UARS and CPAP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Sun Feb 17, 2019 3:41 am

Here's last night with minepap 6:
Bildschirmfoto 2019-02-17 um 10.22.57.png
Bildschirmfoto 2019-02-17 um 10.22.57.png (247.1 KiB) Viewed 12148 times
palerider wrote:
Sat Feb 16, 2019 1:06 am
Pressure goes up because of snores, flow limitations, hypopneas and apneas.
There's pressure increase in the first hour, but sleepyhead doesn't show any events and almost no flow limitations. I have read (here, third post viewtopic/t83387/What-caused-this-press ... rease.html) that flow limitations don't necessarily need to be tagged. I assume this is what's happening here?

Again, thank you!!

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: UARS and CPAP

Post by palerider » Sun Feb 17, 2019 3:52 am

I have no idea why you've got that 1 hour 1cm pressure drift.

Could you confirm all the settings on your machine? (I see what sleepyhead says, but it's buggy in that settings respect.

Also, if you want to zip up your entire SD card contents, I'd be happy to take a look at the data more closely. If you want to do that, upload it to: https://www.dropbox.com/request/0LalWvEsijRIgX8sdM7H

The fact that it's the exact same thing, on two charts, with an OA at the peak of the drift makes me think it's a glitch in the way sleepyhead is displaying the data.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Sun Feb 17, 2019 4:18 am

So the settings are right, except Climate Control which is on auto.

I uploaded it. It's the same pattern on 3 of the last 4 days. Almost all OA are falsely flagged, I think.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: UARS and CPAP

Post by palerider » Sun Feb 17, 2019 1:29 pm

werther wrote:
Sun Feb 17, 2019 4:18 am
So the settings are right, except Climate Control which is on auto.

I uploaded it. It's the same pattern on 3 of the last 4 days. Almost all OA are falsely flagged, I think.
It's a sleepyhead bug, if you look at the mask pressure line, you'll see the pressure is flat till the first OA. The last two of which are legit.

Can I pass your data over to the new people working on the sleepyhead replacement software, OSCAR?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Sun Feb 17, 2019 4:30 pm

palerider wrote:
Sun Feb 17, 2019 1:29 pm
It's a sleepyhead bug, if you look at the mask pressure line, you'll see the pressure is flat till the first OA. The last two of which are legit.
Damn, and I was trying to figure out what's wrong with the curves for an hour :lol:

palerider wrote:
Sun Feb 17, 2019 1:29 pm


Can I pass your data over to the new people working on the sleepyhead replacement software, OSCAR?
Sure!

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: UARS and CPAP

Post by palerider » Sun Feb 17, 2019 5:07 pm

werther wrote:
Sun Feb 17, 2019 4:30 pm
palerider wrote:
Sun Feb 17, 2019 1:29 pm
Can I pass your data over to the new people working on the sleepyhead replacement software, OSCAR?
Sure!
Thanks. It's contributions of data like that which help make the software better and more accurate.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Wed Feb 27, 2019 12:48 am

About two weeks have gone by since my first post: I have used the machine every night, although I am averaging only about three hours. I think I have solved my mouth leak/chipmunk issue now though, but it's probably too early to tell. The goal now is wearing the mask the whole night of course.

I am still having a lot of these kinds of events:


Bildschirmfoto 2019-02-27 um 07.46.03.png
Bildschirmfoto 2019-02-27 um 07.46.03.png (246.39 KiB) Viewed 12038 times
More pressure?

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: UARS and CPAP

Post by palerider » Wed Feb 27, 2019 12:55 am

More pressure is unlikely to affect those kind of breathing patterns.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Wed Feb 27, 2019 1:00 am

Do you have an explanation for them? Should I ignore them?

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: UARS and CPAP

Post by palerider » Wed Feb 27, 2019 3:07 am

werther wrote:
Wed Feb 27, 2019 1:00 am
Do you have an explanation for them?
Nope, things just aren't perfect. Might be your co2 levels got a touch low, temporarily depressing your respiratory drive, till the level built backup and you started to want to breathe again... but, that's just a guess.
werther wrote:
Wed Feb 27, 2019 1:00 am
Should I ignore them?
I would.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Wed Feb 27, 2019 3:16 pm

Ok, thank you palerider!

Hope I'm able to post some full nights/data (and how I feel...) in the next weeks.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Sat Mar 09, 2019 4:36 am

Short update. I was not able to keep the mask on longer than about 3 hours with EPAP 6/PS 4. Also tried out EPAP 8/PS 2. Didn't work either. Also had some weird feeling in my chest as well as aerophagia.
Six nights ago I changed it to Epap 5/PS 4. Was able to keep the mask on longer, averaging about 5 hours. Don't have that weird feeling my chest anymore.

Palerider, I assume you want me to get to EPAP 6? If so, I would increase by 0.2 every few nights, if thats ok.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto
Attachments
Bildschirmfoto 2019-03-09 um 11.34.53.png
Bildschirmfoto 2019-03-09 um 11.34.53.png (248.56 KiB) Viewed 11964 times
Bildschirmfoto 2019-03-09 um 11.34.48.png
Bildschirmfoto 2019-03-09 um 11.34.48.png (252.92 KiB) Viewed 11964 times
Bildschirmfoto 2019-03-09 um 11.35.00.png
Bildschirmfoto 2019-03-09 um 11.35.00.png (249.05 KiB) Viewed 11964 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: UARS and CPAP

Post by palerider » Sat Mar 09, 2019 5:59 am

No harm in sneaking up on it... proceed at your own pace.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Sun Mar 10, 2019 3:48 am

Ok, again, thank you!!!

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto

werther
Posts: 42
Joined: Mon Dec 31, 2018 6:52 am
Location: Germany

Re: UARS and CPAP

Post by werther » Fri Mar 29, 2019 4:30 am

I now can tolerate fixed EPAP 6 with PS 3. No aerophagia and no "weird feeling in my chest". I thought the reason why I would pull the mask off midnight was the pressure increases, but I still don't keep the mask on longer than about 5 hours. Maybe I should get sleeping pills. Had some good nights though, which is giving me hope.
Another thing which might be an issue is that even though I use tape, sometimes my mouth "puffs". No leak, just the air pressuring against the tape. I'm not sure how often this happens (would you be able to tell from the flow rate curve??).
Here's yesterday night and a close up of two events. How do you interpret them? Do I need higher pressure?

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 VAuto
Attachments
Bildschirmfoto 2019-03-29 um 11.18.39.png
Bildschirmfoto 2019-03-29 um 11.18.39.png (245.15 KiB) Viewed 11835 times
Bildschirmfoto 2019-03-29 um 11.19.15.png
Bildschirmfoto 2019-03-29 um 11.19.15.png (36.04 KiB) Viewed 11835 times
Bildschirmfoto 2019-03-29 um 11.22.10.png
Bildschirmfoto 2019-03-29 um 11.22.10.png (33.19 KiB) Viewed 11835 times