Has anybody done a DIY remote on/off switch for resmed AirSense 10?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
feeling_better
Posts: 801
Joined: Sun Jun 01, 2008 7:08 pm

Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by feeling_better » Fri Oct 23, 2020 5:36 pm

[An old member, now not active, returning for help]

My Resmed AirSense 10 used to auto start when I put on the mask, and auto stop when I remove it. For some reason it does not auto stop anymore. I only see one setting for both on and off, which is enabled.

I cannot reach my machine from where I sleep, so it would be very nice to have this switch. I am hoping somebody has done a wired push switch in parallel with the built in top button to do this function. Any references? Thank you for your help!
Resmed S9 Elite cpap mode, H5i Humidifier, Swift FX Bella L nasal pillows

User avatar
Pugsy
Posts: 64159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by Pugsy » Fri Oct 23, 2020 5:39 pm

Will it auto off if you just unhook the mask hose from the long hose?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
chunkyfrog
Posts: 34449
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by chunkyfrog » Fri Oct 23, 2020 6:15 pm

Pugsy wrote:
Fri Oct 23, 2020 5:39 pm
Will it auto off if you just unhook the mask hose from the long hose?
That works for me; but it's easier to just reach over and press the button.
I would assume something has changed--like switching masks.
A perfectly normal thing with some nasal pillows.
It happens 100% of the time for me--two machines, over 5+ years.
Better to have it not switch off when you'd like, than to keep shutting off when you don't.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 64159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by Pugsy » Fri Oct 23, 2020 6:38 pm

chunkyfrog wrote:
Fri Oct 23, 2020 6:15 pm
That works for me; but it's easier to just reach over and press the button.
OP specifically said the on/off button on the machine wasn't easily reached from the bed.

With the P10 my machine wouldn't auto off for me but it does for the Bleep. Same pressures.
OP may have switched masks...showing the Swift FX in the profile but maybe the P10 now.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
chunkyfrog
Posts: 34449
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by chunkyfrog » Fri Oct 23, 2020 7:33 pm

Pugsy wrote:
Fri Oct 23, 2020 6:38 pm
. . .
wasn't easily reached from the bed.
. . .
Scoot the nightstand closer?
:mrgreen:

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 64159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by Pugsy » Fri Oct 23, 2020 7:45 pm

chunkyfrog wrote:
Fri Oct 23, 2020 7:33 pm
Pugsy wrote:
Fri Oct 23, 2020 6:38 pm
. . .
wasn't easily reached from the bed.
. . .
Scoot the nightstand closer?
:mrgreen:
:lol: :lol: Might be like me and I couldn't squeeze a gnat's ass between my night table and my bed and the wall and maybe where the machine is down off the top of the table and in a drawer or something...just out of reach.
I couldn't get my table a quarter of an inch closer if my life depended on it. I put a king size bed in a very small bedroom. If I moved the bed it would be in the doorway.
Or maybe shoulders just don't want to allow reaching anywhere.

Unhooking the mask hose from the long hose should work for auto off to kick in though and once up OP auto off has kicked in then just hook it back up again. If it doesn't work then there's a problem with the machine. It needs to be tested anyway to make sure there isn't a problem with the machine sensors.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
chunkyfrog
Posts: 34449
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by chunkyfrog » Fri Oct 23, 2020 9:37 pm

An excellent place for a quick-connect fitting.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by DreamStalker » Sat Oct 24, 2020 7:20 am

What level of DIY?

Image


or

Image
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
chunkyfrog
Posts: 34449
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by chunkyfrog » Sat Oct 24, 2020 7:29 am

Make Alexa do it.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by DreamStalker » Sat Oct 24, 2020 7:47 am

chunkyfrog wrote:
Sat Oct 24, 2020 7:29 am
Make Alexa do it.
Isn't Alexa a Russian spy?

At least the Clapper is an American spy -- even if bought and paid for the Chinese Communist Party.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
Pugsy
Posts: 64159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by Pugsy » Sat Oct 24, 2020 8:05 am

Please get back on topic folks. This side tracking is not helping the OP.
Further OT chit chat will be removed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by khauser » Sat Oct 24, 2020 11:07 am

Not sure if the idea of using a clapper-type switch was a joke, but if anything like that IS done, the very useful (IMO) feature of the 30 minute cooldown through the hose is forfeit. Personally I would not want to defeat that.

_________________
MachineMask
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by Goofproof » Sat Oct 24, 2020 11:30 am

khauser wrote:
Sat Oct 24, 2020 11:07 am
Not sure if the idea of using a clapper-type switch was a joke, but if anything like that IS done, the very useful (IMO) feature of the 30 minute cooldown through the hose is forfeit. Personally I would not want to defeat that.
While a Clapper would turn a XPAP off, it would NOT turn it on. You would need to buy a Clapper finger extension to press the On button. Also on stormy nights thunder might turn off the XPAP.

You might just buy Lee Press On Nails, in the extra long size (2 foot), to push the Buttons. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Dog Slobber
Posts: 3976
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by Dog Slobber » Sat Oct 24, 2020 11:51 am

Turning the device's power on remotely won't help, because that won't start the blower. I don't see a practical, permanent DIY solution.

Perhaps an extension grabber.

How old is the device, time for a new one?

Image

User avatar
Pugsy
Posts: 64159
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Has anybody done a DIY remote on/off switch for resmed AirSense 10?

Post by Pugsy » Sat Oct 24, 2020 12:01 pm

Dog Slobber wrote:
Sat Oct 24, 2020 11:51 am
How old is the device, time for a new one?
Even a brand new device won't necessarily fix the auto off problem if it is because of pressure and the nasal pillow used.
This is talked about in the ResMed manual it is so common.
Some nasal pillows just offer enough resistance the machine doesn't know the human is no longer attached. Primarily a problem at lower pressures but when I experimented with trying to find the right pressure...I quit at 10 cm because it still wouldn't auto off with the P10 pillow on. Now maybe it might have worked if I had used a larger nasal pillow but I didn't have one to try and I wasn't going to use a larger pillow than I wanted or needed just to get auto off to work.

With the Bleep I no longer have this issue though.. Auto off works just fine with it.

The easiest way to fix this problem (assuming not sensor failing issue) is just unhook the mask hose from the long hose and then hook it back up once auto off has kicked in and the machine is in cool down mode...when someone can't reach the on/off button for whatever reason. If someone can't easily disengage the mask hose from the long hose then one of those quick release adapters could be used.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.