Routine endoscopic screening with known sleep apnea - accommodations?
Routine endoscopic screening with known sleep apnea - accommodations?
It's time for the every-ten-years colonoscopy (and in my case, upper GI endoscopy). It was during the last one that the sleep apnea was discovered since they needed to use a device in my nose to give me more air.
I have told the clinic I have known sleep apnea and will ask them specifically if I need to bring my machine, since I am having monitored anesthesia care. This means I won't know what they're doing while they are doing it, which is the only way I'll have the endoscopy.
Has anyone here who's had routine endoscopy like these been asked to bring their machines? If not, do you know how they accommodated the tendency to develop apnea due to relaxation during anesthesia?
Thanks much.
Deb
I have told the clinic I have known sleep apnea and will ask them specifically if I need to bring my machine, since I am having monitored anesthesia care. This means I won't know what they're doing while they are doing it, which is the only way I'll have the endoscopy.
Has anyone here who's had routine endoscopy like these been asked to bring their machines? If not, do you know how they accommodated the tendency to develop apnea due to relaxation during anesthesia?
Thanks much.
Deb
Re: Routine endoscopic screening with known sleep apnea - accommodations?
With an endoscopy there is a hose down your throat...Very difficult to use a CPAP. Since they use conscious sedation, you should be awake in just a few minutes...plus...there is an anesthesiologist monitoring. I have had several procedures...both ends...and have never needed my CPAP.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Routine endoscopic screening with known sleep apnea - accommodations?
When I had an endoscopy due to my sleep apnoea they would not sedate me instead I had it fully awake with a local anaesthetic throat spray not the most pleasant experience I must say the tenancy to gag was the worstLSAT wrote: ↑Tue Oct 15, 2019 2:58 pmWith an endoscopy there is a hose down your throat...Very difficult to use a CPAP. Since they use conscious sedation, you should be awake in just a few minutes...plus...there is an anesthesiologist monitoring. I have had several procedures...both ends...and have never needed my CPAP.
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
- Bertha deBlues
- Posts: 314
- Joined: Sun Aug 20, 2017 8:36 am
Re: Routine endoscopic screening with known sleep apnea - accommodations?
I had an endoscopy under sedation one year ago with no problems. I was told my machine would not be necessary. When I alerted the anesthesiologist to my SA, she promised she would look after me. She kept her word. For my own peace of mind, my CPAP was in the car so that my husband could retrieve it if needed. Best of luck with your procedure. [Edit: I also had a colonoscopy last year with the same results. In both cases they put a tube down my throat to keep my airway open; slight discomfort upon waking up but it went away quickly.]
_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + P10 Nasal Pillow Mask Bundle |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Last edited by Bertha deBlues on Wed Oct 16, 2019 7:32 am, edited 1 time in total.
And in the end the love you take is equal to the love you make. - Paul McCartney
- ChicagoGranny
- Posts: 14582
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
-
- Posts: 198
- Joined: Wed Apr 18, 2018 12:07 pm
- Location: Chicago-ish
Re: Routine endoscopic screening with known sleep apnea - accommodations?
I have to have a colonoscopy every 3 years. I get sedated and have told them about my sleep apnea. They assured me that they would keep me breathing. I don't have to take my machine. I have no memory of anything going on, so I don't know if they use anything.
Re: Routine endoscopic screening with known sleep apnea - accommodations?
I work with hospitals on medical malpractice liability issues, and as a result, I spend a lot of time talking to risk managers and patient safety officers. There was a time when ignorance about sleep apnea posed some dangers during endoscopic procedures, but everyone is highly aware of the issues now. As long as the team has notice of your sleep apnea, they will take the necessary steps. I've had plenty of colonoscopies myself (colon cancer veteran) and a couple of uppers as well, and I've never been asked to bring my own machine. In the last ten years, as far as I can remember, the only malpractice cases I've seen arising out of failure to prevent hypoxia due to sedation in a patient with sleep apnea have been when the patient had not been diagnosed but exhibited signs or symptoms that should have aroused suspicion and led to testing.
_________________
Machine: AirSense 11 Autoset |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
-
- Posts: 60
- Joined: Sun Nov 26, 2017 6:47 pm
- Location: Oregon
Re: Routine endoscopic screening with known sleep apnea - accommodations?
I'm having a colonoscopy next month. I was told to bring my CPAP machine with me. The machine is only for use in recovery. As others here have mentioned, an anesthesiologist (should) keep a careful eye on you during the procedure itself.
