respironics machine needs repair
i have m series auto c flex, with 1100 hours on it, it stopped working during the night, the alert alarm awoke me,and the screen said service required, i have power to machine but ithats all, message keeps appearing, i will send it back to cpap, per shipping inst,they were kind to me. has any one had this same problem? ps. my machine is still under warranty.thanks a bunch! two stepper!
_________________
Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear |
Additional Comments: swift1,swift2,hybrid, quattro full face,ultra mirage full face |
Re: respironics machine needs repair
It happens occasionally. That's why it's a good idea to have a spare/backup machine handy.
Den
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
-
- Posts: 34
- Joined: Fri Mar 23, 2007 8:55 pm
- Location: florida
Re: respironics machine needs repair
wulfman. thank you for response, yes i do have a backup, it is a resmed s7 not an auto by far,but it is almost new, but it was set at titration of 10 at sleep study. from what i can see this cpap does not have an option to change pressure that i know of, this machine \was sold to me by dme who charged medicare $3500for it, i did not belong to this forum then and did not realize what was going on, after about 1 year i got my sleep study report and read that they had indicated a bipap machine, i called the dme and he said i could not do anything about it now, he said they gave him the wrong report at first, that wasnt mine. then he got corrected, but too late to exchange it, that is when i purchased the m series from cpap, and really liked it, , so till i find out what may be needed to fix it i will go back to resmed, thanks again,two stepper.
_________________
Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear |
Additional Comments: swift1,swift2,hybrid, quattro full face,ultra mirage full face |
Re: respironics machine needs repair
Any machine can have the pressure reset within the range it will run at! The S7 won't operate as an auto or a Bi Pap for that matter. I think this is the instructions to reset the Pressure on the S7:
1) Turn on your machine and wait for the "Welcome "message to complete
2) Press the right button (with the red line on it - it should also have the word EXIT
above it on the display)
3) Now the display should show "S7 Lightweight " or "S7 Elite" or S7 Spirit" and the
word MENU should show over the left button."
4) Press both the right button and the down-arrow button at the same time and hold
down for at least 3 seconds.
5) "CLINICAL MENU - WELCOME" should show in the display, then after 3 seconds the
display will change again to ...
6) "CLINICAL - MENU exit"
7) Press the left button with the word MENU above it
Next you should see "SETTINGS - ENTER exit" (if SETTINGS doesn't show it may be
"RESULTS" or "OPTIONS" or "SERVICING" if it is any of these other words then use
the middle up/down toggle button to navigate to "SETTINGS"
9) When you have "SETTINGS" in the display, press the left button with the word
"ENTER" above it (this now takes you into the SETTINGS menu)
10) Then use the up/down toggle button in the middle of the panel, to navigate to the
settings you want to change - when you have the setting you want to change press the
left button with the word "change" above it.
11) Use the up/down toggle button to vary the settings.
To get out just keep pressing the right button. It will have either "cancel" or "exit"
written above it. Keep pressing it until back to normal ("S7 Elite" or S7 Lightweight" or "S7 Spirit").
Jerry
1) Turn on your machine and wait for the "Welcome "message to complete
2) Press the right button (with the red line on it - it should also have the word EXIT
above it on the display)
3) Now the display should show "S7 Lightweight " or "S7 Elite" or S7 Spirit" and the
word MENU should show over the left button."
4) Press both the right button and the down-arrow button at the same time and hold
down for at least 3 seconds.
5) "CLINICAL MENU - WELCOME" should show in the display, then after 3 seconds the
display will change again to ...
6) "CLINICAL - MENU exit"
7) Press the left button with the word MENU above it
Next you should see "SETTINGS - ENTER exit" (if SETTINGS doesn't show it may be
"RESULTS" or "OPTIONS" or "SERVICING" if it is any of these other words then use
the middle up/down toggle button to navigate to "SETTINGS"
9) When you have "SETTINGS" in the display, press the left button with the word
"ENTER" above it (this now takes you into the SETTINGS menu)
10) Then use the up/down toggle button in the middle of the panel, to navigate to the
settings you want to change - when you have the setting you want to change press the
left button with the word "change" above it.
