Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Motirck
Posts: 10
Joined: Fri Oct 17, 2025 11:53 am

Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by Motirck » Fri Oct 17, 2025 5:52 pm

Hi guys! Do you know if the algorithm of CPAP Resmed AirSense 10 AutoSet is trustworthy? I mean, I have a prescription from my doctor with min and max pressure, but I already had to change this pressure some times because I'm still adapting as I'm using it for 20 days now. Some people say that it works very good, other say that in AutoSet (without setting pressures) it can use an excessive pressure that is unnecessary. I'm using that for 20 days and my reports say that min pressure I achieved was 8.9 and max was 10.7. Right now it os configured min pressure as 9 and max as 11. Should I keep oit set or should I trust on automatic set?
Thank you!

mlmollenkamp
Posts: 55
Joined: Fri Apr 01, 2016 3:46 pm

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by mlmollenkamp » Fri Oct 17, 2025 7:17 pm

For most people most of the time the automatic algorithms are better than fixed pressure. There are exceptions, but for most people the automatic pressure settings work well. There are reasons to fine tune both the minimum and maximum pressure, but within a reasonable range the machine tries to use the minimum pressure that's necessary.

I suggest you get either OSCAR or SleepHQ. Once you have actual data its relatively straight forward to determine if its necessary to change minimum or maximum. There are some other settings as well, EPR and response for example. Get some data and there are people here who can help you interpret it.

_________________
Machine: Aircurve 11 asv
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: P30i is modified with Dreamwear headgear.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by Pugsy » Fri Oct 17, 2025 7:33 pm

I prefer auto adjusting pressures myself.
I learned a long time ago that my pressure needs during REM Stage sleep (where my OSA is much, much worse) were a LOT higher than what I would need during non REM sleep.
If I were to try the fixed pressure route then I might need 15 to 18 cm all night long....compare that to using maybe 8 or 9 the other parts of the night. Which would you want to do???
REM is maybe 20 % of the night.

The auto adjusting algorithm actually won't go higher unless it thinks it has a need to....now sometimes that reason isn't the greatest reason and when that happens fixed pressures might be a better option. Flow limitations come to my mind because sometimes nasal congestion can cause the machine to think it needs to go killing FLS with more pressure but nasal congestion FLs won't respond to more pressure. That's really about the only time that the machine will increase the pressure and get it wrong.

In general the machine's auto algorithm works quite well despite what you may have heard at other websites by people who don't think the same way.
Everyone has the right to their opinion and you are hearing a different opinion.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

ozij
Posts: 10474
Joined: Fri Mar 18, 2005 11:52 pm

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by ozij » Fri Oct 17, 2025 9:11 pm

Pugsy wrote:
Fri Oct 17, 2025 7:33 pm
Everyone has the right to their opinion and you are hearing a different opinion.
Pugsy wrote:
Fri Oct 17, 2025 7:33 pm

And everyone has an obligation to find out what's best for them. Listen to the different opinions, experiment, track your results log how you feel, and go by the info you have about yourself.

You wouldn't set an insulin pump based on general statements, and the same is true of your CPAP.
Some people's sleep is interrupted by changes in pressure - they do better on fixed.
Some people's flow limitation arennot resolved by the higher pressure - the do better on fixed.
The type of setting you choose depends on what results you get from it. In this therapy, you have only yourself to take care of - find out what works for you.

Posted by ozij, who is still wrongfully banned by the so called security program, still using vpn, still ignored by cpaptalk's tech support (assuming such an antity exists...)

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Motirck
Posts: 10
Joined: Fri Oct 17, 2025 11:53 am

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by Motirck » Sat Oct 18, 2025 6:53 am

Thanks guys! I'll test what will work for me... I do appreciate any response :)

User avatar
Dog Slobber
Posts: 4260
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by Dog Slobber » Sat Oct 18, 2025 9:49 am

A properly configured APAP configuration *IS* superior to CPAP.

Jason, LankyLefty27 has done APAP a real disservice by comparing flawed comparisons.

He compares APAP to his properly configured CPAP configuration. The thing is: his APAP minimum pressure configuration is always 2cm lower than his CPAP configuration. So of course APAP will perform poorly when the minimum configuration is too low. This comparison is disingenuous/dishonest.

Here's a thought exercise.

Let's say your optimum CPAP configuration is straight 10 cm.

So turn on APAP, set your minimum pressure to 10, and then set your maximum to 11.

In this circumstance, how possibly could APAP be performing worse than the CPAP? It can't. It will be delivering at least the pressure that the CPAP was, and should the algorithm determine an increase is necessary, it will increase it.

