Minimum pressure on ResMed S9 Autoset

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
wellwood
Posts: 2
Joined: Thu Sep 20, 2012 3:39 am

Minimum pressure on ResMed S9 Autoset

Post by wellwood » Thu Sep 20, 2012 4:10 am

I have my ResMed S9 Autoset set between 4cm and 7cm. My next day's pressure graphs show that it stays at 4cm until an event occurs, when it rises to 7cm (and hopefully does its job). It will then drop to around the 6.2 - 6.6cm pressure (but never lower thereon).
I need to turn the power off, then on again to get the pressure back to the 4cm mark, which is more comfortable.
Surely the CPAP machine should default back to the minimum setting (4cm) when "all is clear". Any thoughts would be appreciated.
PS..I am a newbie from Australia and am overjoyed to have your forum to learn the many and varied aspects of CPAP therapy.

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: Minimum pressure on ResMed S9 Autoset

Post by apneawho » Thu Sep 20, 2012 5:52 am

After an event, it will maintain a pressure that represses another event. It has internal algorithms and sensors that determine this. It make take a pressure of around 6 to keep your otherwise collapsed airway open. My magic numerber is 14.2. Each person's anatomy is different.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
Pugsy
Posts: 64012
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Minimum pressure on ResMed S9 Autoset

Post by Pugsy » Thu Sep 20, 2012 6:05 am

wellwood wrote:Surely the CPAP machine should default back to the minimum setting (4cm) when "all is clear".
APAPs don't increase in the blink of an eye and they don't decrease in a blink of an eye either. If you wake up and pressure of 6 is showing and you don't like it then you don't really have a choice except a quick restart to drop the pressure down to starting point again. Some machines have a ramp button separate from the on/off button but the ResMed doesn't.

I don't know how long you have been on the machine but eventually we do get used to the pressure and eventually that 6 cm pressure won't seem so awful for you. If you are seeing it stay at 6 all night...then there is a reason the machine is staying at 6 or whatever. The APAPs respond to event precursors...warning signs that a collapse of the airway is imminent.
Things like snores or flow limitations. If you wake up and stay awake long enough the machine will drop down to 4 cm again but it does it in gradual steps down. If you are asleep and see it stay at 6..then the machine is doing its job because it senses the need for that pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
KEQ5
Posts: 93
Joined: Sat Jul 21, 2012 7:38 pm

Re: Minimum pressure on ResMed S9 Autoset

Post by KEQ5 » Thu Sep 20, 2012 8:33 pm

4-7 is a very low (and rather narrow) range.

You may just need to get used to the pressure and/or play around with the EPR setting (try it at 1 or 2cm) to make it more comfortable.

If the machine had to go up to 6.5-7.0 in order to stop your events and the limit is 7.0, then you're going to need to look at your charts and see whether it's always bumping up against the 7.0 limit and whether you're still seeing a lot of OAs at the 7.0 pressure.

wellwood
Posts: 2
Joined: Thu Sep 20, 2012 3:39 am

Re: Minimum pressure on ResMed S9 Autoset

Post by wellwood » Thu Sep 20, 2012 11:20 pm

Many thanks to apneawho, Pugsy and KEQ5 for your informative replies.
After studying my graphs, I see that my maximum pressure is indeed bumping along the 7cm mark for a lot of the night (previously set at 4.0 min-7.0 max) and EPR set to OFF.
For tonight's journey into sleepsville, I have set the pressure at 5.0 min and 9.0 max, and turned the EPR to position 1.
My main problem with the increased pressure is that I get a bubble of air in my mouth every now and again (causing a 'blowing bubble effect'), even when I am just lying in bed watching TV, with my mouth firmly closed and breathing regularly.
Will let you know how I perform tonight, thanks again...

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: Minimum pressure on ResMed S9 Autoset

Post by robysue » Fri Sep 21, 2012 7:24 am

wellwood wrote: My main problem with the increased pressure is that I get a bubble of air in my mouth every now and again (causing a 'blowing bubble effect'), even when I am just lying in bed watching TV, with my mouth firmly closed and breathing regularly.
Will let you know how I perform tonight, thanks again...
If the air in your mouth problem starts turning into an air in your stomach problem, you may want to consider backing the pressure back down to your original range of 4-7 if your AHI < 5 at that range and you were not seeing significant clustering of events occurring when the pressure was at the max.

