SD Card being erased and rewritten

Post by Citation4444 » Fri Dec 20, 2013 2:15 pm

I have been removing my SD card from my S9 Autoset each day and importing the data into ResScan and Sleepyhead. However, each time I put the card back in the S9, it erases and rewrites the card. How do I prevent it from doing this? I've tried a bunch of things, but maybe not all. My standard procedure, which doesn't work, is to turn off the machine by holding the I/O button in, followed by removing the power cord. I then write protect the card when using with the external programs, then put the switch back before putting the card back in the S9. What is the correct procedure to remove the card so that the data is not erased upon putting the card back in??

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: SD Card being erased and rewritten

Post by idamtnboy » Fri Dec 20, 2013 2:58 pm

Citation4444 wrote:I have been removing my SD card from my S9 Autoset each day and importing the data into ResScan and Sleepyhead. However, each time I put the card back in the S9, it erases and rewrites the card. How do I prevent it from doing this? I've tried a bunch of things, but maybe not all. My standard procedure, which doesn't work, is to turn off the machine by holding the I/O button in, followed by removing the power cord. I then write protect the card when using with the external programs, then put the switch back before putting the card back in the S9. What is the correct procedure to remove the card so that the data is not erased upon putting the card back in??
Take the card out of the S9 like you're doing and then put it back in without putting it in your PC. Does it read it OK or erase the data? If it's OK then your PC is somehow writing something to your card. Make absolutely certain the write protect tab is in the lock position when you put it in the PC. If the S9 erases the data without having put the card in the PC then either the card is bad or your S9 has a problem. Also, there's no need to turn the S9 completely off.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: SD Card being erased and rewritten

Post by chunkyfrog » Fri Dec 20, 2013 3:03 pm

What kind of computer is it?
I've heard that with a Mac you need to protect the card before ever reading it.
Otherwise the OS adds stuff that confuses the S9.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: SD Card being erased and rewritten

Post by Sir NoddinOff » Fri Dec 20, 2013 3:22 pm

For the last year I've always used Sleepyhead, but a few days ago I installed ResScan 4.3 for a test run. Now I can load my SD cards data just fine into SH and it shows the usual flow lines, leaks etc for the whole night... however ResScan refused to load the 'detailed data', in other words, I just get the summary data reports and when I click on the detailed tab it sez no data loaded, or something to that effect. Without the detailed lines/graphs I'm kind of in the dark regarding the details of my sleep. I'll bet the OPs question and mine are related somehow. BTW, with SH I've never had to slide the lock tab on the SD card. I'm using Win 8 and my machine is an S9 VPAP Adapt model 36007.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
Rob108
Posts: 86
Joined: Sat Jul 27, 2013 4:08 pm

Re: SD Card being erased and rewritten

Post by Rob108 » Fri Dec 20, 2013 3:31 pm

Are you sure that ResScan & Sleepyhead are pointing to the correct download location? They are different and need to be correct or import won't see data.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure 10.0cmH2O; EPR 3.0cmH2O; ResScan version 4.3.0.7081

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SD Card being erased and rewritten

Post by Citation4444 » Fri Dec 20, 2013 3:37 pm

idamtnboy wrote:
Citation4444 wrote:I have been removing my SD card from my S9 Autoset each day and importing the data into ResScan and Sleepyhead. However, each time I put the card back in the S9, it erases and rewrites the card. How do I prevent it from doing this? I've tried a bunch of things, but maybe not all. My standard procedure, which doesn't work, is to turn off the machine by holding the I/O button in, followed by removing the power cord. I then write protect the card when using with the external programs, then put the switch back before putting the card back in the S9. What is the correct procedure to remove the card so that the data is not erased upon putting the card back in??
Take the card out of the S9 like you're doing and then put it back in without putting it in your PC. Does it read it OK or erase the data? If it's OK then your PC is somehow writing something to your card. Make absolutely certain the write protect tab is in the lock position when you put it in the PC. If the S9 erases the data without having put the card in the PC then either the card is bad or your S9 has a problem. Also, there's no need to turn the S9 completely off.
If I take out the card and then put it back in, it reads the card and doesn't erase the data. I'm certain the protect tab is in the lock position. I've read elsewhere that I should copy my card into a folder and then on to another card for reading into SH and ResScan. I haven't tried that yet, but I will tomorrow. Even though the write protect in on, somehow using the card with SH and/or ResScan is affecting the card, causing an erase and rewrite when put back in the S9.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: SD Card being erased and rewritten