It's probably the same for an endoscopic screening. They probably want you to have your machine on hand for the recovery time.
I'll bring my machine but I doubt if I'll need it. Whenever I've had a colonoscopy they bring me out of sedation/anesthesia just as the doctor is finishing up. I don't need to sleep afterward. The recovery time is (mostly?) for liability reasons.
It's probably the same for an endoscopic screening. They probably want you to have your machine on hand for the recovery time.
I'll bring my machine but I doubt if I'll need it. Whenever I've had a colonoscopy they bring me out of sedation/anesthesia just as the doctor is finishing up. I don't need to sleep afterward. The recovery time is (mostly?) for liability reasons.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
-
- Posts: 210
- Joined: Tue Feb 11, 2014 6:03 pm
- Location: MA
Re: Routine endoscopic screening with known sleep apnea - accommodations?
Speaking as a gastroenterologist who has performed many tens of thousands of procedures, there are many variables at play here. After administering the IV sedation to my patients for many years, our hospital system switched over to using propafol as the sedation agent administered by anesthesia trained MDs and nurse anesthetics. Sleep apnea is quite noticeable after sedation and it is something we dealt with daily. Most of the time, these patients did not have a diagnosis of OSA yet.
None of my patients ever brought a CPAP machine to their procedure that I was aware of. My main points are to realize that we would see numerous instances of obstructive apnea every day...it was nothing unusual. And most occurred in patients who were not diagnosed.
The intra-procedure staff and recovery staff are well versed in dealing with the issue.
None of my patients ever brought a CPAP machine to their procedure that I was aware of. My main points are to realize that we would see numerous instances of obstructive apnea every day...it was nothing unusual. And most occurred in patients who were not diagnosed.
The intra-procedure staff and recovery staff are well versed in dealing with the issue.
_________________
Machine: AirSense™ 10 Elite CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Humidifier: XT Heated Humidifier |
Additional Comments: Sleepyhead software via SD card |
- chunkyfrog
- Posts: 34459
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Routine endoscopic screening with known sleep apnea - accommodations?
Every one should have essential medical information wtitten down, and available
for medical personnel to consult.
You never know when you might miss mentioning something for one reason or another.
It is a shame that medical alert jewelry and accessories are treated like a cash cow.
for medical personnel to consult.
You never know when you might miss mentioning something for one reason or another.
It is a shame that medical alert jewelry and accessories are treated like a cash cow.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Routine endoscopic screening with known sleep apnea - accommodations?
I get an endoscopy every two years. About 10 years ago, I was asked to bring the cpap. It is for the recovery period. But recently, they did not ask for CPAP. When I asked they said it is not necessary.
atab
atab
- ChicagoGranny
- Posts: 14582
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Routine endoscopic screening with known sleep apnea - accommodations?
You guys may be on the front lines of apnea diagnosis. I have two friends who were told to get sleep studies after their gastroenterologists observed apnea during their colonoscopies.RobertS975 wrote: ↑Mon Oct 21, 2019 12:03 pmSpeaking as a gastroenterologist who has performed many tens of thousands of procedures, there are many variables at play here.
Heart surgeons are also finding it.
Re: Routine endoscopic screening with known sleep apnea - accommodations?
Went under in the main hospital last year for a GI Endoscopy because the way my chest caves in scared the outpatient anesthesiologist team. Didn't need the CPAP machine but they did monitor my apnea while I was sleeping. I will say this though, I'm still a bit disturbed how alert I was when I woke up in the recovery area, making fart jokes and joining in on the Lottery fever discussion.
_________________
Machine: DreamStation Auto CPAP Machine with Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Phillips Dreamstation 2 Auto CPAP with Humidifier **
Re: Routine endoscopic screening with known sleep apnea - accommodations?
I had a recent colonoscopy & upper GI scope early this year since being diagnosed with apnea. I told them I had apnea before they proceeded & they informed me they were already aware of that. Both procedures happened with anesthesia & without me needing to bring my CPAP equipment.
_________________
Machine: AirStart™ 10 Auto CPAP with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
“We are what we repeatedly do. Excellence, then, is not an act, but a habit. ” —Aristotle.
-
- Posts: 30
- Joined: Fri Sep 13, 2019 1:06 am
Re: Routine endoscopic screening with known sleep apnea - accommodations?
My dad had endoscopy operation, he was not able to use CPAP mask. Now I am having sleep apnea, but luckily I can use CPAP mask. I did not have any operation.