Well, I think you just described ideal diagnostic & calibration procedures.Sludge wrote:cpaman wrote:I don't know if we could just change out the sensor, and if the chip would automatically re-adjust / re-calibrate the pressure.I would say it can't. In order to calibrate the flow and pressure sensors, it would have to send in a series of known flows and pressures (see above Service Report where this is done). Without additional flow and pressure sensors, it can never be sure what the flow and pressure is at any given time.-SWS wrote:Dunno. But I'd be surprised if Respironics didn't write firmware code to compensate for ordinary sensor drift. Zero-sum inspiratory & expiratory flow reversals should make that a fairly easy task.
Further, "IMHO" it can't even zero. It would have to be sure there is no flow or pressure, and that scenario can never be assumed to be true in the field. The procedure would, out of necessity, have to be done at start up, and since plenty of people put the mask on first, then turn on the machine, there is never a true zero state.
That said, if there were such a routine, then there should be an indicator if it fails, and if we look at the available Error Codes for flow and pressure:
Pressure Sensor Errors
E-70 ERR_PRESSURE_SENSOR_ABSENT
E-72 ERR_PSENS_UNABLE_TO_OBTAIN_BUS
E-73 ERR_SENSOR_PRESS_OFFSET_STOP
Flow Sensor Errors
E-80 ERR_UNABLE_TO_INIT_FLOW_SENSOR
E-81 ERR__FLOW_SENSOR_TABLE
E-82 ERR_FLOW_SENSOR_OFFSET
E-83 ERR_FSENS_UNABLE_TO_OBTAIN_BUS
E-84 ERR_FLOW_SENSOR_STOP
E-85 ERR_FLOW_SENSOR_OCCLUDED
E-86 ERR_FLOW_SENSOR_ABSENT
E-87 ERR_FLOW_SENSOR_BUS
there are those OFFSET things, but it cannot necessarily be assumed that where you're offsetting from is actually zero.
That said2, in a review of Respironics technology, it seems that Flow Offset only comes into play when the patient is actually "doing something":
That said3, that certainly does not mean that once one exceeds zero flow and/or pressure, that those delivered parameters are accurate."FlowOffset" corresponds to a final value of patient flow during inspiration responsive to the final value being positive, otherwise FlowOffset is set to zero.
By contrast, I'm referring to a practical and on-the-fly sensor-drift compensation routine, which I think is algorithmically possible. Given that sensor drift is inevitable over time, I suspect the manufacturers might reference zero-flow on-the-fly, by noting where inspiratory and expiratory volumes break even. Sensor drift and orifice flow would each be fixed functions, with non-fixed leaks as the routine's potential spoiler. While that type of zero-flow referencing would be less desirable than bench calibration, I think not compensating for inevitable sensor drift may be even less desirable. Then again, that thought begs the question of whether the typical sensor-drift range stands to impair xPAP efficacy.