Post by squid13 » Fri Dec 20, 2013 3:45 pm

When the box comes up and says start the download there are 2 small rectangular boxes off to the side. Open them up and select detailed data and what ever else you need and it will download all your data.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: SD Card being erased and rewritten

Post by idamtnboy » Fri Dec 20, 2013 4:06 pm

Citation4444 wrote:If I take out the card and then put it back in, it reads the card and doesn't erase the data. I'm certain the protect tab is in the lock position. I've read elsewhere that I should copy my card into a folder and then on to another card for reading into SH and ResScan. I haven't tried that yet, but I will tomorrow. Even though the write protect in on, somehow using the card with SH and/or ResScan is affecting the card, causing an erase and rewrite when put back in the S9.
Put the card in the PC, download the data, remove the card, and put it back in the PC again. Look at the directory listing and see if there is a new file on the card. If so, the write protect feature of the card reader in your PC may not be working correctly. Simply reading the card will not cause the S9 to erase the data. Somehow something is being added to the card that is foreign to the S9 data structure and that's why it erases it.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: SD Card being erased and rewritten

Post by Jay Aitchsee » Fri Dec 20, 2013 4:13 pm

Squid may have nailed it for Sir Noddinoff. The default for ResScan is Summary only. See the two dialogue boxes below.
Image
Click "Select" on the first to get the second.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: SD Card being erased and rewritten

Post by squid13 » Fri Dec 20, 2013 4:25 pm

Thanks Jay for putting a photo of the box up, I knew where it was just couldn't quite remember what it said.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: SD Card being erased and rewritten

Post by Sir NoddinOff » Fri Dec 20, 2013 5:36 pm

Typically, I'm not easily impressed, but you guys really nailed it regarding my problem, especially Squid. I did follow Squid's somewhat abbreviated instructions and that worked perfectly. I would simply add that there is a small checkbox at the bottom left of the window that allows you to set the new xfer of detailed data from the SD card as the default... be sure to check that.
As a bonus all the 'detailed data' goes all the way back for 30 days.
Sweet.
Side Note: I'm still working on getting my year long record of SH data converted into a format that ResScan will load. JediMark... where are you when I need you? Just kidding. Have a great holiday season Mark and thanks for all the great work on your program. Reference this post:
viewtopic.php?f=1&t=94532&p=873801#p873621

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SD Card being erased and rewritten

Post by Citation4444 » Mon Dec 23, 2013 8:28 am

I solved my problem! It turned out to be my card reader. I bought one of these http://www.walmart.com/ip/Targus-USB-2. ... r/16617535 at Walmart and it solved my problem. I've now removed my card several times with no problems. By the way, I can read the card into my MacBook Pro with no problem, but with my pc using the older card reader is when I got the problem. But, with this new (and cheap) card reader all problems are solved!

Bob

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: SD Card being erased and rewritten

Post by idamtnboy » Mon Dec 23, 2013 1:40 pm

Citation4444 wrote:I solved my problem! It turned out to be my card reader.
That's good to hear! Sometimes it can be really difficult to determine if a problem is hardware or software. For instance, I've been dealing with a network connectivity problem with 4 computers at church. I think I finally have it solved. A DSL filter was not installed in the phone line to the fax machine. That's the same line the DSL modem is connected to and I think fax calls may have been upsetting the modem. I've been somewhat focusing on software issues, but it looks like it's been a hardware issue instead.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: SD Card being erased and rewritten

Post by archangle » Mon Dec 23, 2013 6:08 pm

Sounds like you found your problem, but it's been reported that Macs and Windows 8 machines will write data to the SD card whenever you insert it, and this screws up S9 cards.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SD Card being erased and rewritten

Post by Citation4444 » Mon Dec 23, 2013 7:10 pm

archangle wrote:Sounds like you found your problem, but it's been reported that Macs and Windows 8 machines will write data to the SD card whenever you insert it, and this screws up S9 cards.
I use both a MacBook Pro and a Windows 8.1 system with both ResScan and SleepyHead on both machines, and have no problems with my S9 card. I turn off the S9 before removing the card and I immediately slide the write-protect lever to protect the card. No problems with either machine now that I've got the new card reader.