Hi everyone!
Trying to understand how stuff works... The device is supposed to keep a constant pressure (adjusting through the night but more or less constant during a breath) BUT we inhale, exhale, so is it true that it needs to adjust, otherwise the pressure would go up at every exhale?
Not sure what makes sense, but I do hear the motor working harder to keep up during inhales.
The second question, the airsense 10 seems to stop the pressure at the end of each exhale. In the sense that I turn it on, before breathing, pressure is there (can be felt and measured). Inhale, pressure is there. Exhale, pressure is there. Stop the exhale slowly, pressure drops to zero for a second or so before returning. If I try to inhale during that break I would have no pressure at all. EPR is turned off.
Is there a way to remove that break? It feels like a lag/delay between my breath and the device reacting.
Maybe some settings?
Resmed airsense 10 takes a break at end of breath
Re: Resmed airsense 10 takes a break at end of breath
Impossible. The machine can't EVER drop the pressure below 4 cm. Simply won't/can't do it.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- Respirator99
- Posts: 343
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: Resmed airsense 10 takes a break at end of breath
G'day BlackBags, welcome to the forum.
It sounds like the EPR (expiratory pressure relief) isn't properly synching with your breathing. If you have EPR turned on, it reduces pressure slightly as you exhale. It then increases it again as you inhale. The available settings are 0 (off), 1, 2 or 3 - where the number corresponds to the number of cm reduction. As Pugsy said, the machine won't go below 4 cm, even if EPR is set to the maximum.
To overcome your problem, try turning EPR off. You'll find it takes a little more effort to exhale, but you may eliminate that synchronisation problem.
It sounds like the EPR (expiratory pressure relief) isn't properly synching with your breathing. If you have EPR turned on, it reduces pressure slightly as you exhale. It then increases it again as you inhale. The available settings are 0 (off), 1, 2 or 3 - where the number corresponds to the number of cm reduction. As Pugsy said, the machine won't go below 4 cm, even if EPR is set to the maximum.
To overcome your problem, try turning EPR off. You'll find it takes a little more effort to exhale, but you may eliminate that synchronisation problem.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: Resmed airsense 10 takes a break at end of breath
Thank you for your replies Pugsy and Respirator99!
Very strange... I did buy a calibrated manometer (pressure sensor) - mostly to check if the motor was still working properly, as the DME would have done (they switched from patient care and device maintenance to remote compliance checks only, so trying to learn how to do it myself, but that's a different story).
That sensor is measuring 0 pressure (in the mask) during that short time.
Will check EPR again.
Very strange... I did buy a calibrated manometer (pressure sensor) - mostly to check if the motor was still working properly, as the DME would have done (they switched from patient care and device maintenance to remote compliance checks only, so trying to learn how to do it myself, but that's a different story).
That sensor is measuring 0 pressure (in the mask) during that short time.
Will check EPR again.