Should I get an oxygen monitor
-
- Posts: 20
- Joined: Sat Jun 30, 2018 12:33 pm
- Location: Alabama
Should I get an oxygen monitor
I have been using my APAP 2 years. I averaged around 3.5 AHI's. Mid-November, my AHI's started averaging 7.9 with a high of 14.1 and a low of 3.4. Several things changed. 1. I started intermittent fasting. 2. I started a healthy, low carb, low-inflammation, keto diet. 3. My right shoulder (the side on which I normally sleep) is painful, so I am sleeping more on my left (which triggers my AFIB more). OSCAR shows that many of the AHI episodes are during periodic breathing. I contacted both my pulmonologist NP and my cardiologist. Neither seems concerned. The NP did up the pressure Min from 8 to 10 and max from 12 to 15. I don't see a change, yet. My concern is that it seems a long time in which I might not be getting sufficient oxygen. Would you advise getting a 24 hr monitoring device, and if so, which one? Thanks! OSCAR charts are attached for the worst night. Thank You!
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: ResMed AirTouch™ F20 Mask For Her with Headgear + 2 Replacement Cushions |
Humidifier: DreamStation Heated Humidifier |
- Attachments
-
- Small excerpt during the periodic breathing
- Small zoomed period.png (87.31 KiB) Viewed 1393 times
-
- Full night graph
- Full Night CPAP graph.png (86.55 KiB) Viewed 1393 times
Re: Should I get an oxygen monitor
Any chance you were awake during the clustering of CAs/Centrals?
I was looking at 12/9 at 3:36 and there's a cluster and it ends with the machine being turned off. Were you awake for a period of time before you turned the machine off?
Do you ever spend much time awake with the mask and machine on?
Try turning off Flex exhale relief and see what happens.
If you still get a large number of CAs/centrals and you were for sure asleep, then yes...I think if it were me I would invest in an overnight recording pulse oximeter.
I was looking at 12/9 at 3:36 and there's a cluster and it ends with the machine being turned off. Were you awake for a period of time before you turned the machine off?
Do you ever spend much time awake with the mask and machine on?
Try turning off Flex exhale relief and see what happens.
If you still get a large number of CAs/centrals and you were for sure asleep, then yes...I think if it were me I would invest in an overnight recording pulse oximeter.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Should I get an oxygen monitor
The contec CMS50F and CMS50i are wrist mounted oximeters that record overnight, the F only records one session, the I will record multiple sessions, as well as the perfusion index. They're both supported by Oscar, so you can sync the time on the oximeter, then import the data into Oscar to display along with your cpap data.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Should I get an oxygen monitor
This is the one I got, from Amazon.com
https://www.amazon.com/gp/product/B0796 ... UTF8&psc=1
I have a Mac, so their software doesn't work, but it syncs just fine with Oscar and is included in the daily report.
https://www.amazon.com/gp/product/B0796 ... UTF8&psc=1
I have a Mac, so their software doesn't work, but it syncs just fine with Oscar and is included in the daily report.
_________________
Mask: Pico Nasal CPAP Mask with Headgear |
Additional Comments: Resmed AirCurve 10 ASV and Humidifier, Oscar for Mac |
KatieW
-
- Posts: 20
- Joined: Sat Jun 30, 2018 12:33 pm
- Location: Alabama
Re: Should I get an oxygen monitor
Thanks, so much, ya'll. I will put a pen and paper by the bed and write down when I am awake. I will also look at more recent nights to help determine, if I was awake. I think that the episodes and AFIB woke me, but I will check this out. I will also try the setting change. Then, if I am asleep, I will pursue one of these O2 monitors. Again, THANK YOU!
