Autostart not shutting off with Climatelineair

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
nospamprl
Posts: 9
Joined: Sat Mar 09, 2024 1:34 pm

Autostart not shutting off with Climatelineair

Post by nospamprl » Thu Apr 11, 2024 12:25 am

Just bought my Airsense 10 and noticed that the autostart feature doesnt shut down the machine when the mask is removed when the Climateairline is used.

It turns on automatically, but only shuts off if the standard tube is used.

Is this the expected behaviour?

Thanks
Airsense 10 Autoset w/ ClimateAirLine
Started therapy April 2024

User avatar
Conrad
Posts: 510
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Autostart not shutting off with Climatelineair

Post by Conrad » Thu Apr 11, 2024 6:10 am

I believe that there's an auto-start and an auto-off, do you have them both enabled?
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

User avatar
Dog Slobber
Posts: 3966
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Autostart not shutting off with Climatelineair

Post by Dog Slobber » Fri Apr 12, 2024 9:47 am

LillianFelly wrote:
Thu Apr 11, 2024 4:39 am
There are a couple of possibilities:

Leak Sensor: The Climateairline might be causing a small leak that the machine misinterprets as breathing, preventing auto-off.
Settings: Some settings might need adjustment for optimal autostart functionality with the Climateairline.
This is not how AutoStop works, in fact, any leaking would make it more likely that AutoStop shuts down the machine.

Autostop doesn't detect breathing it shuts down the machine when it detects excessive leaking, and attributes that to nobody being attached to the mask.
Conrad wrote:
Thu Apr 11, 2024 6:10 am
I believe that there's an auto-start and an auto-off, do you have them both enabled?
Not on the AS10.

On the AS11 SmartStart and SmartStop are two separate configuration items. On the AS10 they are ether both on or both off.
nospamprl wrote:
Thu Apr 11, 2024 12:25 am
It turns on automatically, but only shuts off if the standard tube is used.

Is this the expected behaviour?
No, and it's not tied to the climateline hose.

Sometimes SmartStop isn't reliable because their is too much resistance because of the mask. Enough for the leak threshold not to be sufficiently high enough to interpret as no longer being attached to the person. Small nasal pillows and low pressure contribute to this.