CPAP without Apnea?
-
STLWantRest
- Posts: 2
- Joined: Tue Dec 21, 2010 1:30 am
CPAP without Apnea?
I am a heavy snorer and wanted to see if a CPAP might work.
I have snored very loudly my entire life. I've had nasal passage surgeries and nothing has helped. Had a Sleep Study done and they said that I didn't have Sleep Apnea. I'm not overweight, even snored when I was real skinny as a kid, but I'm tired all day and want a solution!
Should I get a used CPAP and try it?
I have snored very loudly my entire life. I've had nasal passage surgeries and nothing has helped. Had a Sleep Study done and they said that I didn't have Sleep Apnea. I'm not overweight, even snored when I was real skinny as a kid, but I'm tired all day and want a solution!
Should I get a used CPAP and try it?
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: CPAP without Apnea?
I would.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
- BlackSpinner
- Posts: 9742
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: CPAP without Apnea?
Bad study or lousy sleep doc. I have friends who use prescribed cpap for snoring.STLWantRest wrote:I am a heavy snorer and wanted to see if a CPAP might work.
I have snored very loudly my entire life. I've had nasal passage surgeries and nothing has helped. Had a Sleep Study done and they said that I didn't have Sleep Apnea. I'm not overweight, even snored when I was real skinny as a kid, but I'm tired all day and want a solution!
Should I get a used CPAP and try it?
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
Re: CPAP without Apnea?
I am assuming you are in St. Louis also. Welcome to the forum. You have found a great resource here.
_________________
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Machine: S9 VPAP Adapt (eASV) 36037 |
Mode: ASVAuto, Min EPAP: 4, Max EPAP: 15, Min PS: 3, Max PS: 15, Ramp: Off - Original Titration: 18
-
no_more_headaches
- Posts: 160
- Joined: Sun Mar 16, 2008 3:20 pm
Re: CPAP without Apnea?
My wife snored loudly and felt tired all the time. She had a sleep study and it showed she did not have Apnea but sever Hypopnea (partial collapse) and her O2 levels drop to the 70s. I have severe apnea and stopped breathing up to a minute but my O2 levels only hit 89%.
I would think her situation is worse with the desats but I had an uncle die in his sleep at my age so ....
I would think her situation is worse with the desats but I had an uncle die in his sleep at my age so ....
OSA Sleep Test 11/23/07
AHI 29.5, Sat. 88%
Supine AHI 78.9 REM AHI 16
Titrated 1/18/08, Pressure 9
AHI 29.5, Sat. 88%
Supine AHI 78.9 REM AHI 16
Titrated 1/18/08, Pressure 9
Re: CPAP without Apnea?
That question is best answered by someone who has evaluated your sleep study. The odds are that experimenting with xPAP wouldn't cause you any harm, but it COULD. It is irresponsible for anyone with no knowledge of your case to recommend going out on your own. I recommend you ask your Dr about your plan and the associated risks before proceeding.
But, if you DO go ahead, I would suggest you find a used AutoPAP machine. It will adjust the pressure setting to meet your body's needs. If you use straight CPAP, you wouldn't have a clue what setting to use.
I hope you find a solution!
But, if you DO go ahead, I would suggest you find a used AutoPAP machine. It will adjust the pressure setting to meet your body's needs. If you use straight CPAP, you wouldn't have a clue what setting to use.
I hope you find a solution!
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...
Unlike his passengers who died screaming as the car went over the cliff...
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: CPAP without Apnea?
I would add to linkc's comments that you would be best be served by an Auto cpap machine with full data;
That way you would have feedback to give your doctor. I would also get a O2 monitor--(just my thought)
It would be best to have a doctor in on this with you--maybe it would have to be a different doc.
You might even have to pester the doctor to find the real reason for your fatigue.
Don't give up!
That way you would have feedback to give your doctor. I would also get a O2 monitor--(just my thought)
It would be best to have a doctor in on this with you--maybe it would have to be a different doc.
You might even have to pester the doctor to find the real reason for your fatigue.
Don't give up!
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
-
STLWantRest
- Posts: 2
- Joined: Tue Dec 21, 2010 1:30 am
Re: CPAP without Apnea?
Thanks to all for the rapid responses.
I will get a different doctor involved. Is it typically a Neurologist or ENT that you work with concerning the CPAP? My ENT was a very well-known and respected doctor in St. Louis. Not sure why he didn't go the CPAP route.
Thanks again!
I will get a different doctor involved. Is it typically a Neurologist or ENT that you work with concerning the CPAP? My ENT was a very well-known and respected doctor in St. Louis. Not sure why he didn't go the CPAP route.
Thanks again!
-
no_more_headaches
- Posts: 160
- Joined: Sun Mar 16, 2008 3:20 pm
Re: CPAP without Apnea?
ENTs are typically surgeons and in my experience are not as involved.
Sleep doctors can be pulmonogists or neurologists but it is a specialty.
I would look specifically for sleep doctors if you can.
Sleep doctors can be pulmonogists or neurologists but it is a specialty.
I would look specifically for sleep doctors if you can.
