Defective machine?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
jbenenson
Posts: 63
Joined: Fri Oct 08, 2010 1:17 pm

Defective machine?

Post by jbenenson » Tue Nov 30, 2010 4:07 pm

I have had my S9 Autoset for about a month and have experimented with different settings (thanks to this forum for how-to). I have found that regardless of what I do to change the SmartStart feature it remains grayed out. Have you other S9 Autoset owners been able to adjust it or is it inoperable in your machine too? If I have a defective machine I have only 23 months remaining to get a new one under warranty.
Never, never, never, never say never

Janknitz
Posts: 8413
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Defective machine?

Post by Janknitz » Tue Nov 30, 2010 6:42 pm

Is the option turned off in the clinician's menu? Do you have access to the clinician's menu???
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
BanjoPaterson
Posts: 93
Joined: Mon Nov 01, 2010 8:15 pm
Location: Canberra, Australia

Re: Defective machine?

Post by BanjoPaterson » Tue Nov 30, 2010 6:51 pm

Hi - this happened to me too...

I bet you have a full face mask or the mask selection is full face. The smartstart doesn't happen with a full face mask, only nasal or pillows (for whatever reason).

_________________
MaskHumidifier

User avatar
jbenenson
Posts: 63
Joined: Fri Oct 08, 2010 1:17 pm

Re: Defective machine?

Post by jbenenson » Tue Nov 30, 2010 6:54 pm

Thanks for your replies. I have the Clinician's Manual (that I downloaded from here) but I couldn't find anything relating to SmartStart and full face masks. I guess I'll have my S9 for a while.
Never, never, never, never say never

sleepnationtv
Posts: 283
Joined: Mon Jul 19, 2010 7:23 pm

Re: Defective machine?

Post by sleepnationtv » Tue Nov 30, 2010 8:24 pm

The smart start will only work with nasal and pillows masks. If you have a fullface mask you can still set the mask type as nasal and the smart start should work. go into the clinical menu under the options tab and turn it on.

Good luck,
Dane Schapper, Founder cpaplibrary.com
Check out the cpaplibrary video bloghttp://www.cpaplibrary.com