Tiny Hole in CPAP Hose - Only six weeks old
- SuperGeeky
- Posts: 414
- Joined: Mon May 19, 2008 3:55 pm
- Location: MaryEtta, Gawga (Now a part of the Union)
Tiny Hole in CPAP Hose - Only six weeks old
I was washing my CPAP 6' Hose this morning when I noticed a tiny stream coming out. Totally shocked!! I had never anticipated looking for a small pinhole in the hose...
I had trouble breathing last night. I just couldn't get enough air. And, I took the mask off and slept without it the last two hours or so. I checked my Encore Report this morning, leak rate was a solid 30-35 when it should have been 18-20. Mysterious, because I'll usually wake up to a Mask leak.
Lesson learned:
Literally, pressure check everything.
1. Cusp one end of hose and either blow into it checking for pressure leaks or like this morning, fill it with water from the sink.
2. Cusp the end of the mask and Diffuser, blow into it as hard as possible to detect any leaks.
I was so lucky to find this problem. Cognitively, it could of gone on for quite some time. **The hose was only six weeks old**
Input welcome,,
SG
I had trouble breathing last night. I just couldn't get enough air. And, I took the mask off and slept without it the last two hours or so. I checked my Encore Report this morning, leak rate was a solid 30-35 when it should have been 18-20. Mysterious, because I'll usually wake up to a Mask leak.
Lesson learned:
Literally, pressure check everything.
1. Cusp one end of hose and either blow into it checking for pressure leaks or like this morning, fill it with water from the sink.
2. Cusp the end of the mask and Diffuser, blow into it as hard as possible to detect any leaks.
I was so lucky to find this problem. Cognitively, it could of gone on for quite some time. **The hose was only six weeks old**
Input welcome,,
SG
Re: Tiny Hole in CPAP Hose - Only six weeks old
Many people carry a spare hose as a backup as they are inexpensive.
In the mean time you could repair the hose using gorilla tape which can be found at most hardware stores.
In the mean time you could repair the hose using gorilla tape which can be found at most hardware stores.
Re: Tiny Hole in CPAP Hose - Only six weeks old
Maybe a cat, dog, mouse, sharp corner on the nightstand, fingernail, subconsciously chewing on it during sleep, or pulled to hard.
I had the cat thing happen to it once. They don't taste as good as when I have them at a restaurant.
Two words for something to keep by the bed. Duct Tape
Allen
I had the cat thing happen to it once. They don't taste as good as when I have them at a restaurant.
Two words for something to keep by the bed. Duct Tape
Allen
- SuperGeeky
- Posts: 414
- Joined: Mon May 19, 2008 3:55 pm
- Location: MaryEtta, Gawga (Now a part of the Union)
Re: Tiny Hole in CPAP Hose - Only six weeks old
It's not a matter of fixing it.... It just caught me off guard!! DME usually replaces every three months which gives you a certain peace of mind. I wasn't expecting this...
When I'm tired, as I am now, I'll go to sleep even though I'm aware of not getting enough air. I'm just too tired to put two and two together. Really, I have to push myself hard to create a higher state of awareness. Takes a lot of energy.... And cognitive abilities can be dull....
Now I know what the Astronauts must feel when they space walk. I'm really surprised a small pinhole could do this to my therapy!!
SG
When I'm tired, as I am now, I'll go to sleep even though I'm aware of not getting enough air. I'm just too tired to put two and two together. Really, I have to push myself hard to create a higher state of awareness. Takes a lot of energy.... And cognitive abilities can be dull....
Now I know what the Astronauts must feel when they space walk. I'm really surprised a small pinhole could do this to my therapy!!
SG
- SuperGeeky
- Posts: 414
- Joined: Mon May 19, 2008 3:55 pm
- Location: MaryEtta, Gawga (Now a part of the Union)
Re: Tiny Hole in CPAP Hose - Only six weeks old
I forgot to add, I had a towel wrapped around the hose to prevent 'rainout'. Dah, guess I wouldn't notice a leak....
