S9 Autoset not delivering correct pressure

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: S9 Autoset not delivering correct pressure

Post by palerider » Tue Jan 27, 2015 9:29 pm

Burkebang wrote:
palerider wrote:
the machine is set to a fairly tight range, but, you might notice the pressure summary line:

minimum: 8.22 median: 13.62 95%: 14.42 max: 15.18
The pressure summary line is not interesting, the detailed pressure graph is. I have been doing this for a while and I do know something about it. I stand by what I say. When a person has 13,6 min pressure, a flat pressuregraph, no snoring and almost no events, the min pressure is unnessecarily high..
so, if I get what you're saying....

someone whose pressure doesn't max out and jump all over the place... has too much pressure.
someone whose machine is preventing almost all events, ... has too much pressure.
someone who's machine is keeping the snoring pretty much non-existent... has too much pressure.

I'd say that, going by what we're generally trying to achieve here, that's actually

someone who's machine is adjusted just about right

now, if the OP wants to listen to your crackpot theories, and wants to have MORE events, and MORE disturbed sleep, then, sure, lower the pressure till he's gasping and choking and having a higher ahi.

now, if the person didn't have ANY events, and their pressure NEVER went above minimum, (which, though you don't care about it, this person's pressure does) then sure, I'd say try lowering it a bit and see, IF the current pressure is bothering them.

but, I'd say, again, you're full of fertilizer.... however, people are free to follow your tragically stupid advice if they want, I wish them well.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Burkebang
Posts: 297
Joined: Fri Dec 09, 2011 4:30 pm
Location: Oslo, Norway

Re: S9 Autoset not delivering correct pressure

Post by Burkebang » Tue Jan 27, 2015 10:23 pm

palerider wrote:
Burkebang wrote:
palerider wrote:
the machine is set to a fairly tight range, but, you might notice the pressure summary line:

minimum: 8.22 median: 13.62 95%: 14.42 max: 15.18
The pressure summary line is not interesting, the detailed pressure graph is. I have been doing this for a while and I do know something about it. I stand by what I say. When a person has 13,6 min pressure, a flat pressuregraph, no snoring and almost no events, the min pressure is unnessecarily high..
so, if I get what you're saying....

someone whose pressure doesn't max out and jump all over the place... has too much pressure.
someone whose machine is preventing almost all events, ... has too much pressure.
someone who's machine is keeping the snoring pretty much non-existent... has too much pressure.

I'd say that, going by what we're generally trying to achieve here, that's actually

someone who's machine is adjusted just about right

now, if the OP wants to listen to your crackpot theories, and wants to have MORE events, and MORE disturbed sleep, then, sure, lower the pressure till he's gasping and choking and having a higher ahi.

now, if the person didn't have ANY events, and their pressure NEVER went above minimum, (which, though you don't care about it, this person's pressure does) then sure, I'd say try lowering it a bit and see, IF the current pressure is bothering them.

but, I'd say, again, you're full of fertilizer.... however, people are free to follow your tragically stupid advice if they want, I wish them well.
The goal of the therapy is to reduce the AHI to normal, that means below 5 using as little intervention - pressure as possible. The goal is to sleep well and most people, specially newbees sleeps much better with low pressures. This increases compliancy and makes people stick to their therapy. People like you cause MANY people to unnessecarily up their pressures and drop their therapy and that costs lifes.

I guess if you take sleep mediction, you take the whole bottle also?

Why buy an expensive APAP, if you are not going to trust the algorithm and USE it as a APAP? Check my signature for the sample Sleepyhead report. It shows a pressure graph that shows my APAP working as an APAP SHOULD be working. What would be your response to my pressure grapf? Up the min pressure to 13,6?

People like you is what drove me away from this forum, I see that things have not improved around here...