And no, don't interpret this as a hit piece agsainst Jason. I like Jason, love his videos, enthusiasm, humour and general knowledge. He's just wrong on his condemnation of APAP, which is always a comparison of a misconfigured device.,
Battery Backup: EcoFlow Delta 2

User avatar
Motirck
Posts: 10
Joined: Fri Oct 17, 2025 11:53 am

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by Motirck » Sat Oct 18, 2025 12:06 pm

Dog Slobber wrote:
Sat Oct 18, 2025 9:49 am
A properly configured APAP configuration *IS* superior to CPAP.

Jason, LankyLefty27 has done APAP a real disservice by comparing flawed comparisons.

He compares APAP to his properly configured CPAP configuration. The thing is: his APAP minimum pressure configuration is always 2cm lower than his CPAP configuration. So of course APAP will perform poorly when the minimum configuration is too low. This comparison is disingenuous/dishonest.

Here's a thought exercise.

Let's say your optimum CPAP configuration is straight 10 cm.

So turn on APAP, set your minimum pressure to 10, and then set your maximum to 11.

In this circumstance, how possibly could APAP be performing worse than the CPAP? It can't. It will be delivering at least the pressure that the CPAP was, and should the algorithm determine an increase is necessary, it will increase it.

And no, don't interpret this as a hit piece agsainst Jason. I like Jason, love his videos, enthusiasm, humour and general knowledge. He's just wrong on his condemnation of APAP, which is always a comparison of a misconfigured device.,
Thanks for the explanation. I also think APAP is better than CPAP as it can vary automatically inside the min and max pressure configured. However, my most important question is: Should I let APAP configured from 4cmH2O to 20cmH2O and trust it will only deliver what I really need? Or it is better I set min and max more closely?
As for reference, I'm using APAP for 20 days and I never had a pressure less than 8.9 and more than 10.7cm. So I had set now it as min 9 and max 11. Is it a good approach or you recommend let 4-20cmH2O?

Thanks a lot!

User avatar
Dog Slobber
Posts: 4260
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by Dog Slobber » Sat Oct 18, 2025 1:15 pm

Motirck wrote:
Sat Oct 18, 2025 12:06 pm
Should I let APAP configured from 4cmH2O to 20cmH2O and trust it will only deliver what I really need?
Absolutely not.
4-20 is the very definition of a horribly misconfigured APAP.
Battery Backup: EcoFlow Delta 2

User avatar
ChicagoGranny
Posts: 15369
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by ChicagoGranny » Mon Oct 20, 2025 1:03 pm

Dog Slobber wrote:
Sat Oct 18, 2025 1:15 pm
4-20 is the very definition of a horribly misconfigured APAP.
I still smile fondly every time this comes up. My former sleep doctor, who was very good otherwise, would get visibly aggravated when I told her I would not use 4 - 20. I had done enough experimenting to know that I needed a minimum of 10.

The only other point we clashed on was foam cervical collars. She did not approve! After she left, my first visit was a year later with her assistant. I mentioned wearing a collar. She said, "Yes, that's good thing to do." :lol:
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

DreamDolphin
Posts: 6
Joined: Wed Mar 13, 2024 5:45 am

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by DreamDolphin » Mon Oct 27, 2025 3:33 am

Since no two nights are exactly alike, I prefer APAP over fixed CPAP. Also, I require a slightly higher pressure later in the night and APAP takes care of that. I currently keep the range to only 1cm, some people need a bit more.

Yes, setting to a range of 4-20cm and relying on the algorithm to automatically deliver your optimal pressure is definitely not the way to go. You will experience massive pressure swings throughout the night, introducing its own set of problems. These algorithms assume you're properly titrated around a certain pressure and are meant to provide a small boost during obstruction. They were not designed to self-titrate algorithmically. Maybe someday we'll see machines that can do that effectively, and that would be welcome innovation, but there are none in existence at the moment.

_________________
Machine: Luna II Auto CPAP Machine with Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: I also use Resvent iBreeze & Luna/BMC travel PAP machines.

User avatar
ChicagoGranny
Posts: 15369
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Resmed AirSense 10 AutoSet - Is automatic setup better than fixed pressure?

Post by ChicagoGranny » Tue Oct 28, 2025 1:25 pm

DreamDolphin wrote:
Mon Oct 27, 2025 3:33 am
are meant to provide a small boost during obstruction
No pressure increases during obstruction - see chart. Sorry to be so precise.


Image
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.