Hi, I'm new here but have been using CPAP for fifteen years; REALLY GLAD that I found this board. I opted for a new System One machine and am having a difficult time getting enough humidity out of it. Set on 5, it uses very little water, about 1/3 that small tank. So I turned the auto system off via the user menu and switched it to conventional mode. That was even worse, the thing used perhaps 1/4 tank of water and now my sinus is irritated badly.
I did take the machine back to the supplier who checked it by turning it on and feeling the plate, that was prior to the last two nights use. Why did I waste my time with that trip? I would think this machine on conventional mode would collect water in the hose if turned all the way up. One other issue was that I noticed the pressure was set a 11, where I use 10, so I dropped it back today. Fortunately I found a website that shows how to get into the same setup menu as the clinic uses.
Thanks in advance!
I did take the machine back to the supplier who checked it by turning it on and feeling the plate, that was prior to the last two nights use. Why did I waste my time with that trip? I would think this machine on conventional mode would collect water in the hose if turned all the way up. One other issue was that I noticed the pressure was set a 11, where I use 10, so I dropped it back today. Fortunately I found a website that shows how to get into the same setup menu as the clinic uses.
Thanks in advance!
Re: Need Help with System One Machine...
Welcome to the board.
Just to be sure, does the "system one humidification" option say "off" and the humidification say "C5"?
Just to be sure, does the "system one humidification" option say "off" and the humidification say "C5"?
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Need Help with System One Machine...
Thank you for the welcome
Would that C5 be the dial setting on the top of the machine? If so, yes.. The system one definately said "off" and I had the dial for the humidity set all the way up to five with the system both off and on.
Would that C5 be the dial setting on the top of the machine? If so, yes.. The system one definately said "off" and I had the dial for the humidity set all the way up to five with the system both off and on.
Re: Need Help with System One Machine...
With the machine turned off, turn the dial to "Setup". Press the dial. Turn the knob to "humidifier." Be sure it says "C5," not just "5". That indicates the smart humidification is off and that it's set to the highest level.Bayrat wrote:Thank you for the welcome
Would that C5 be the dial setting on the top of the machine? If so, yes.. The system one definately said "off" and I had the dial for the humidity set all the way up to five with the system both off and on.
Use the machine for a while, turn it off, and see if the plate on the bottom of the humidifier tank is warm.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Need Help with System One Machine...
OK, I will hook it back up and check that particular setting, as is I have replaced it with my old S8 again until I can figure out what is up. The plate does get warm, I did that test, but not as warm as I thought it should based upon my other machine.
Re: Need Help with System One Machine...
I have a system one Cpap. My machine uses very little water also. I can go about 4 nights on a tank with setting 3. System one humidification is
turned to on. I don't have any problems with dry sinus's, though. My nose is always kinda wet and runny.
Jerry
turned to on. I don't have any problems with dry sinus's, though. My nose is always kinda wet and runny.
Jerry
Re: Need Help with System One Machine...
The prs1's do a very good job of preventing rainout. They have a relative humidity sensor and as a resultthe humidifier is very efficient at putting enough humidity in the air without going overboard. So you should never see moisture in the hose and the water usage will be very low unless you live in a very dry climate.
This may be very different than your old machine. With the humidifier set to C5, you may have too much humidity and that can be the reason from your sinus irritation.
This may be very different than your old machine. With the humidifier set to C5, you may have too much humidity and that can be the reason from your sinus irritation.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Need Help with System One Machine...
If you want more moisture you can add an HC 150 inline with your machine.Bayrat wrote:I would think this machine on conventional mode would collect water in the hose if turned all the way up.
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
Re: Need Help with System One Machine...
Thanks everyone for all the ideas. I will mess with it some more this week but for now the S8 is back on the nightstand. As for too much humidity, I guess anything is possible but in the past, this same sinus issue has occurred due to operator error i.e. not checking the reservoir and running out of water. Even with a mask collecting water I have never had a sinus problem. We all react differently, the first time on CPAP at the sleep center my sinus were dried out utilizing their "passover" humidifier.