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Mirage FX nasal mask a lot. Contec CMS-50D+ Pulseoximeter and Zeo Mobile tracks the quality of my therapy.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: S9 Autoset not delivering correct pressure

Post by palerider » Tue Jan 27, 2015 10:54 pm

Burkebang wrote: The goal of the therapy is to reduce the AHI to normal, that means below 5 using as little intervention - pressure as possible.
most people with any experience here would disagree with your idea that 5 is a good ahi to have... you do know that means, that, on average, you're getting roused every 12 minutes? but... you don't care about others, that's pretty clear.
Burkebang wrote: the goal is to sleep well and most people, specially newbees sleeps much better with low pressures.
and the OP, who's therapy you want to screw up, is far from a newbie, so your blathering is irrelevant.
Burkebang wrote:I guess if you take sleep mediction, you take the whole bottle also?
there's a term professionals have for this kind of bullshit "question", but I prefer to just call it 'bullshit'.
Burkebang wrote: Why buy an expensive APAP, if you are not going to trust the algorithm and USE it as a APAP? Check my signature for the sample Sleepyhead report. It shows a pressure graph that shows my APAP working as an APAP SHOULD be working. What would be your response to my pressure grapf? Up the min pressure to 13,6?
well, I'd turn the minimum up above 4, which is what you've got it set at, since you're having more events, and flow limitations than you need to, every time it tries to get back down to 4. if you actually KNEW anything, you'd know that the prevailing wisdom is that 4cm min pressure is stupid.

and, no, dipshit, I wouldn't suggest setting it at 13.6, since your pressure didn't get that high.

btw, since you're incapable of actually READING things, like numbers, and comprehending them, I'll 'splain something to you. I'll try to use small words, and type slow so you can follow along.

the OP's minimum base pressure is 11.6, except when he inhales, at which point it goes up to 13.6, this is because he's using EPR, which your machine doesn't have... which a limited bi-level feature. his machine, using that auto feature you soooo want to see doing more, took his ipap up to 15.18, so at some point, he DID need more pressure, and it gave it to him. you also might notice, if your head wasn't firmly buried in that stinky place, that the screenshot is zoomed in so you're not even seeing the whole pressure trace... you're just making asinine ASSumptions.

but, I don't expect you to be able to understand these complicated things.
Burkebang wrote: People like you is what drove me away from this forum, I see that things have not improved around here...
obviously, they were clear thinking, intelligent people. perhaps they'll get out their pitchforks and drive you and your bad, ignorant ideas back out again.

cheers!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sun Apr 26, 2015 11:21 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Burkebang
Posts: 297
Joined: Fri Dec 09, 2011 4:30 pm
Location: Oslo, Norway

Re: S9 Autoset not delivering correct pressure

Post by Burkebang » Wed Jan 28, 2015 2:23 am

Oh man... I will not even quote all that foul language...

I guess people will just have to make up their own mind about which one of us is intelligent and which one is full of... Because one of us is definatly a complete moron

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Mirage FX nasal mask a lot. Contec CMS-50D+ Pulseoximeter and Zeo Mobile tracks the quality of my therapy.

User avatar
ChicagoGranny
Posts: 14569
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: S9 Autoset not delivering correct pressure

Post by ChicagoGranny » Wed Jan 28, 2015 7:16 am

Burkebang wrote:The goal of the therapy is to reduce the AHI to normal, that means below 5
Not acceptable for me!

If my AHI hits 3.5 or higher, my energy level the next day is poor.

AHI of 1.0 or less is my target and that gives this old man unbelievable (My friends tell me) energy for any age.

But I do believe the OP should experiment with lowering his minimum pressure a bit.

ChicagoGramps

rcs914
Posts: 32
Joined: Tue Sep 04, 2012 12:04 pm

Re: S9 Autoset not delivering correct pressure

Post by rcs914 » Wed Jan 28, 2015 10:33 am

It is kind of funny to me considering I came here because I was pretty much convinced that my machine is not delivering enough pressure to be told to lower it. In the past I did have the lower end of the pressure at 12.0 - I had raised it before because of the feeling like I wasn't getting enough air.

