Airsense 10 - Knocking/Ticking Noise When Breathing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jagstyles
Posts: 26
Joined: Tue Jun 05, 2018 8:16 pm

Airsense 10 - Knocking/Ticking Noise When Breathing

Post by jagstyles » Wed Sep 19, 2018 8:45 am

Lately, I've started to hear a knocking sound on each breath I take with the machine. It sounds like it's coming from the water chamber and I've tried removing the water chamber and re-inserting thinking perhaps it wasn't snugly in place but it was. There was still water in the chamber so not sure what might be causing this. It kept me up for most of the night until I just got fed up and turned off the machine. Anyone experience this before?

_________________
MachineMask

User avatar
LSAT
Posts: 13232
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by LSAT » Wed Sep 19, 2018 9:14 am

Check for water/moisture in the hose...."rainout".

jagstyles
Posts: 26
Joined: Tue Jun 05, 2018 8:16 pm

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by jagstyles » Wed Sep 19, 2018 10:19 am

LSAT wrote:
Wed Sep 19, 2018 9:14 am
Check for water/moisture in the hose...."rainout".
Thanks will check that. If this is the issue, do I decrease the temperature setting or the humidity level?

_________________
MachineMask

User avatar
Pugsy
Posts: 64044
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by Pugsy » Wed Sep 19, 2018 10:45 am

For unwanted condensation in the mask or hose you have to either INCREASE the hose air temp to prevent the hose air from cooling to the point of releasing the moisture or maybe reduce the humidity setting.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

jagstyles
Posts: 26
Joined: Tue Jun 05, 2018 8:16 pm

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by jagstyles » Wed Sep 19, 2018 12:18 pm

Didn't see any condensation or moisture in the tube. my current temperate setting is 28.0 degrees Celsius and humidity level is only set to 2.0. Wonder if the knocking noise is something else?

_________________
MachineMask

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by Jas_williams » Thu Sep 20, 2018 1:12 am

I had a clunking noise when breathing due to the humidifier not sealing correctly, I pulled of the seals reseated and problem went away

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
Nick Danger
Posts: 621
Joined: Mon Jun 09, 2014 3:13 pm

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by Nick Danger » Thu Sep 20, 2018 7:49 am

My Airsense 10 humidifier frequently makes that knocking noise. Reseating the gasket works for a few weeks before the noise comes back.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: APAP mode, minimum pressure = 9. No ramp, EPR = 3, medium. Soft cervical collar. Sleepyhead software.

kuk_sleeper
Posts: 54
Joined: Sat Jan 13, 2018 11:06 am

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by kuk_sleeper » Fri Sep 21, 2018 8:30 am

Does it sound like this https://streamable.com/5rt8s? I had this sound for a week and then the machine died.
CPAP: Resmed Airsense 10
Mask: ResMed Airfit F20

SewTired
Posts: 1737
Joined: Thu Apr 16, 2015 8:33 am
Location: Minneapolis area

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by SewTired » Sat Sep 22, 2018 3:59 pm

Jas_williams wrote:
Thu Sep 20, 2018 1:12 am
I had a clunking noise when breathing due to the humidifier not sealing correctly, I pulled of the seals reseated and problem went away
I did the same thing. That doesn't mean that there aren't other causes, but it is a frequent cause.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)

Dave-o
Posts: 1
Joined: Sat Mar 16, 2024 7:20 am

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by Dave-o » Sat Mar 16, 2024 7:32 am

I get tube knocking after about 6 hrs. of use lately. I can stop it by adding more water to tank. Of course this means disconnecting hose and getting up to get water. Anyone use this technique to stop their knocking sound? Other suggested methods?

User avatar
Pugsy
Posts: 64044
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 10 - Knocking/Ticking Noise When Breathing

Post by Pugsy » Sat Mar 16, 2024 7:34 am

Zombie thread alert folks. Old thread. Watch your responses.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.