Question regarding Sleepyhead Daily...
Question regarding Sleepyhead Daily...
I would like to know if the tiny, red numbers you can see when you hover over an individual daily chart event is indicative of the number of seconds that you aren't breathing?
_________________
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: Machine is the ResMed Air Curve 10 ASV. I also wear a Dr. Dakota under the chin wedge. Software is Sleepyhead and Oscar. |
- ChicagoGranny
- Posts: 14570
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Question regarding Sleepyhead Daily...
Look on the left side of the screen, and click on the Events tab. You will see lists of the events including the length of each event. If you click on an individual event it will zoom in on the event in the flow chart. I know that doesn't answer your question, but it is a neat tool for analysis.
Re: Question regarding Sleepyhead Daily...
You are right...a very neat tool for analysis. Thank you!ChicagoGranny wrote: ↑Thu Apr 11, 2019 11:16 amLook on the left side of the screen, and click on the Events tab. You will see lists of the events including the length of each event. If you click on an individual event it will zoom in on the event in the flow chart. I know that doesn't answer your question, but it is a neat tool for analysis.
_________________
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: Machine is the ResMed Air Curve 10 ASV. I also wear a Dr. Dakota under the chin wedge. Software is Sleepyhead and Oscar. |
- Dog Slobber
- Posts: 3982
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Question regarding Sleepyhead Daily...
For reasons I don't understand Hypopneas (H) are always reported as 10 seconds.
Not sure, but I believe Unclassified Apneas (UA) ar also always reports as 10 seconds. I don't have enough of them to verify.
Not sure, but I believe Unclassified Apneas (UA) ar also always reports as 10 seconds. I don't have enough of them to verify.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Re: Question regarding Sleepyhead Daily...
Thanks, Dogslobber. That is a mystery to me too. Maybe someone will come along and enlighten us both.Dog Slobber wrote: ↑Thu Apr 11, 2019 12:18 pmFor reasons I don't understand Hypopneas (H) are always reported as 10 seconds.
Not sure, but I believe Unclassified Apneas (UA) ar also always reports as 10 seconds. I don't have enough of them to verify.
_________________
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: Machine is the ResMed Air Curve 10 ASV. I also wear a Dr. Dakota under the chin wedge. Software is Sleepyhead and Oscar. |
Re: Question regarding Sleepyhead Daily...
In the case of ResMed machines like you are using, for (Edit: for CA & OA events, DogSlobbers right on H's being weird, who knows what ResMed means by durations on those) they are durations in seconds.
Just a reference for those with Philips Respironics machines, unfortunately we don't yet exactly know how the event data fields correlate to event durations, so you need to turn on Custom CPAP User Event Flagging and the Resync machine detected events to try and detect the real event durations. (This only works when Flow Waveform is present, and only for Hypopneas, Clear Airways and Obstructive events, and doesn't catch them all... I never had enough time left over to work on my waveform processing code.)
Just a reference for those with Philips Respironics machines, unfortunately we don't yet exactly know how the event data fields correlate to event durations, so you need to turn on Custom CPAP User Event Flagging and the Resync machine detected events to try and detect the real event durations. (This only works when Flow Waveform is present, and only for Hypopneas, Clear Airways and Obstructive events, and doesn't catch them all... I never had enough time left over to work on my waveform processing code.)
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: CMS50D+/F Oximeter, S9 VPAP Auto/Adapt, PRS1 Auto, Intellipap Auto, SleepyHead :) |
Author of the free, cross platform, open-source sleep tracking software SleepyHead.
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code
Download http://sleepyhead.jedimark.net
Source Code http://gitlab.com/sleepyhead/sleepyhead-code
Re: Question regarding Sleepyhead Daily...
Thanks, JediMark, Great to hear from the author of this wonderful software. I am very impressed with it and have learned so much more about apnea in general by utilizing Sleepyhead. Thank you for helping so many people!jedimark wrote: ↑Thu Apr 11, 2019 2:53 pmIn the case of ResMed machines like you are using, for (Edit: for CA & OA events, DogSlobbers right on H's being weird, who knows what ResMed means by durations on those) they are durations in seconds.
Just a reference for those with Philips Respironics machines, unfortunately we don't yet exactly know how the event data fields correlate to event durations, so you need to turn on Custom CPAP User Event Flagging and the Resync machine detected events to try and detect the real event durations. (This only works when Flow Waveform is present, and only for Hypopneas, Clear Airways and Obstructive events, and doesn't catch them all... I never had enough time left over to work on my waveform processing code.)
_________________
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: Machine is the ResMed Air Curve 10 ASV. I also wear a Dr. Dakota under the chin wedge. Software is Sleepyhead and Oscar. |
- ChicagoGranny
- Posts: 14570
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Question regarding Sleepyhead Daily...
Are you using a ResMed machine?Dog Slobber wrote: ↑Thu Apr 11, 2019 12:18 pmFor reasons I don't understand Hypopneas (H) are always reported as 10 seconds.
This is using ResMed machine -
Re: Question regarding Sleepyhead Daily...
From near as I can tell ResMed changed hyponea duration reporting with the AirSense 10 models....all hyponeas are 10 seconds no matter how long they are.ChicagoGranny wrote: ↑Sat Apr 13, 2019 5:56 pmAre you using a ResMed machine?Dog Slobber wrote: ↑Thu Apr 11, 2019 12:18 pmFor reasons I don't understand Hypopneas (H) are always reported as 10 seconds.
This is using ResMed machine -
When I was using the S9 model I got actual duration in seconds like you do...but now that I am using the AirSense model every single hyponea I ever have flagged is exactly 10 seconds...and ResScan shows no duration at all. I think that it's just how they get reported now.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- Dog Slobber
- Posts: 3982
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Question regarding Sleepyhead Daily...
Yup ResMed AS 10 AutoSet for HerChicagoGranny wrote: ↑Sat Apr 13, 2019 5:56 pmAre you using a ResMed machine?Dog Slobber wrote: ↑Thu Apr 11, 2019 12:18 pmFor reasons I don't understand Hypopneas (H) are always reported as 10 seconds.
This is using ResMed machine -
hypopnea.JPG
Every Hypopnea presented is 10 seconds. In both SH 1.0.0 and SH 1.1.0
Glad to know it's a AirSense 10 v S9 thing.From near as I can tell ResMed changed hyponea duration reporting with the AirSense 10 models....all hyponeas are 10 seconds no matter how long they are.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
- ChicagoGranny
- Posts: 14570
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Question regarding Sleepyhead Daily...
While we are on SH, the Overview graphs, which once worked well, quit working for me. I uninstalled SH and then installed it again using the link in Pugsy's tutorial (v1.0.0-beta-2). The Overview graphs still don't work.
Re: Question regarding Sleepyhead Daily...
jimbud writes:
Same thing happened to me about a month ago. After a few days they came back and are there for now. Buggy? JPBChicagoGranny wrote: ↑Sun Apr 14, 2019 11:42 amWhile we are on SH, the Overview graphs, which once worked well, quit working for me. I uninstalled SH and then installed it again using the link in Pugsy's tutorial (v1.0.0-beta-2). The Overview graphs still don't work.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleep on a Buckwheat Hull Pillow. |
- ChicagoGranny
- Posts: 14570
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Question regarding Sleepyhead Daily...
jimbud writes:
Thank you. I think?
JPB