While it goes against conventional wisdom here at cpaptalk, there's nothing automatically "wrong" with allowing an APAP to hit its max pressure setting (frequently) if your max setting is enough to control your apnea---as in your overall AHI < 5 and you are NOT seeing significant clusters of events when you are at maximum pressure. In other words, for a few people, very minor snoring and flow limitations can cause an auto machine to increase pressure to the point where it triggers aerophagia and/or spontaneous arousals caused by discomfort. And if you're one of these people (I am), increasing the max pressure beyond what you strictly need to control the apneas and hypopneas can lead to diminishing returns: The higher max pressure doesn't necessarily bring the AHI down any further, but it can increase restlessness and discomfort during the night, which then leads to less overall sleep and less high quality sleep, which leads to feeling lousy in the daytime.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

luiggi

Re: Minimum pressure on ResMed S9 Autoset

Post by luiggi » Sun Jul 07, 2013 12:30 pm

I just got my Resmed S9. When I turn it on a blast of air (like a leaf blower) comes out of the mask. The air pressure shows 4.0 . If I put the mask on I feel that my head is going to blow out. Is this normal?? When I was at the clinic the air started to come out very slow . Is there anything wrong with the equipment or ???

Wulfman...

Re: Minimum pressure on ResMed S9 Autoset

Post by Wulfman... » Sun Jul 07, 2013 12:37 pm

luiggi wrote:I just got my Resmed S9. When I turn it on a blast of air (like a leaf blower) comes out of the mask. The air pressure shows 4.0 . If I put the mask on I feel that my head is going to blow out. Is this normal?? When I was at the clinic the air started to come out very slow . Is there anything wrong with the equipment or ???
Strap the mask on your face BEFORE you start the machine. Otherwise, it's going to do what you describe.


Den

.

Luiggi

Re: Minimum pressure on ResMed S9 Autoset

Post by Luiggi » Sun Jul 07, 2013 2:04 pm

I tried to start the resmed 9 with the mask on and the air continues to blow very hard without going down. I feel that my nose and my lungs are going to explote!!!. It only shows 6 cm.......What next ??? Thank you.....

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Minimum pressure on ResMed S9 Autoset

Post by DreamDiver » Sun Jul 07, 2013 5:55 pm

wellwood wrote:... My main problem with the increased pressure is that I get a bubble of air in my mouth every now and again (causing a 'blowing bubble effect'), even when I am just lying in bed watching TV, with my mouth firmly closed and breathing regularly. ...
If you're getting "chipmunk cheeks", and air is leaking out of your mouth, then you may be a mouth leaker, in which case you may want to try a full-face mask. (Some people like to try mouth taping - you can search this forum for taping.) Whether you elect to try the full face mask or try taping, a PapCap is a great addition to stop chipmunk cheeks. If you are getting air into your belly (aerophagia) causing painful belly aches and gassiness, then you may want to do as robysue suggests and reduce your settings back down so that you no longer experience aerophagia.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Minimum pressure on ResMed S9 Autoset

Post by DreamDiver » Sun Jul 07, 2013 5:57 pm

Luiggi wrote:I tried to start the resmed 9 with the mask on and the air continues to blow very hard without going down. I feel that my nose and my lungs are going to explote!!!. It only shows 6 cm.......What next ??? Thank you.....
Hi Luiggi - Welcome. Why not become a member, fill out your equipment profile and allow us to learn more about what you're going through on a new thread for you?

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF

User avatar
billbolton
Posts: 2266
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: Minimum pressure on ResMed S9 Autoset

Post by billbolton » Sun Jul 07, 2013 6:21 pm

wellwood wrote:even when I am just lying in bed watching TV, with my mouth firmly closed and breathing regularly.
The scoring algorithms on any xPAP system only function effectively when you are asleep and breathing autonomously

Anytime you are awake, you will be breathing differently than when you are asleep, so whatever you experience while using xPAP while awake is no guide to what will occur will you when you are asleep.

Cheers,

Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6
Last edited by billbolton on Sun Jul 07, 2013 6:23 pm, edited 1 time in total.

Luiggi

Re: Minimum pressure on ResMed S9 Autoset

Post by Luiggi » Sun Jul 07, 2013 6:22 pm

Hi dreamdiver: I'm a member already.

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Minimum pressure on ResMed S9 Autoset

Post by DreamDiver » Sun Jul 07, 2013 6:46 pm

Luiggi wrote:Hi dreamdiver: I'm a member already.
Unfortunately, you are not logged in right now, and you have not yet posted while logged in with the name "Luiggi". Your name will then be bold, high-lighted and linked to your profile so we can learn more about your machine, mask and other stuff. (See image below comparing your avatar and mine.)
Image

If you see "Login" instead of "Logout [ Luiggi ]" you can be pretty sure you're not logged in on the computer you're using.
How about logging in?
Image

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF