donnaclark@gci.net wrote:rested gal wrote:What is the full name of your machine, as written on top of the machine?
If it is not an M series machine and doesn't have a blue sticker on top that says "C-Flex", it doesn't have C-flex.
Sounds like it is probably the older Respironics REMstar Auto (and I'm guessing that's the "full name" you'll see on top) before they added C-Flex to their autos. If so, I have a Provider manual for your older machine and will be glad to email that to you. Please PM me your email address. Use the "PM Me" button at the bottom of my post here.
I have the same unit, will you email me a copy too? Please.
Try these
Den
REMstar Auto w/C-Flex Setup
When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.
To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Auto w/C-Flex beeps twice. (about 5 seconds)
Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
(If you press the Pressure start/stop button, you will exit the Setup Menu.)
A. Compliance hours/nights: (recommend leaving alone, but CAN be cleared at this point)
Select next setting with >
B. Therapy mode: (CPAP/CFLE/APAP/AFLE) select with ^ or v
Select next setting with >
B1. If CPAP or CFLE select pressure setting: Select with ^ or v
Select next setting with >
B1a. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
B2. If APAP or AFLE, select minimum pressure setting: Select with ^ or v
Select next setting with >
B2a. If APAP or AFLE, select maximum pressure setting: Select with ^ or v
Select next setting with >
B2b. C-Flex mode (if you chose AFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >
C: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0)
(if using APAP or AFLE mode, Ramp Length Setting MUST be set to 0:00)
Select next setting with >
D. Ramp pressure setting: Select with ^ or v (use only in CPAP or CFLE mode)
Select next setting with >
E. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >
F. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)
_________________
| Mask | Humidifier | |
 |  | |
| Additional Comments: Either lead, follow or get out of the way..... |
(4) REMstar Autos w/C-Flex & (6) Pro 2 CPAPs w/C-Flex - Pressure Setting=12 cm - C-Flex Setting=2