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: ResMed AirTouch™ F20 Mask For Her with Headgear + 2 Replacement Cushions |
Humidifier: DreamStation Heated Humidifier |
-
- Posts: 20
- Joined: Sat Jun 30, 2018 12:33 pm
- Location: Alabama
Re: Should I get an oxygen monitor
Pugsy,
You are definitely on to something. I turned the flex off. I know that I need many samples to be significantly different, but I am definitely encouraged. Prior to turning the flex off: AHI-9.15, PB-14.52, CA-4.90, Obs-1.29, HA-2.96, RERA-1.16. After turning off the flex: AHI-3.37, PB-5.39%, CA-0.15, Obs-2.30, HA-0.92, RERA-4.14. It looks like I am very susceptible to the pressure changes. And, I was awake during many of the events. One series of events looks like it may have precipitated waking. I recorded that at 3:38, I woke and my heart was fluttery (or I was jittery). Last night was a terrible night for sleep, so the chart is very deceptive. We walked yesterday, so I could not get comfortable to go to sleep and did not drift off till sometime after 02:30. At 3:38, I awoke. Again, at 5:59, I awoke. But the AHI and periodic breathing while I was asleep are much better. I listened to a YouTube that LankyLefty had on the EPR/FLEX, and I think that my pressure ranges may need to be narrowed as well. THANK YOU, THANK YOU, THANK YOU!
You are definitely on to something. I turned the flex off. I know that I need many samples to be significantly different, but I am definitely encouraged. Prior to turning the flex off: AHI-9.15, PB-14.52, CA-4.90, Obs-1.29, HA-2.96, RERA-1.16. After turning off the flex: AHI-3.37, PB-5.39%, CA-0.15, Obs-2.30, HA-0.92, RERA-4.14. It looks like I am very susceptible to the pressure changes. And, I was awake during many of the events. One series of events looks like it may have precipitated waking. I recorded that at 3:38, I woke and my heart was fluttery (or I was jittery). Last night was a terrible night for sleep, so the chart is very deceptive. We walked yesterday, so I could not get comfortable to go to sleep and did not drift off till sometime after 02:30. At 3:38, I awoke. Again, at 5:59, I awoke. But the AHI and periodic breathing while I was asleep are much better. I listened to a YouTube that LankyLefty had on the EPR/FLEX, and I think that my pressure ranges may need to be narrowed as well. THANK YOU, THANK YOU, THANK YOU!
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: ResMed AirTouch™ F20 Mask For Her with Headgear + 2 Replacement Cushions |
Humidifier: DreamStation Heated Humidifier |
- Attachments
-
- CPAP Flex off.png (86.37 KiB) Viewed 1289 times
Re: Should I get an oxygen monitor
Sometimes easy peasy works....let's see how you do on a night where you know your sleep quality was better and you slept more soundly.
There's probably some false positive awake events getting flagged last night.
They muddy up the evaluation process.
Sometimes when people use exhale relief it creates a bilevel situation which can trigger centrals in numbers that are alarming.
It creates an unstable breathing pattern that perpetuates itself. It's not extremely common but it can happen.
Turning off exhale relief (or sometimes just reducing it) seems to keep that unstable breathing from ever starting or feeding itself.
Need more nights where you sleep more soundly to see for sure.
There's probably some false positive awake events getting flagged last night.
They muddy up the evaluation process.
Sometimes when people use exhale relief it creates a bilevel situation which can trigger centrals in numbers that are alarming.
It creates an unstable breathing pattern that perpetuates itself. It's not extremely common but it can happen.
Turning off exhale relief (or sometimes just reducing it) seems to keep that unstable breathing from ever starting or feeding itself.
Need more nights where you sleep more soundly to see for sure.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Should I get an oxygen monitor
What you're susceptible to is the extra ventilation that flex causes, which drives down your CO2 levels, which depresses your respiration.CorgiWalker wrote: ↑Sun Dec 15, 2019 1:49 pmIt looks like I am very susceptible to the pressure changes.
