Dreamstation 2 won'r turn on
Dreamstation 2 won'r turn on
My Dreamstation 2 used to turn on automatically. Now no matter how hard I huff and puff, it stays off. Water chamber is fitting snuggly
_________________
Machine | Mask | Humidifier | ||
Additional Comments: Getting tired of feeling exhausted so often during the day and having to nap |
Coming back to learn some more
Re: Dreamstation 2 won'r turn on
Sounds like your Auto on feature has got turned off by someone.
You should be able to go into the clinical settings menu and turn it back on. Do you know how to do that?
But for now....what happens if you just push the on/off button to turn the machine on and off?
You should be able to go into the clinical settings menu and turn it back on. Do you know how to do that?
But for now....what happens if you just push the on/off button to turn the machine on and off?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Dreamstation 2 won'r turn on
Nope, auto on and off on the Dream Station 2 is enabled or disabled through Orchestrator (DME or other care provider) and can not be done from the clinical menu.Pugsy wrote: ↑Sun Jan 07, 2024 9:40 pmSounds like your Auto on feature has got turned off by someone.
You should be able to go into the clinical settings menu and turn it back on. Do you know how to do that?
But for now....what happens if you just push the on/off button to turn the machine on and off?
May not be the problem, but perhaps the fine air filter is plugging up with dust and restricts air flow so the machine is not sensing your breathing as it had done in the past. My own DS2 will auto turn on, but I have to cover the vent holes on my full face mask (F&P Vitera) to get it to do so. There is a setting for Respironics mask types in the settings (OFF, X1-X5) which allow compensation for air flow resistance of different masks. Perhaps experimenting with those values may affect the sensitivity of the auto on function.
Have you changed your mask type or hose size which might affect air flow and the machines sensitivity to the auto on function?
_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Resmed AirSense 10 Auto with Humidifier |
- chunkyfrog
- Posts: 34458
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Dreamstation 2 won'r turn on
Or there could be a loose connection somewhere in the path of airflow.
It could even be an imperfect seal on the tank.
It could even be an imperfect seal on the tank.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- ChicagoGranny
- Posts: 14573
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
- babydinosnoreless
- Posts: 2325
- Joined: Fri Nov 02, 2018 2:53 pm
Re: Dreamstation 2 won'r turn on
Grumpy48 wrote: ↑Mon Jan 08, 2024 7:15 amNope, auto on and off on the Dream Station 2 is enabled or disabled through Orchestrator (DME or other care provider) and can not be done from the clinical menu.Pugsy wrote: ↑Sun Jan 07, 2024 9:40 pmSounds like your Auto on feature has got turned off by someone.
You should be able to go into the clinical settings menu and turn it back on. Do you know how to do that?
But for now....what happens if you just push the on/off button to turn the machine on and off?
Wow the more I hear about the respironics the happier I am I have resmed devices. .
After the Dme made my settings unusable, I figured out how to set it myself on the resmed.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Re: Dreamstation 2 won't turn on
The implication of the original posting is that the machine can be used as long as it's turned on and off manually. Assuming that's correct, he does not have an emergency on his hands.
- ChicagoGranny
- Posts: 14573
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Dreamstation 2 won't turn on
No one said it was an emergency. They were just trying to restore a convenient feature.
- chunkyfrog
- Posts: 34458
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Dreamstation 2 won'r turn on
The feature can be very welcome for many patients with diminished vision or motor abilities.
So, for some, this could indeed be a crisis.
Eff respironics! Is their plan to make their junk completely useless?
So, for some, this could indeed be a crisis.
Eff respironics! Is their plan to make their junk completely useless?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Dreamstation 2 won'r turn on
While the auto on/off setting may not be user selectable, the default setting is 'Auto On' is 'On' and 'Auto Off' is 'Off' and this can be verified in OSCAR. I would be inclined to believe this is how most people would want it set. The default 'Auto On' enabled is good if one forgets to press the start button and provides a bit of insurance that the therapy should start if the button press is neglected. Setting 'Auto On' to 'Off' introduces a risk of therapy not starting if the start button is not pushed and likely why Philips chose to not make it user selectable. I would not rely on the Auto On function as I believe that for it to work reliably relies on the type of mask used ('resistance' values & vent leakage) and perhaps the filter condition which could affect air flow. For me the default Auto On/Off function is transparent and I have no need or desire to change it.
Presuming patients with diminished vision or motor abilities have a healthcare provider or DME (Philips likely believes you have one or the other or both), I would assume the Auto On/Off function could then be changed as needed via Orchestrator on request.
Presuming patients with diminished vision or motor abilities have a healthcare provider or DME (Philips likely believes you have one or the other or both), I would assume the Auto On/Off function could then be changed as needed via Orchestrator on request.
_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Resmed AirSense 10 Auto with Humidifier |