11) Use the up/down toggle button to vary the settings.
To get out just keep pressing the right button. It will have either "cancel" or "exit"
written above it. Keep pressing it until back to normal ("S7 Elite" or S7 Lightweight" or "S7 Spirit").
Jerry
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto |
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS
I still play Cowboys and Bad Guys but now I use real bullets. CAS
-
- Posts: 34
- Joined: Fri Mar 23, 2007 8:55 pm
- Location: florida
Re: respironics machine needs repair
thank you so much!i got it set to correct pressure, no problem at all, you made it so easy for me ,i will be using it till my auto is repaired, !two stepper
_________________
Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear |
Additional Comments: swift1,swift2,hybrid, quattro full face,ultra mirage full face |
Re: respironics machine needs repair
I have an error code E - 87 can you advise. (maching is respironics Model no. 1017453 - REMaster Auto c/C-Flex) thanks rs :
Re: respironics machine needs repair
E 87richard smeltzer wrote:I have an error code E - 87 can you advise. (maching is respironics Model no. 1017453 - REMaster Auto c/C-Flex) thanks rs :
Occluded, restrictive or defective flow sensor.
Needs to be replaced. Or replace PCA.
http://www.acbio.com/ this place fixes cpap machines.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
Last edited by Pugsy on Tue Jul 31, 2012 1:53 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: respironics machine needs repair
My suggestion:richard smeltzer wrote:I have an error code E - 87 can you advise. (maching is respironics Model no. 1017453 - REMaster Auto c/C-Flex) thanks rs :
Go up to the Search function.......click on the "Advanced Search" and look up the forum member "Accurate".
Or, go to the "acbio (dot) com" website. (also listed in the forum member's signature)
Contact them and they should be able to help you.
.
Re: respironics machine needs repair
I've S7 elite. When i turn it on, screen shows welcome message and start to scan with ">>>>>>" signs increasing on screen. Then in the middle it says "system error call service"
I don't have service facility in my area. Its too far. Can anybody help me trace fault and fix it.
I don't have service facility in my area. Its too far. Can anybody help me trace fault and fix it.
Re: respironics machine needs repair
How old is your machine? Do you know the number of hours on it?
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: respironics machine needs repair
For most people self repair isn't a option. The DME you got it from could send it in for repair, but it's a 4 generation old machine. Replacement is probably a better option. The locations listed in the other posts can probably repair it but it will cost for them to see and probably not a good investment.shary wrote:I've S7 elite. When i turn it on, screen shows welcome message and start to scan with ">>>>>>" signs increasing on screen. Then in the middle it says "system error call service"
I don't have service facility in my area. Its too far. Can anybody help me trace fault and fix it.
You can also join the forum, and keep informed we have some honest sellers of used equipment, but remember buyer beware. Also Second Wind, sells used units. I bought a new APAP from our host cpap dot com., you won't find a more honest dealer. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: respironics machine needs repair
Displays error message:shary wrote:I've S7 elite. When i turn it on, screen shows welcome message and start to scan with ">>>>>>" signs increasing on screen. Then in the middle it says "system error call service"
I don't have service facility in my area. Its too far. Can anybody help me trace fault and fix it.
SYSTEM ERROR
Call service!
Component failure. Return your unit for servicing
That's all the manual says.
You need to get a newer/different one. The S7 generation of machines was made prior to (about) 2006.
That's why we have spares and backup machines. Nothing lasts forever.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: respironics machine needs repair
Ahem! One might note the dates of the original post. 2008!
(It's happened to me, too, sometimes they reappear like ghosts).
(It's happened to me, too, sometimes they reappear like ghosts).
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: respironics machine needs repair
Yes....."original" posts are old.......but, "Shary" posted today.Janknitz wrote:Ahem! One might note the dates of the original post. 2008!
(It's happened to me, too, sometimes they reappear like ghosts).
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: respironics machine needs repair
That's how we know it's the weekend, people drag up old posts and Whack-O-Dootle them. It was only 8 years old, wrong brand, must have caught me in a good mood, (almost next go impossible) jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire