Whats best time of day to download ResScan Data?
- LittleRedTruck
- Posts: 183
- Joined: Wed Feb 22, 2006 9:19 am
Whats best time of day to download ResScan Data?
I seem to recall there was something important about 12 noon for downloading ResScan data. But don't know why. So wondering when to download last nites data, before noon or afternoon? Is the Data retained on a noon to noon basis or am I totally wrong ? Probably so !!! Dan
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 Vpap Adapt SV |
Re: Whats best time of day to download ResScan Data?
Yes, that's correct, however if you do all of your sleeping between noon of one day and noon the next, then it doesn't matter when you download the SD card info.
However, if you sleep, say 9am until 5pm, then your data will be split. I have no idea how to remedy that, or if there is a way, but if there is - someone here would have found it.
Cheers,
xena
However, if you sleep, say 9am until 5pm, then your data will be split. I have no idea how to remedy that, or if there is a way, but if there is - someone here would have found it.
Cheers,
xena
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-) |
Re: Whats best time of day to download ResScan Data?
What you are vaguely recalling is the need to look at the overnight data on the S9's LCD before noon if you want to see the overnight data on the LCd. The S9 resets the LCD's displayed "one day" data at noon local time each day.LittleRedTruck wrote:I seem to recall there was something important about 12 noon for downloading ResScan data.
ResScan doesn't care when the download is done. It downloads all the data on the card whenever you do it. You don't even have to download every day. Once you download all the data that's been written to the card since the last time you downloaded the data should be visible.
There is one caveat, however: If you want to see the detailed wave form data, you MUST download "All available data" at least once a week. The reason why is that the S9 overwrites the waveform data when it's eight days old---in other words, there are only seven nights of waveform data on the SD card at any time if I recall correctly.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: Whats best time of day to download ResScan Data?
You do recall correctly.robysue wrote:LittleRedTruck wrote:..in other words, there are only seven nights of waveform data on the SD card at any time if I recall correctly.
_________________
Mask: Mirage™ SoftGel Nasal CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Ray
Diagnosed in 1997
Diagnosed in 1997
Re: Whats best time of day to download ResScan Data?
In looking at data and thinking about this I would suggest a 9am to 5pm sleeper to set the CPAP clock back 12 hours, so that the CPAP clock changes its recording day at midnight real clock time. This way the date stamp on the downloaded data will correspond to the sleep date. The S9 assigns data collected between midnight and noon (CPAP clock time) to the date of the day before.xenablue wrote:However, if you sleep, say 9am until 5pm, then your data will be split. I have no idea how to remedy that, or if there is a way, but if there is - someone here would have found it.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7 |