Re: Need Help with System One Machine...
Oh, and to reiterate, the medical supplier set my pressure at 11 rather than 10 and I did not notice that until today. It seemed the darn thing was blowing too hard but then I attributed that to a worn out predecessor going soft over time and me getting used to that. Surprising how much difference 1 cm makes.
Re: Need Help with System One UPDATE!!
Update: I checked and yes, the setting internally is C5 and when used as a conventional CPAP humidification is almost nil. Took my sinus several days to recover from being cooked.
I contacted Phillips, explained that my medical supplier "tested" the machine by feeling the heat on the plate. Phillips returned my inquiry and told me to contact my medical supplier or my doctor for resolution, basically blowing me off. NEVER again will I deal with this company. I will continue to use my old S8 and perhaps I can fix this Phillips POS System One enough to have it for a spare. Don't know but certainly dissatisfied with the unit.
I contacted Phillips, explained that my medical supplier "tested" the machine by feeling the heat on the plate. Phillips returned my inquiry and told me to contact my medical supplier or my doctor for resolution, basically blowing me off. NEVER again will I deal with this company. I will continue to use my old S8 and perhaps I can fix this Phillips POS System One enough to have it for a spare. Don't know but certainly dissatisfied with the unit.
Re: Need Help with System One Machine...
Hi Bayrat,
Just thought I'd tell you what my machine does. If I turn the humidifier to 1, it uses up all the water by 3am and I have to wake up and add water. I know it's a thermostat thing...but I'm happy with it on zero. Plenty of humidity.
Jamis
Seriously, don't you think you're being a little silly here. What did you expect, that they'd send out an emergency factory service crew to fix your humidifier? If you call the Ford Corporate office with an issue, they send you to the dealer. Same thing here.Bayrat wrote: Phillips returned my inquiry and told me to contact my medical supplier or my doctor for resolution, basically blowing me off. NEVER again will I deal with this company.
Just thought I'd tell you what my machine does. If I turn the humidifier to 1, it uses up all the water by 3am and I have to wake up and add water. I know it's a thermostat thing...but I'm happy with it on zero. Plenty of humidity.
Jamis
- Carl LaFong
- Posts: 163
- Joined: Tue Sep 06, 2011 4:21 pm
- Location: Maryland
Re: Need Help with System One Machine...
OP, never mind how much water it uses -- At what setting do your sinuses not become irritated?
Re: Need Help with System One UPDATE!!
Does ResMed do any better in terms of being willing to deal directly with the patient?Bayrat wrote:Phillips returned my inquiry and told me to contact my medical supplier or my doctor for resolution, basically blowing me off. NEVER again will I deal with this company.
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: Need Help with System One Machine...
Silly you say?jamiswolf wrote:Hi Bayrat,
Seriously, don't you think you're being a little silly here. What did you expect, that they'd send out an emergency factory service crew to fix your humidifier? If you call the Ford Corporate office with an issue, they send you to the dealer. Same thing here.Bayrat wrote: Phillips returned my inquiry and told me to contact my medical supplier or my doctor for resolution, basically blowing me off. NEVER again will I deal with this company.
Just thought I'd tell you what my machine does. If I turn the humidifier to 1, it uses up all the water by 3am and I have to wake up and add water. I know it's a thermostat thing...but I'm happy with it on zero. Plenty of humidity.
Jamis
What I expect is that a problem with a NEW machine, under warranty, would be addressed by the manufacturer after being informed that the medical supply company is NOT going to do any more with it than see if the plate gets warm. THAT is what I expect, not certain what your expectations are after purchasing a new product. WTF is my doctor going to do with it, perform surgery on it?? Sorry, but I don't lay down just because someone kicks me i.e. this manufacturer.