Re: Tiny Hole in CPAP Hose - Only six weeks old
Addendum:
3. Don't wash your hose. You'll poke holes in it.
3. Don't wash your hose. You'll poke holes in it.
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...
Unlike his passengers who died screaming as the car went over the cliff...
Re: Tiny Hole in CPAP Hose - Only six weeks old
LinkC wrote:Addendum:
3. Don't wash your hose. You'll poke holes in it.
Ditto.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
- BlackSpinner
- Posts: 9745
- Joined: Sat Apr 25, 2009 5:44 pm
- Location: Edmonton Alberta
- Contact:
Re: Tiny Hole in CPAP Hose - Only six weeks old
My hose is now almost a year old and covered with duct tape patches where the cat stepped on it and slid off, digging in her nails.
_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine |
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear |
Additional Comments: Quatro mask for colds & flus S8 elite for back up |
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal
- bearded_two
- Posts: 459
- Joined: Mon Aug 10, 2009 8:01 pm
Re: Tiny Hole in CPAP Hose - Only six weeks old
A few tine pinholes aren't going to materially affect your leak rate. Compare the size of the pin holes to the size of the exhaust hole(s).
Re: Tiny Hole in CPAP Hose - Only six weeks old
I have been using My original Hose for over two years now with no Problems but I store the Hose along with the Machine Daily as i have small G'Children here--Not worth the risk of Damage as Kids will be Kids and their curosity gets the best of them. I have never washed my Hose and only rinsed it out a couple times in over two years. I have several new Hoses stored just in case though. Could be that your Hose was Flawed from the Manufacturer--Hard to say but never the less a Hole is a Hole.
_________________
Mask: Mirage Liberty™ Full Face CPAP Mask with Nasal Pillows With Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Re: Tiny Hole in CPAP Hose - Only six weeks old
Hoses are an easy repair. Duct tape works like a charm. Takes longer to find the tape than to make the repair. It will last a long time also.
JeffH
JeffH
- SuperGeeky
- Posts: 414
- Joined: Mon May 19, 2008 3:55 pm
- Location: MaryEtta, Gawga (Now a part of the Union)
Re: Tiny Hole in CPAP Hose - Only six weeks old
Don't wash the hose??
Re: Tiny Hole in CPAP Hose - Only six weeks old
When you say duct tape are you speaking about the aluminium tape or the one exterior gray, sticky side white with a kind of threads inside?
I have never seen gorilla tape here only gorila glue and that's not good.
Thanks
I have never seen gorilla tape here only gorila glue and that's not good.
Thanks
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Headrest not modified, Hose Lift System, SleepyHead software. Pressure settings 7 cmH2O constant. |
- bearded_two
- Posts: 459
- Joined: Mon Aug 10, 2009 8:01 pm
Re: Tiny Hole in CPAP Hose - Only six weeks old
Gorilla glue is great stuff, and a drop of it might fix a small hole in a hose; urethane glues stick to about anything, including many plastics. Urethane glues are very sticky when wet, expand as they cure, and can be impossible to remove after it cures. Alcohol can usually remove it before it cures, you may have to sand it off or file it down after it cures.
- OldLincoln
- Posts: 780
- Joined: Wed Mar 26, 2008 7:01 pm
- Location: West Coast
Re: Tiny Hole in CPAP Hose - Only six weeks old
The latest hose I received on my regular DME delivery is quite unlikely to get a leak as it is very thick. That the good, but the bad is it is so stiff it stands up when I turn over rather than flex and swivel. I called and they will replace when I get around to it. I'm back to my old one now. Moral is take thin and flexible and patch as needed.
ResMed AirSense 10 AutoSet / F&P Simplex / DME: VA
It's going to be okay in the end; if it's not okay, it's not the end.
It's going to be okay in the end; if it's not okay, it's not the end.