Hi, My 420 E keeps developing leaks where the water container fits into the machine.
It doesn't leak all the time, so it has caught me out a few times and has caused havoc with my readings. I blamed my taping until I realized what the cause was, the sneaky little thing just waits for me to forget to check it for leaks. I am still using the original soft fitting connector which hasn't been used a lot as I only use the humidifier in the winter.
The same fitting on my spare Resmed machine has a metal ring in it, which is probably better.
I just wondered if anyone else has had this type of leaking problem?
PB 420 E leaks
Re: PB 420 E leaks
there should be a round clear silicone coupler between machine and the tank. Push down on the tank, slide it back from the machine, check the coupler, insert your finger and pull it out a bit, makes sure machine is seated on the platform, slide the tank back into the machine until it stops.
Make sure the tiny clear hose is connected to the little barb next to the hose, that clear hose goes down the center of the larger hose through a hole in the coupler on the end of the hose, that has to be connected to the machine.
Make sure the tiny clear hose is connected to the little barb next to the hose, that clear hose goes down the center of the larger hose through a hole in the coupler on the end of the hose, that has to be connected to the machine.
someday science will catch up to what I'm saying...
Re: PB 420 E leaks
If you haven't used the coupling a lot, could it be that you're using it backwards? It has a shallow side and a deep side. The deep side is the machine side, the shallow side should be towards the humidifier.
![Image](https://www.cpap.com/viewImage.php?PNum=1873&Image=humid-coupler.jpg)
O.
![Image](https://www.cpap.com/viewImage.php?PNum=1873&Image=humid-coupler.jpg)
O.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
- Posts: 225
- Joined: Tue Apr 18, 2006 12:08 pm
Re: PB 420 E leaks
If you put a small tiewrap around the collar to hold it on to the humidifier tank.
This will stop the leaks.
Billy
This will stop the leaks.
Billy