Just as a test, if I purposefully stop breathing with the mask on, shouldn't I be able to feel it try to increase pressure in order to force me to breathe?

I appreciate the suggestions, and it is re-assuring that my numbers look good, but I still can't shake the feeling that there is something not quite right. Manometer test for me, although I know that the machine is capable of blowing more air (see my comments about the mask fit test) so I am not certain if this will work or not. Can't hurt anything by trying.

Chris

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: S9 Autoset not delivering correct pressure

Post by OkyDoky » Wed Jan 28, 2015 10:37 am

Your machine will not force you to breath. It supports your breathing with pressure when you breath.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
LSAT
Posts: 13238
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: S9 Autoset not delivering correct pressure

Post by LSAT » Wed Jan 28, 2015 10:51 am

rcs914 wrote:It is kind of funny to me considering I came here because I was pretty much convinced that my machine is not delivering enough pressure to be told to lower it. In the past I did have the lower end of the pressure at 12.0 - I had raised it before because of the feeling like I wasn't getting enough air.

Just as a test, if I purposefully stop breathing with the mask on, shouldn't I be able to feel it try to increase pressure in order to force me to breathe?

I appreciate the suggestions, and it is re-assuring that my numbers look good, but I still can't shake the feeling that there is something not quite right. Manometer test for me, although I know that the machine is capable of blowing more air (see my comments about the mask fit test) so I am not certain if this will work or not. Can't hurt anything by trying.

Chris
This is the 3rd day you have been concerned about this. Why haven't you just jumped into your can and driven to ANY DME that provides CPAP equipment and asked them to check your machine. That way you will know whether it is you or your machine. (My $$ is on you)

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: S9 Autoset not delivering correct pressure

Post by Jay Aitchsee » Wed Jan 28, 2015 11:04 am

rcs914 wrote:Just as a test, if I purposefully stop breathing with the mask on, shouldn't I be able to feel it try to increase pressure in order to force me to breathe?
No, As OkyDoky said, unless you have a ventilating machine, it does not force you to breathe. In fact, many users become so accustomed to the pressure, they wonder if the machine is on, especially after waking in the night.
I'm with LSAT, just do it. If you do it at home, don't forget the video .
My money is on Chris, along with LSAT. Any takers?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: S9 Autoset not delivering correct pressure

Post by palerider » Wed Jan 28, 2015 11:22 am

rcs914 wrote: Just as a test, if I purposefully stop breathing with the mask on, shouldn't I be able to feel it try to increase pressure in order to force me to breathe?
as others have noted, your machine will *not* increase the pressure during an apnea, your machine does not attempt to breath for you. others will, such as the VPAP ST, VPAP Adapt, and VPAP ST-A.

what will happen is that if you hold your breath for about 6 seconds, you might feel the very faint rapid fluctuation of the pressure as the machine uses forced pressure oscillations (FOT) to attempt to see if your airway is open, then it will flag an apnea.

additionally, since you're using EPR, which is lowering your pressure by 2cm when you're not inhaling, after holding your breath for 15 seconds, EPR will timeout and raise your pressure by 2cm, and wait for your next breath to begin before it starts working again. you *might* feel that slight pressure change. similar to the description of EPR on the s8 series below.
EPR Timeout
If a patient's exhalation period exceeds 15 seconds, EPR immediately suspends. The treatment pressure reverts to set CPAP and remains suspended until the next inhalation phase is detected. EPR helps patients experience the highest level of comfort and benefit from their therapy. EPR is available on our S8 Elite™ and S8 AutoSet Vantage™ flow generators.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

rcs914
Posts: 32
Joined: Tue Sep 04, 2012 12:04 pm

Re: S9 Autoset not delivering correct pressure

Post by rcs914 » Wed Jan 28, 2015 2:25 pm

palerider wrote:
rcs914 wrote: Just as a test, if I purposefully stop breathing with the mask on, shouldn't I be able to feel it try to increase pressure in order to force me to breathe?
as others have noted, your machine will *not* increase the pressure during an apnea, your machine does not attempt to breath for you. others will, such as the VPAP ST, VPAP Adapt, and VPAP ST-A.

