Camping with an ASV cpap
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: Camping with an ASV cpap
I am thinking it would work....
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
Re: Camping with an ASV cpap
Your husband is *WRONG*. the Resmed 12V converter works with... 12v.Barb (Seattle) wrote: ↑Tue May 08, 2018 8:41 pmmy husband said it won't work (the converter) as we have a 12v plug. and, that we need a 12v cpap. *shrug* it's a 24v machine and we have a 12 v plug I guess is what he means.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 1739
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Camping with an ASV cpap
The ResMed converter will work. I have the S10 version running from a 12 volt battery and it is very efficient.Barb (Seattle) wrote: ↑Tue May 08, 2018 8:41 pmmy husband said it won't work (the converter) as we have a 12v plug. and, that we need a 12v cpap. *shrug* it's a 24v machine and we have a 12 v plug I guess is what he means.
I thought it would convert it TO 12 v...thought that's what the plug does?
this is confusing.
See my earlier posting with the link.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: Camping with an ASV cpap
good deal then I kind of thought that it made sense that using a 12v cord with a 24v cpap would work....it converts. that's what it's supposed to do.
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
-
- Posts: 1739
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Camping with an ASV cpap
The converter has a 12V lighter plug like you could use in a car. I think it also comes with a socket & clips to adapt to any 12 volt source.Barb (Seattle) wrote: ↑Wed May 09, 2018 5:13 amgood deal then I kind of thought that it made sense that using a 12v cord with a 24v cpap would work....it converts. that's what it's supposed to do.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Camping with an ASV cpap
Barb as the others have said get the Resmed DC adapter it will allow you to plug it into your 12v supply buy your bed and provide 24v for the S9 Adapt. But if your insurance is going to get you a new ASV they may provide the Aircurve 10 ASV which uses a different DC adapter I don't want you wasting your money on an adapter you will only use once
Jason
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: Camping with an ASV cpap
true....I want to get some type of battery back up pack for when our lights go out....they're quite expensive thoughJas_williams wrote: ↑Wed May 09, 2018 1:49 pmBarb as the others have said get the Resmed DC adapter it will allow you to plug it into your 12v supply buy your bed and provide 24v for the S9 Adapt. But if your insurance is going to get you a new ASV they may provide the Aircurve 10 ASV which uses a different DC adapter I don't want you wasting your money on an adapter you will only use once
Jason
unless we create a 12v plug for our house somehow...has anyone ever done that
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
Re: Camping with an ASV cpap
I know you don't want to hear it but there is another option.
Your DreamStation at a pressure of around 15 with no exhale relief and no pressure support.
Your DreamStation at a pressure of around 15 with no exhale relief and no pressure support.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Camping with an ASV cpap
You said several times that a battery backup is quite expensive - but a battery plus charger is between $80 and $200 depending on size. Is that too expensive? If by "12v plug for our house" you mean a battery on a charger with a socket, then yes, that's exactly what I have next to my bed.Barb (Seattle) wrote: ↑Wed May 09, 2018 2:25 pmtrue....I want to get some type of battery back up pack for when our lights go out....they're quite expensive though
unless we create a 12v plug for our house somehow...has anyone ever done that
The possibility that you might get an upgrade doesn't have to paralyze you. If its only for one trip you can use a small inverter (you should have one anyways on the RV). Just understand you'll use almost twice as much power as with the converter. Would you be keeping the S9? Maybe you'd use that for travel.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Quattro™ Air Full Face Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid |
Hark, how hard he fetches breath . . . Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Re: Camping with an ASV cpap
Indeed... I've got one of the s9 12v converters... they work perfectly.. and they give you much longer battery use than an inverter will. (up to 2x as much time)Barb (Seattle) wrote: ↑Wed May 09, 2018 5:13 amgood deal then I kind of thought that it made sense that using a 12v cord with a 24v cpap would work....it converts. that's what it's supposed to do.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: Camping with an ASV cpap
the battery pack I looked at at first was $499...there is a cheaper alternative called the Freedom which would work with the 12v charger for my resmed.CapnLoki wrote: ↑Wed May 09, 2018 3:16 pmYou said several times that a battery backup is quite expensive - but a battery plus charger is between $80 and $200 depending on size. Is that too expensive? If by "12v plug for our house" you mean a battery on a charger with a socket, then yes, that's exactly what I have next to my bed.Barb (Seattle) wrote: ↑Wed May 09, 2018 2:25 pmtrue....I want to get some type of battery back up pack for when our lights go out....they're quite expensive though
unless we create a 12v plug for our house somehow...has anyone ever done that
The possibility that you might get an upgrade doesn't have to paralyze you. If its only for one trip you can use a small inverter (you should have one anyways on the RV). Just understand you'll use almost twice as much power as with the converter. Would you be keeping the S9? Maybe you'd use that for travel.
and it was $209. then, with the charger, another $65-75 or so...I think it might be a viable alternative....I have to save some $$$ though. if the power goes out, I could just go to the motorhome and plug it in
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: Camping with an ASV cpap
nooooooooooooooooooooooooooooooooooo
*runs from the room*
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
Re: Camping with an ASV cpap
No, dear, the cheaper alternative is a 60$ scooter battery, which will get you four or five nights without the humidifier. plus a charger.Barb (Seattle) wrote: ↑Wed May 09, 2018 8:43 pmthere is a cheaper alternative called the Freedom which would work with the 12v charger for my resmed.
and it was $209. then, with the charger, another $65-75 or so...I think it might be a viable alternative....I have to save some $$$ though. if the power goes out, I could just go to the motorhome and plug it in
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Camping with an ASV cpap
These packs you're looking at are very expensive because they are using Lithium cells to save weight. This is very handy if you need portability, but it can drive the cost up by as much as a factor of ten! For home backup (and for RV/Boat/car camping) you can get far more power for less money with an AGM battery. For example, a scooter battery (as use by millions of seniors and safe for indoor use) that has 35 amp-hours capacity can be had for $65. A charger plus a few minor cables adds $35-$50. Assembly is "plug and play" - the most complicated bit is bolting the color coded cables to the battery. A lithium based pack with this much power would be about $500. AGM batteries come in a variety of sizes so you can get anything from 4 to 100 amp-hours, price and weight varies appropriately. These batteries are used in everything from UPS power supplies, jump starters, scooters and wheelchairs, to lawn mowers, etc. Unlike tradition lead acid batteries, they are sealed and can be used upside down, and don't vent any gas. With a gentle charger, they are perfectly safe at home. The major downside is the weight, so a while an AGM battery that can power your pump with humidity through a 2-3 day power outage may be affordable, you won't be moving it much!Barb (Seattle) wrote: ↑Wed May 09, 2018 8:43 pmthe battery pack I looked at at first was $499...there is a cheaper alternative called the Freedom which would work with the 12v charger for my resmed.
and it was $209. then, with the charger, another $65-75 or so...I think it might be a viable alternative....I have to save some $$$ though. if the power goes out, I could just go to the motorhome and plug it in
For more details, look at the thread linked to in my signature. A post near the bottom of the first page lists the details of my setup.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: Quattro™ Air Full Face Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid |
Hark, how hard he fetches breath . . . Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html
-
- Posts: 1739
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Camping with an ASV cpap
I use an AGM battery as a backup but, because I want it for power outages, I bought a 100 watt solar panel system to charge it.
I can get several days (no heat or humidity) from the AGM battery without recharging, if needed. I live in a rural area so I do not want to depend solely on the power grid for charging.
I can get several days (no heat or humidity) from the AGM battery without recharging, if needed. I live in a rural area so I do not want to depend solely on the power grid for charging.