I started using the RemZzzs CPAP mask liners two nights ago. When I checked the readings on my machine in the morning, it has displayed the grumpy face symbol the past two mornings, indicating an improper mask fit. I haven't had any leaks that have awakened me the past two nights and have felt generally well rested the past two days. I intend to check the Sleepyhead graphs tonight to verify the leak level. Could the machine possibly be indicating an improper fit simply because of the liners? Thank you for any information.
Chuck
RemZzzs CPAP Mask Liners; grumpy face on CPAP machine.
- Chuck Connors
- Posts: 257
- Joined: Fri Apr 20, 2012 7:55 am
Re: RemZzzs CPAP Mask Liners; grumpy face on CPAP machine.
Some people report that leaks are worse with mask liners but that at least the leaks aren't so loud that they keep waking a person up.
Mr Grumpy means that for at least 30% or more of the night your leaks were above the 24 L/min mark. He doesn't show up until the 30% mark is hit.
Now maybe if the worse you hit was 27 or 28 L/min then it might not be that big of a deal but if you spent much of the night over 30 L/min the therapy could potentially be compromised and the data reported be inaccurate.
It's not like things go totally in the toilet if you just barely cross the 24 L/min line...the further you go the worse it gets. So how bad things really were has to be evaluated by looking at the leak line itself.
Generally if Mr Grumpy shows up things are usually pretty bad though. I have had some pretty ugly leak lines and yet still not earn a Mr Frowny.
Mr Grumpy means that for at least 30% or more of the night your leaks were above the 24 L/min mark. He doesn't show up until the 30% mark is hit.
Now maybe if the worse you hit was 27 or 28 L/min then it might not be that big of a deal but if you spent much of the night over 30 L/min the therapy could potentially be compromised and the data reported be inaccurate.
It's not like things go totally in the toilet if you just barely cross the 24 L/min line...the further you go the worse it gets. So how bad things really were has to be evaluated by looking at the leak line itself.
Generally if Mr Grumpy shows up things are usually pretty bad though. I have had some pretty ugly leak lines and yet still not earn a Mr Frowny.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: RemZzzs CPAP Mask Liners; grumpy face on CPAP machine.
The machine doesn't know you are using liners....It just knows there is a leak.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: RemZzzs CPAP Mask Liners; grumpy face on CPAP machine.
I found the Remzzz liners to be thin, loosely woven and very prone to curling and getting stretched out of shape (size). Expensive, too. I bought some thick/densely woven micro-fiber cloths and cut mask liners from them. They're working 1000% better than Remzzz, super soft and only a tiny fraction of the price of the Remzzz.
_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
- Chuck Connors
- Posts: 257
- Joined: Fri Apr 20, 2012 7:55 am
Re: RemZzzs CPAP Mask Liners; grumpy face on CPAP machine.
Thank you all for your input and suggestions. Very helpful.
Chuck
Chuck
Re: RemZzzs CPAP Mask Liners; grumpy face on CPAP machine.
To Gasper62:
Can you please gove me more specifics about the thick/densely woven micro-fiber cloths?
Who sold them to you?
Do you have contact information (phone or email)?
Can you please gove me more specifics about the thick/densely woven micro-fiber cloths?
Who sold them to you?
Do you have contact information (phone or email)?