New, Severe, Really Concerned
-
- Posts: 132
- Joined: Sat May 09, 2020 6:52 am
Re: New, Severe, Really Concerned
So at what time would you ever lower the pressure? If you have AHI is it a general rule that you keep rising until you no longer have AHI?
I am trying to understand the WHY not just what to do.
I am trying to understand the WHY not just what to do.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Humidifier: DreamStation Heated Humidifier |
Re: New, Severe, Really Concerned
Min. pressure is usually only ever lowered if someone can't find other answers for aerophagia, or if a new sleep study (or Oscar) shows the higher setting is unnecessary. The max setting may be lowered on occasion for other reasons depending on the situation, patient and doctors, but those usually are few and far between.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: New, Severe, Really Concerned
pretty much, though there are exceptions. what we want to try to accomplish with a higher minimum pressure so that the machine has a better chance at getting hypopneas and obstructives before they get too bad. better yet, so your airway is held open fully throughout the night.NewToSleepApnea wrote: ↑Tue May 19, 2020 2:19 pmIf you have AHI is it a general rule that you keep rising until you no longer have AHI?
I am trying to understand the WHY not just what to do.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: New, Severe, Really Concerned
The below graph is for someone with one of the popular brand Auto-style machines.
We titrate Min and Max to find the sweet spot - the Minimum pressure required to keep the airway open. For most people we leave Max open to 20 and bring the Min pressure up to keep the airway open all night. However some people have trouble with aerophagia. Others begin to get clusters of centrals when pressure goes too high. For both of these examples, sometimes it helps to bring the Max down to a reasonable level. Not everyone is challenged by aerophagia, and everyone's pressure regime is unique, so the graph is different for all of us. We each have to find our own pressure "sweet spot."
The green arrows point to an area in the green gradient that is the general sweet spot for this individual, compensating for positional and flow limitation pressure needs within set Min and Max pressure limits.
Hope this helps.
Chris
We titrate Min and Max to find the sweet spot - the Minimum pressure required to keep the airway open. For most people we leave Max open to 20 and bring the Min pressure up to keep the airway open all night. However some people have trouble with aerophagia. Others begin to get clusters of centrals when pressure goes too high. For both of these examples, sometimes it helps to bring the Max down to a reasonable level. Not everyone is challenged by aerophagia, and everyone's pressure regime is unique, so the graph is different for all of us. We each have to find our own pressure "sweet spot."
The green arrows point to an area in the green gradient that is the general sweet spot for this individual, compensating for positional and flow limitation pressure needs within set Min and Max pressure limits.
Hope this helps.
Chris
_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice
Be well,
Chris
Re: New, Severe, Really Concerned
what is the source of this graph? i want to have it on hand!DreamDiver wrote: ↑Tue May 19, 2020 8:44 pmThe below graph is for someone with one of the popular brand Auto-style machines.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: New, Severe, Really Concerned
I created it.zonker wrote: ↑Wed May 20, 2020 10:56 amwhat is the source of this graph? i want to have it on hand!DreamDiver wrote: ↑Tue May 19, 2020 8:44 pmThe below graph is for someone with one of the popular brand Auto-style machines.
https://i.imgur.com/oPsbvEw.jpg
C
_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice
Be well,
Chris
-
- Posts: 132
- Joined: Sat May 09, 2020 6:52 am
Re: New, Severe, Really Concerned
Last night was good. I did raise the min. pressure to 10 but I find it interesting that my pressure never really went up. Basically stayed around the minimum the whole time.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Humidifier: DreamStation Heated Humidifier |
Re: New, Severe, Really Concerned
<sigh> you and dog slobber are much more creative than i.DreamDiver wrote: ↑Wed May 20, 2020 11:01 amI created it.zonker wrote: ↑Wed May 20, 2020 10:56 amwhat is the source of this graph? i want to have it on hand!DreamDiver wrote: ↑Tue May 19, 2020 8:44 pmThe below graph is for someone with one of the popular brand Auto-style machines.
https://i.imgur.com/oPsbvEw.jpg
C
thanks!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: New, Severe, Really Concerned
Congratulations NewToSleepApnea!NewToSleepApnea wrote: ↑Wed May 20, 2020 11:13 amLast night was good. I did raise the min. pressure to 10 but I find it interesting that my pressure never really went up. Basically stayed around the minimum the whole time.
5-20-20.png
I think I remember LSAT saying something about how when you're at the right Min pressure, your pressure curve tends to be more even and regular. That's what's happening here, eh?
Why not leave it there for a few days to see how it averages out? You could tweak further, sure, but you're under 1.0 AHI. This is a good time to make small changes only infrequently.
Well done!
Chris
_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice
Be well,
Chris
-
- Posts: 132
- Joined: Sat May 09, 2020 6:52 am
Re: New, Severe, Really Concerned
Update. 2 Nights ago my AHI was low but I felt terrible when I got up. Turns out I kept waking up in the middle of the night and falling asleep without the mask on. My O2 dropped 40 times that night but it seems as if they were all with the mask off. Which worries me because without the mask I'm apparently more of a mess than ever.
So I was determined last night to keep that mask on. I did, but again, couldn't get to good sleep until 430am. I had a few good days but I'm back to feeling like there is a bag over my head. I know it says AHI has been good, but I just feel sleep deprived.
So I was determined last night to keep that mask on. I did, but again, couldn't get to good sleep until 430am. I had a few good days but I'm back to feeling like there is a bag over my head. I know it says AHI has been good, but I just feel sleep deprived.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Humidifier: DreamStation Heated Humidifier |
- Attachments
-
- 5-22-20.png (121.51 KiB) Viewed 2845 times
-
- Posts: 132
- Joined: Sat May 09, 2020 6:52 am
Re: New, Severe, Really Concerned
How Normal / acceptable is it for my O2 to drop to 86 with sleeping with the cpap. It happens every night, but really only in the first half when I’m trying to get to sleep.
I have a recorder. Here is what it looked like for 3 hours tonight (first hour away - all major activity during attempts to sleep).
I have a recorder. Here is what it looked like for 3 hours tonight (first hour away - all major activity during attempts to sleep).
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Humidifier: DreamStation Heated Humidifier |
Re: New, Severe, Really Concerned
It's actually not worth keeping, it makes too many unfounded assumptions.zonker wrote: ↑Wed May 20, 2020 10:56 amwhat is the source of this graph? i want to have it on hand!DreamDiver wrote: ↑Tue May 19, 2020 8:44 pmThe below graph is for someone with one of the popular brand Auto-style machines.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 132
- Joined: Sat May 09, 2020 6:52 am
Re: New, Severe, Really Concerned
Also, so every sleep doctor I’ve talked to says “you must have had this without knowing it” yet, before this started last month - I only needed 5/6 hours of sleep and felt good. Now, with the CPAP, I need to be in bed for 10/11 hours to feel okay.
How is this possible too? Is this common for anyone?
How is this possible too? Is this common for anyone?
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Humidifier: DreamStation Heated Humidifier |
Re: New, Severe, Really Concerned
Nope, it's backwards.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: New, Severe, Really Concerned
I am having the same issue. It was a sudden onset for me. I have 1000 questions for @ newtosleepapnea and anyone else who is experiencing this. I am absolutely restless and nobody knows what's going on. It's the same.... sleep apnea, and this cannot be sleep apnea based on what I am experiencing. Your story is my story to the T. I would like to connect with you if possible?