OSA Sleep Test 11/23/07
AHI 29.5, Sat. 88%
Supine AHI 78.9 REM AHI 16
Titrated 1/18/08, Pressure 9
AHI 29.5, Sat. 88%
Supine AHI 78.9 REM AHI 16
Titrated 1/18/08, Pressure 9
-
brazospearl
- Posts: 704
- Joined: Sun Nov 15, 2009 5:51 pm
Re: CPAP without Apnea?
Do you have a copy of your sleep study results? I think it'd be worth reading to see if you had any apneas or hypopneas, what your oxygen levels were, if there were any limb movements, etc. If, for instance, your doctor said you "didn't have sleep apnea" because you had 4 events instead of the standard of 5, but you also had hypopneas or desats, that would be good to know.
_________________
| Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
- parabellum
- Posts: 10
- Joined: Wed Dec 15, 2010 4:10 pm
Re: CPAP without Apnea?
Few notes:
1) I have "herioc" snoring. That basically means LOUD as hell. :-/ But the good news is that it almost completely stops when I use the CPAP. My GF described it as dropping from a 10 to a 0, with some rare nights where I may snore at a 1 or 2 level of loudness.
2) The Resmed S9 records snore levels. I'm not sure exactly how it does that, I presume there is a mic in the unit that listens through the tube...but it does do a very good job of showing me exactly when I do snore a bit.
If you can get the script, definitely give it a try, but watch out, if your pressure is too high it can have adverse effects (such as triggering central apneas).
1) I have "herioc" snoring. That basically means LOUD as hell. :-/ But the good news is that it almost completely stops when I use the CPAP. My GF described it as dropping from a 10 to a 0, with some rare nights where I may snore at a 1 or 2 level of loudness.
2) The Resmed S9 records snore levels. I'm not sure exactly how it does that, I presume there is a mic in the unit that listens through the tube...but it does do a very good job of showing me exactly when I do snore a bit.
If you can get the script, definitely give it a try, but watch out, if your pressure is too high it can have adverse effects (such as triggering central apneas).
Severe OSA. (Untreated AHI=65, O2 Desat to 77%)
Resmed Mirage Full Face Mask
Resmed S9 Elite w Humidity
Happy CPAP user.
Resmed Mirage Full Face Mask
Resmed S9 Elite w Humidity
Happy CPAP user.
- Lizistired
- Posts: 2835
- Joined: Tue Dec 14, 2010 10:47 pm
- Location: Indiana
Re: CPAP without Apnea?
If you or your insurance company has already paid for the sleep study, get the "Full Scored Data Summary Report with Condensed Graphs". (I think that's what it's called. If not someone will be kind enough to correct me.) Then you can read it and can take it to a sleep doctor.
My AHI was only 11, but I had 49 arousals an hour and awakenings that were unrelated to apneas or hypopneas. Tired wasn't the word for it. The cpap cleared most of the arousals too. I felt 100% better just on the 2 good hours of sleep I got during the titration. Go see a sleep doctor.
My AHI was only 11, but I had 49 arousals an hour and awakenings that were unrelated to apneas or hypopneas. Tired wasn't the word for it. The cpap cleared most of the arousals too. I felt 100% better just on the 2 good hours of sleep I got during the titration. Go see a sleep doctor.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Swift FX sometimes, CMS-50F, Cervical collar sometimes, White noise, Zeo... I'm not well, but I'm better. |
ResScan: http://www.resmed.com/int/assets/html/s ... c=patients
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
ResScan Tutorial- http://montfordhouse.com/cpap/resscan_tutorial/
Machines Video: http://www.cpaplibrary.com/machine-education
Re: CPAP without Apnea?
OK. I'm confused. I have a S9 Autoset, and when I run a detailed report I only get reporting on four variables; Pressure, Leak, AHI & AI, and Usage. I don't see any snoring data whatsover. Please explain.parabellum wrote:Few notes:
1) I have "herioc" snoring. That basically means LOUD as hell. :-/ But the good news is that it almost completely stops when I use the CPAP. My GF described it as dropping from a 10 to a 0, with some rare nights where I may snore at a 1 or 2 level of loudness.
2) The Resmed S9 records snore levels. I'm not sure exactly how it does that, I presume there is a mic in the unit that listens through the tube...but it does do a very good job of showing me exactly when I do snore a bit.
If you can get the script, definitely give it a try, but watch out, if your pressure is too high it can have adverse effects (such as triggering central apneas).
Re: CPAP without Apnea?
wonka,
When you download the data, make sure you select "download all available data." The snoring data is not in the summary graphs; it's only in the detailed graphs. When you select the detailed data graphs in ResScan, there's a little tab off to the right somewhere that controls which graphs show up on both the navigation (large) time scale at the top and the detailed time scale (at the bottom). Check there to make sure that the Snore graph is selected as one of the ones to show. Then it will show up as a function of time. The y-axis is a scale of how loud the snoring is and the x-axis is the time of night the snoring is occurring.
When you download the data, make sure you select "download all available data." The snoring data is not in the summary graphs; it's only in the detailed graphs. When you select the detailed data graphs in ResScan, there's a little tab off to the right somewhere that controls which graphs show up on both the navigation (large) time scale at the top and the detailed time scale (at the bottom). Check there to make sure that the Snore graph is selected as one of the ones to show. Then it will show up as a function of time. The y-axis is a scale of how loud the snoring is and the x-axis is the time of night the snoring is occurring.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5 |