I have the clinician's manual, but between hubby and I we can't see where to change the time.
Have to presume the S9 Elite autosets for DST?
Thanks for any input.
Cheers,
Xena
Can't find where to change time on my machine
Can't find where to change time on my machine
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-) |
Re: Can't find where to change time on my machine
No, the S9 doesn't reset itself to non-daylight savings time.
Here's how to change the time:
Step 1) Get into the clinician's menu by pressing the double check mark button and the big round button simultaneously for about 3 seconds.
Step 2) Use the big round button to slide FROM the icon on the left that looks like a pair of gears (where the therapy settings are) down to the next icon.
Step 3) Click the big round button to select this submenu and scroll down until you get to time. Click to select time as the thing to change.
Step 4) Click to set the hour and rotate the round button appropriately. Click when the hour is correct.
Step 5) Use the big round button to slide all the way down to Home and click to get out of the clinician's menu.
Here's how to change the time:
Step 1) Get into the clinician's menu by pressing the double check mark button and the big round button simultaneously for about 3 seconds.
Step 2) Use the big round button to slide FROM the icon on the left that looks like a pair of gears (where the therapy settings are) down to the next icon.
Step 3) Click the big round button to select this submenu and scroll down until you get to time. Click to select time as the thing to change.
Step 4) Click to set the hour and rotate the round button appropriately. Click when the hour is correct.
Step 5) Use the big round button to slide all the way down to Home and click to get out of the clinician's menu.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |
Re: Can't find where to change time on my machine
Thanks, robysue! Sheesh! I couldn't find it in the manual - my printer must've skipped that page
Cheers,
Xena
Cheers,
Xena
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-) |
Re: Can't find where to change time on my machine
Thanks Robysue and Zenablue. Glad you are there to help.
Will
Will