HumidAire 4i disassembly for cleaning?

Post by wcj1 » Wed Aug 05, 2009 11:33 am

I have been doing therapy for 6 months. I always use distilled water in my humidfier, except when I travel (about 24 days out of the past 6 months).

Since it has been 6 months, I am eligible for a new mask and the first night I tried it out, I decided to increase the humidity level. Previously my humidifier was only set at 1.5 and I did not think about the consequences of a higher level. When I went to bed last night, I discovered that the tank was dry from the night before and I had some brown stains in it. Luckily, I had just picked up a replacement water chamber too, so I could swap it out and not worry about it right then.

I have not tried to clean this yet, but I think it would be easier if I took the HumidAire apart. Is it really designed to be taken apart? I cannot tell and I do not want to break it in trying. And, if it can be taken apart, is it recommended? (i.e. will it still work ok later, or might I have problems getting it back together and sealed properly?)

Thanks,
BillJ

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: HumidAire 4i disassembly for cleaning?

Post by jdm2857 » Wed Aug 05, 2009 11:44 am

Do not attept to open the standard H4i chamber. It is held together by plastic pins the end of which are melted to form rivet-like ends. Look at the bottom to see. You can't open it without breaking it.

ResMed makes a more expensive chamber for the H4i that opens and is dishwasher safe.

Or just use some vinegar in the dirty tank. Vinegar will dissolve any mineral deposits.
jeff

wcj1
Posts: 21
Joined: Sun Jan 11, 2009 11:19 pm

Re: HumidAire 4i disassembly for cleaning?

Post by wcj1 » Wed Aug 05, 2009 6:59 pm

Thanks. I thought they looked like they might break, so I wanted to check before I tried anything. I will try the vinegar.
BillJ