what will happen is that if you hold your breath for about 6 seconds, you might feel the very faint rapid fluctuation of the pressure as the machine uses forced pressure oscillations (FOT) to attempt to see if your airway is open, then it will flag an apnea.

additionally, since you're using EPR, which is lowering your pressure by 2cm when you're not inhaling, after holding your breath for 15 seconds, EPR will timeout and raise your pressure by 2cm, and wait for your next breath to begin before it starts working again. you *might* feel that slight pressure change. similar to the description of EPR on the s8 series below.
EPR Timeout
If a patient's exhalation period exceeds 15 seconds, EPR immediately suspends. The treatment pressure reverts to set CPAP and remains suspended until the next inhalation phase is detected. EPR helps patients experience the highest level of comfort and benefit from their therapy. EPR is available on our S8 Elite™ and S8 AutoSet Vantage™ flow generators.
Good to know, thank you. And to others - I have a one year old, and PLENTY of other stuff on my plate on a daily basis. I'm guessing that there likely is a DME around where I live, but I haven't looked it up yet, or called them to ask them if they would do a pressure test. I'm not going to just walk in

I appreciate the assistance.

User avatar
Tatooed Lady
Posts: 984
Joined: Sat Jun 07, 2014 6:18 pm
Location: Central Wisconsin

Re: S9 Autoset not delivering correct pressure

Post by Tatooed Lady » Wed Jan 28, 2015 7:58 pm

Well golly, hasn't this thread been a hotbed of activity and taunts? And I had NOTHING to do with it!! FWIW, I have found PRs grasp on the subject of technology as a tool to be pretty spot on...so I clear him from being a moron, idiot or other in this case. In another setting, maybe...but gizmos and gadgets are his life.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Precious and POW are very very good to me.
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: S9 Autoset not delivering correct pressure

Post by palerider » Wed Jan 28, 2015 8:03 pm

Tatooed Lady wrote:Well golly, hasn't this thread been a hotbed of activity and taunts? And I had NOTHING to do with it!! FWIW, I have found PRs grasp on the subject of technology as a tool to be pretty spot on...so I clear him from being a moron, idiot or other in this case. In another setting, maybe...but gizmos and gadgets are his life.
I'll paypal you the 20 bucks for the kind words right now *giggle*

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Tatooed Lady
Posts: 984
Joined: Sat Jun 07, 2014 6:18 pm
Location: Central Wisconsin

Re: S9 Autoset not delivering correct pressure

Post by Tatooed Lady » Wed Jan 28, 2015 8:11 pm

palerider wrote:
Tatooed Lady wrote:Well golly, hasn't this thread been a hotbed of activity and taunts? And I had NOTHING to do with it!! FWIW, I have found PRs grasp on the subject of technology as a tool to be pretty spot on...so I clear him from being a moron, idiot or other in this case. In another setting, maybe...but gizmos and gadgets are his life.
I'll paypal you the 20 bucks for the kind words right now *giggle*

Lol...redirect it to ghost...and make it $40 and I'll make up....er....throw in....other nice stuff.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Precious and POW are very very good to me.
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!

User avatar
saltydawg2
Posts: 48
Joined: Thu Sep 18, 2014 5:03 pm

Re: S9 Autoset not delivering correct pressure

Post by saltydawg2 » Wed Jan 28, 2015 8:34 pm


Well golly, hasn't this thread been a hotbed of activity and taunts? And I had NOTHING to do with it!! FWIW, I have found PRs grasp on the subject of technology as a tool to be pretty spot on...so I clear him from being a moron, idiot or other in this case. In another setting, maybe...but gizmos and gadgets are his life.
Have to agree! I learned more about right, and stupid in this exchange than I have for a long time! One day, I may understand half what was written. Like I have said before, "I don't do tech".

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure 11.2 min. to 14max EPR 1