NOT the pressure changes, of which there are many, that reduce your obstructive breathing events.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 20
- Joined: Sat Jun 30, 2018 12:33 pm
- Location: Alabama
Re: Should I get an oxygen monitor
Thank you! I didn’t state that well. By pressure changes, I meant the reduce by 3 from the inhale to the exhale. I will go awhile and then feedback.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: ResMed AirTouch™ F20 Mask For Her with Headgear + 2 Replacement Cushions |
Humidifier: DreamStation Heated Humidifier |
Re: Should I get an oxygen monitor
Aflex doesn't reduce by 3... those setting numbers have no direct relation to the amount of exhale relief.CorgiWalker wrote: ↑Sun Dec 15, 2019 4:14 pmThank you! I didn’t state that well. By pressure changes, I meant the reduce by 3 from the inhale to the exhale. I will go awhile and then feedback.
I correct these misinterpretations to prevent others from getting the wrong idea.
A small percentage of people are sensitive to the extra ventilation and end up with central apneas... you're apparently one of that small group.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 20
- Joined: Sat Jun 30, 2018 12:33 pm
- Location: Alabama
Re: Should I get an oxygen monitor
Thanks! Must be an algorithm to some change.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: ResMed AirTouch™ F20 Mask For Her with Headgear + 2 Replacement Cushions |
Humidifier: DreamStation Heated Humidifier |
Re: Should I get an oxygen monitor
Hi CorgiWalker,
One of the things about intermittent fasting and a keto diet is that it's difficult to get enough potassium. Some of the foods that are highest in potassium are cut out of your diet.
I had some palpitations before I started with keto, and they only got worse. One of the reasons I was going keto was to help my heart. I thought I was getting enough potassium, but I wasn't. I was supplementing, but the largest dose you can buy is 100mg. The RDA is 4700mg. That is a lot.
There are many videos on YouTube about keto and potassium.
Also, here is a link to a study about AFIB and potassium.
Cheers!
https://www.ncbi.nlm.nih.gov/pubmed/24012173
One of the things about intermittent fasting and a keto diet is that it's difficult to get enough potassium. Some of the foods that are highest in potassium are cut out of your diet.
I had some palpitations before I started with keto, and they only got worse. One of the reasons I was going keto was to help my heart. I thought I was getting enough potassium, but I wasn't. I was supplementing, but the largest dose you can buy is 100mg. The RDA is 4700mg. That is a lot.
There are many videos on YouTube about keto and potassium.
Also, here is a link to a study about AFIB and potassium.
Cheers!
https://www.ncbi.nlm.nih.gov/pubmed/24012173
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
Using weight loss, general exercise, and tongue/throat exercises I managed to get my AHI down to approx 5.
Not using a machine currently.
Not using a machine currently.
Re: Should I get an oxygen monitor
I'm sure I found this info somewhere on this forum, but I found Mac software compatible with the CMS50F here:KatieW wrote: ↑Sat Dec 14, 2019 6:05 pmThis is the one I got, from Amazon.com
https://www.amazon.com/gp/product/B0796 ... UTF8&psc=1
I have a Mac, so their software doesn't work, but it syncs just fine with Oscar and is included in the daily report.
http://torontek.com/product/torontek-b400/
The last time I tried OSCAR it wasn't yet compatible with newer CMS50F firmware, maybe it is now.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
-
- Posts: 20
- Joined: Sat Jun 30, 2018 12:33 pm
- Location: Alabama
Re: Should I get an oxygen monitor
Thanks! I supplement potassium, but no where near this amount. I will rearch.
And, although my phone and tablet are iOS, my computer is Windows, so I can use OSCAR.
I told my allergist about OSCAR this week. He is a hose head and was very interested in it. He said that if he gets a 90%, he wanted to see what and when it was happening. I told him that this would show all the charts through the night.
And, although my phone and tablet are iOS, my computer is Windows, so I can use OSCAR.
I told my allergist about OSCAR this week. He is a hose head and was very interested in it. He said that if he gets a 90%, he wanted to see what and when it was happening. I told him that this would show all the charts through the night.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: ResMed AirTouch™ F20 Mask For Her with Headgear + 2 Replacement Cushions |
Humidifier: DreamStation Heated Humidifier |