No, I don't. If you understood more, you wouldn't think that though.
Dreamstation 4-5 hours a night - Asking for Help
Re: Dreamstation 4-5 hours a night - Asking for Help
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Dreamstation 4-5 hours a night - Asking for Help
Thanks for sharing your wisdom with those who don't know as much as you. I guess next time someone asks for help and they have centrals I can just say ramp up the pressure to knock out all types of apneas and point back to your post here.
10 to 15 cm h20
Re: Dreamstation 4-5 hours a night - Asking for Help
The next time someone asks for help on any subject here, you should say *NOTHING*.USMCVet wrote: ↑Sun May 20, 2018 1:22 pmThanks for sharing your wisdom with those who don't know as much as you. I guess next time someone asks for help and they have centrals I can just say ramp up the pressure to knock out all types of apneas and point back to your post here.
You have very little understanding about the nuances of treatment, you have no clue about the big picture, yet you're very argumentative.
That's a very bad combination for 'helping' anybody at all.
I'm sure you're very accomplished at some things in your life. Stick to those.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Dreamstation 4-5 hours a night - Asking for Help
Just argumentative with the bullies. I asked you to elaborate so I could genuinely learn something. Do you not agree asking questions is important to learn?palerider wrote: ↑Sun May 20, 2018 1:28 pmThe next time someone asks for help on any subject here, you should say *NOTHING*.USMCVet wrote: ↑Sun May 20, 2018 1:22 pmThanks for sharing your wisdom with those who don't know as much as you. I guess next time someone asks for help and they have centrals I can just say ramp up the pressure to knock out all types of apneas and point back to your post here.
You have very little understanding about the nuances of treatment, you have no clue about the big picture, yet you're very argumentative.
That's a very bad combination for 'helping' anybody at all.
I'm sure you're very accomplished at some things in your life. Stick to those.
If someone is coming here for help and getting advice from strangers it's important to ask questions. If someone isn't willing to explain their reasoning then how can people vet the information they receive so they can make a informed decision?
If you don't like me that's fine. But you really should answer all relevant questions for those who may read a topic later.
10 to 15 cm h20
Re: Dreamstation 4-5 hours a night - Asking for Help
Anybody that trots out the lame "bullies" attempt at insult really REALLY needs to look in the mirror, long and hard.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Dreamstation 4-5 hours a night - Asking for Help
Honestly I'm not even sure what you're trying to say here.
noun
plural noun: trots
1.
a trotting pace.
"our horses slowed to a trot"
2.
informal
diarrhea.
"a bad case of the trots"
10 to 15 cm h20
Re: Dreamstation 4-5 hours a night - Asking for Help
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Dreamstation 4-5 hours a night - Asking for Help
I'm 100% serious. I asked my g/f as well to make sure I wasn't just being an idiot. I'm starting to understand the meaning of your avatar.
10 to 15 cm h20
Re: Dreamstation 4-5 hours a night - Asking for Help
There's this thing called "google".
Perhaps you're really not nearly as smart as you think you are.Definition of trot out.
transitive verb. 1 : to lead out and show the paces of (a horse) 2 : to bring forward for display or use. trotted out a new excuse.
Trot Out | Definition of Trot Out by Merriam-Webster
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Dreamstation 4-5 hours a night - Asking for Help
Are you serious? I did Google. I posted definition I found for key word. I asked another human being.palerider wrote: ↑Sun May 20, 2018 7:41 pmThere's this thing called "google".
Perhaps you're really not nearly as smart as you think you are.Definition of trot out.
transitive verb. 1 : to lead out and show the paces of (a horse) 2 : to bring forward for display or use. trotted out a new excuse.
Trot Out | Definition of Trot Out by Merriam-Webster
I don't claim to be smart. I claim to want to learn. Teach me.
10 to 15 cm h20
Re: Dreamstation 4-5 hours a night - Asking for Help
What more do you need than what's above?USMCVet wrote: ↑Sun May 20, 2018 7:49 pmAre you serious? I did Google. I posted definition I found for key word. I asked another human being.palerider wrote: ↑Sun May 20, 2018 7:41 pm
There's this thing called "google".
Perhaps you're really not nearly as smart as you think you are.Definition of 'trot out'.
transitive verb. 1 : to lead out and show the paces of (a horse) 2 : to bring forward for display or use. trotted out a new excuse.
Trot Out | Definition of Trot Out by Merriam-Webster
I don't claim to be smart. I claim to want to learn. Teach me.
Sometimes the failure to adequately google is not googling specifically enough, sometimes, as in this case, it's a failure to adequately broaden the search.
Definition 2 above is my exact usage.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Dreamstation 4-5 hours a night - Asking for Help
Moderators - Please step in and stop this contest.
Re: Dreamstation 4-5 hours a night - Asking for Help
You specifically said "Anybody that trots out the lame "bullies" attempt at insult really REALLY needs to look in the mirror, long and hard."palerider wrote: ↑Sun May 20, 2018 8:04 pmWhat more do you need than what's above?USMCVet wrote: ↑Sun May 20, 2018 7:49 pmAre you serious? I did Google. I posted definition I found for key word. I asked another human being.palerider wrote: ↑Sun May 20, 2018 7:41 pm
There's this thing called "google".
Perhaps you're really not nearly as smart as you think you are.Definition of 'trot out'.
transitive verb. 1 : to lead out and show the paces of (a horse) 2 : to bring forward for display or use. trotted out a new excuse.
Trot Out | Definition of Trot Out by Merriam-Webster
I don't claim to be smart. I claim to want to learn. Teach me.
Sometimes the failure to adequately google is not googling specifically enough, sometimes, as in this case, it's a failure to adequately broaden the search.
Definition 2 above is my exact usage.
As such I googled trots which made absolutely no sense.
So what your saying if I correctly understand you is that i am the bully?
Pot calling the kettle black?
I would hate to think the advice you give here can be as confusing especially when you won't explain your rationale! Yikes!!!
Maybe if I have to guess and broaden my search as to your meaning you're not being clear enough?
10 to 15 cm h20
Re: Dreamstation 4-5 hours a night - Asking for Help
Page 1 of this thread contains useful information.
Page 2....not particularly useful.
If you guys want to continue hijacking this thread then I will remove the bickering posts and place those posts in a thread devoted to the bickering so you guys can really duke it out in an OT thread.
So...please just stop with the bickering....please.
If you can't say something on topic pertaining to OP's question....how about just not saying anything at all?
Now if you just want to fight...I will fix up a place so you guys can fight all you want to and it won't disrupt a newbie helping thread.
You all know how I feel about disrupting/hijacking newbie helping threads.
Page 2....not particularly useful.
If you guys want to continue hijacking this thread then I will remove the bickering posts and place those posts in a thread devoted to the bickering so you guys can really duke it out in an OT thread.
So...please just stop with the bickering....please.
If you can't say something on topic pertaining to OP's question....how about just not saying anything at all?
Now if you just want to fight...I will fix up a place so you guys can fight all you want to and it won't disrupt a newbie helping thread.
You all know how I feel about disrupting/hijacking newbie helping threads.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Dreamstation 4-5 hours a night - Asking for Help
Well said Pugsy. This forum would be well served if the bullying, sarcasm, and mean spirited comments were reduced. Thanks for your leadership.