 9 it would be in "APAP" mode and yet stay at 9 cmH2O. For me though if that was what the was I'd want to go with a pressure range of 7 to 11 or thereabout. Using the range of pressure and software I can tell where I need to be, not where the doctor is thinking I need to be. Doctors and Respiratory Therapists are using educated 'guesses' (It's called practicing medicine because they've not perfected it yet) by looking at 1 or 2 nights of data on someone who is just starting to learn to sleep with a snake in their bed, not to mention an octupus on their face. This is probably not the best nights sleep to determine the 'best' pressure

When I first started I was started at a pressure of 15, and used a S8 autoset I believe for a week and returned it to the doctor. They looked at the data and dropped it to 13. From there I got a S8 escape (brick) and used it for a year or so, using 13 cmH2O. I went ahead and purchased a S9 autoset so I could start monitoring my therapy. The doctor prescribed a pressure of 10-20 cmh2O with EPRx3. For the past year, my 95% pressure has been 11.04 So the Dr ordered 15, modified it to 13 and yet for the majority of the time my pressure is 11 or below. No straight CPAP device could have determined the optimum pressure as quickly and accurately as a APAP machine.

-Bill

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Question about Lincare and doctor's orders

Post by palerider » Thu Jul 24, 2014 8:44 am

wm_hess wrote:
Guest wrote: The APAP machines usually won't set at a specific pressure (S9 Escape Auto might be able to do CPAP only) so a range must be set.
If you set the minimum to say 9 and the maximum to 9 it would be in "APAP" mode and yet stay at 9 cmH2O.
one thing I've learned here is to pretty much ignore anything posted by 'guest'.

every apap that I've gotten my grubby little paws on has had the option of setting into straight cpap mode.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
wm_hess
Posts: 334
Joined: Sat Oct 16, 2010 2:39 am

Re: Question about Lincare and doctor's orders

Post by wm_hess » Thu Jul 24, 2014 1:30 pm

palerider wrote:
wm_hess wrote:
Guest wrote: The APAP machines usually won't set at a specific pressure (S9 Escape Auto might be able to do CPAP only) so a range must be set.
If you set the minimum to say 9 and the maximum to 9 it would be in "APAP" mode and yet stay at 9 cmH2O.
one thing I've learned here is to pretty much ignore anything posted by 'guest'.

every apap that I've gotten my grubby little paws on has had the option of setting into straight cpap mode.
While I understand the (sometimes) futility of a drive by guest, if someone else some day down the road looks at this thread the last word won't be one that is unhelpful.

-Bill

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead

User avatar
Jeff241
Posts: 49
Joined: Thu Jul 17, 2014 6:58 am
Location: MN

Re: Question about Lincare and doctor's orders

Post by Jeff241 » Thu Jul 24, 2014 1:42 pm

Guest wrote:Really people shouldn't be changing the pressure without guidance from a respirologist, however, it's your body and if you want to take your therapy into your own hands there are plenty of online resources to help you do that. The APAP machines usually won't set at a specific pressure (S9 Escape Auto might be able to do CPAP only) so a range must be set. Also, government guidelines for funding your APAP may specify that the range must be +/- a certain pressure for APAP. A DME will get paid for the machine but they might be trying to help you out with qualifying for funding - not necessarily being greedy. My question would be the same as a previous post - why give an APAP when a CPAP would suffice? If the DME offers compliance downloads and the respirologist is worth his/her pay they should see if the pressure setting is effective and change it for the customer.
Let me answer two of things you mentioned. Yes an APAP can run as a straight CPAP. I am running my S9 Autoset in CPAP mode.
The reason to get an APAP when just a CPAP will work is that my insurance will only allow me to get a new machine every 5 years. If at some point in the next 5 years the CPAP doesn't work for me then I have to try to convince my insurance that I need an APAP before the 5 year period.

_________________
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Straight CPAP @ 12 cmH2O, EPR 2, Sleepyhead Software v0.9.8-1