ResScan not recognizing my SD card

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

ResScan not recognizing my SD card

Post by SleepingUgly » Sat Oct 13, 2012 7:07 pm

First, does ResScan 3.16 work with S9 VPAP Adapt? I couldn't get it to recognize an SD card from an Adapt machine. I put ResScan 4.1 on and it doesn't see the SD card. We put it on my husband's computer and once it read the data, and since then it doesn't recognize there's a card there. I've tried two different SD cards. Anyone know what's going on?
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
NateS
Posts: 1716
Joined: Wed Dec 14, 2011 8:53 pm
Location: Kaatskill Mts-Washington Irving

Re: ResScan not recognizing my SD card

Post by NateS » Sat Oct 13, 2012 8:00 pm

SleepingUgly wrote:First, does ResScan 3.16 work with S9 VPAP Adapt? I couldn't get it to recognize an SD card from an Adapt machine. I put ResScan 4.1 on and it doesn't see the SD card. We put it on my husband's computer and once it read the data, and since then it doesn't recognize there's a card there. I've tried two different SD cards. Anyone know what's going on?
What version of Windows are you using, and are you using Windows on a PC or Windows as a virtual machine on a Mac; if the latter, what version of OSX are you using and what brand and version of VM software are you using?

Knowing this may help considerably in narrowing down the problem and in getting responses from members with a setup similar to yours.

Nate

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: ResMed AirCurve 10 ASV; Dreamwear Nasal Mask Original; CPAPMax Pillow; ResScan & SleepyHead
Central sleep apnea AHI 62.6 pre-VPAP. Now 0 to 1.3
Present Rx: EPAP: 8; IPAPlo:11; IPAPHi: 23; PSMin: 3; PSMax: 15
"I've had a perfectly wonderful evening, but this wasn't it." —Groucho Marx

User avatar
Pugsy
Posts: 64014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan not recognizing my SD card

Post by Pugsy » Sat Oct 13, 2012 8:01 pm

I assume that you have confirmed that the computer can see the SD card and the data files within the SD card?
So you have eliminated the slot or drive where the SD card sits as being the problem.
Just saw that SleepyHead won't read it either.
While I know that some Respironics ASV machines won't work with SleepyHead...I haven't heard of the ResMed VPAP high end machines having a similar problem.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: ResScan not recognizing my SD card

Post by SleepingUgly » Sat Oct 13, 2012 8:05 pm

NateS wrote:What version of Windows are you using, and are you using Windows on a PC or Windows as a virtual machine on a Mac
Windows 7, Professional version, on a PC, not a virtual machine
Pugsy wrote:I assume that you have confirmed that the computer can see the SD card and the data files within the SD card?
Yes, we can see the SD card and the data files in it. I can email the data and other people can unzip and download and open in their ResScan.
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
Pugsy
Posts: 64014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan not recognizing my SD card

Post by Pugsy » Sat Oct 13, 2012 8:15 pm

How many uninstalls and installs have you tried with ResScan? Same download source?
Are talking about only version 4.1....or are you having problems with all versions?

When you tried SleepyHead...did you try creating a new user profile? I assume you probably did because you know your computer stuff but it helps us to know exactly what you have tried and come up empty with.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Slartybartfast
Posts: 1633
Joined: Wed Sep 01, 2010 12:34 pm

Re: ResScan not recognizing my SD card

Post by Slartybartfast » Sat Oct 13, 2012 8:16 pm

FWIW, I just tried, without success, to get ResScan 3.16 to work with my Win 7 Home Edition 32 bit system today. I tried several times running it in compatibility mode as Win XP SP3 and the window kept saying the software was still incompatible. So I downloaded and successfully ran ResScan 4.1.

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: ResScan not recognizing my SD card

Post by SleepingUgly » Sat Oct 13, 2012 8:49 pm

How many uninstalls and installs have you tried with ResScan? Same download source?
I downloaded the latest version several times. No difference.
Are talking about only version 4.1....or are you having problems with all versions?
Before I installed version 4.1, I attempted to read the ASV data from the previous version that I was using which was 3.16. That couldnt even see the ASV SD card in my machine. Mind you, 3.16 was working flawlessly in importing data from my S9 bilevel machine. The problem happened when I moved to the ASV machine.
When you tried SleepyHead...did you try creating a new user profile?
Yes, I created a totally new profile. Sleepyhead doesn't even see the ASV data on the card. It's just blank. I even copied the data from the SD card to a separate directory on my computer and tried to import from that. Same problem -- the directory appeared blank. Then I put in a SD card with my old S9 data on it (before I got the ASV machine) and it worked perfectly. So there is some issue, I suspect, that the Sleepyhead software has with ASV data (at least from my machine).
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
Pugsy
Posts: 64014
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResScan not recognizing my SD card

Post by Pugsy » Sat Oct 13, 2012 8:57 pm

SleepingUgly wrote:So there is some issue, I suspect, that the Sleepyhead software has with ASV data (at least from my machine).
Very possible.
I know that some people are able to use SleepyHead perfectly with their PR S1 ASV machine and someone else using the very same model ASV type of machine can't get SH to work for them at all.
You are the first that I have heard of this happening with the ResMed line though. We may see more of it later.
Don't you just love it that here is another thing that you are "special" in? Like the Gods are looking down and saying "Let's mess with SU just a little more"..

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: ResScan not recognizing my SD card

Post by SleepingUgly » Sun Oct 14, 2012 8:59 am

Pugsy wrote:Don't you just love it that here is another thing that you are "special" in? Like the Gods are looking down and saying "Let's mess with SU just a little more"..
No kidding!

BTW, in the interest of full disclosure, I'm not tech savvy, so my husband is dealing with my computer problems and will occasionally check posts here and respond.
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: ResScan not recognizing my SD card

Post by SleepingUgly » Mon Oct 15, 2012 11:23 am

Bumping this up in case anyone else has ideas.
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: ResScan not recognizing my SD card

Post by JohnBFisher » Mon Oct 15, 2012 2:00 pm

I just posted another note about this.

You might need a card reader that can handle an SDHC card. That's what they use in the S9. The SDHC card in the s9 is 4GB in size. Here's an example card reader at NewEgg.com

http://www.newegg.com/Product/Product.a ... 12K0958344

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: ResScan not recognizing my SD card

Post by SleepingUgly » Mon Oct 15, 2012 2:25 pm

I'm using the same cards I always was. I don't think that's it, but thanks.
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

cpap adict

Re: ResScan not recognizing my SD card

Post by cpap adict » Mon Jan 26, 2015 2:04 pm

Hi,

A late answer, but I recently fell in the same trap using ResScan and Windows Vista.
Simply run ResScan as administrator (left click in the start menu).

Regards.