Water yuckies
Water yuckies
I have a question. I have had my AirSense™ 10 for several months now and am having a problem. Wondering if any body else has had this problem and what the solution would be or could possibly be. When I wake up in the morning, I go to pour out my remaining water from the reservoir and notice there is "stuff" in it. It's kind of gross. I have cleaned and dried my humidifier daily, I am purchasing distilled water from Walmart. Everything looks clean when I pour it in. But in the morning there is "stuff" Any ideas what's happening? Tks. for any suggestions.
Re: Water yuckies
Two questions: 1) do you ever *wash* the reservoir? And 2) do you have a filter on the blower unit?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Backup/travel unit is an identical S9 AutoSet for Her w/Eson nasal mask |
-
- Posts: 1739
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Water yuckies
Welcome to CPAPTalk!Hel8951 wrote: ↑Sat Apr 14, 2018 5:10 pmI have a question. I have had my AirSense™ 10 for several months now and am having a problem. Wondering if any body else has had this problem and what the solution would be or could possibly be. When I wake up in the morning, I go to pour out my remaining water from the reservoir and notice there is "stuff" in it. It's kind of gross. I have cleaned and dried my humidifier daily, I am purchasing distilled water from Walmart. Everything looks clean when I pour it in. But in the morning there is "stuff" Any ideas what's happening? Tks. for any suggestions.
I have the Airsense 10 AutoSet and use the Walmart distilled water with the magenta colored label & cap in my humidifier.
I usually just top it up for the next night but I occasionally empty and put in fresh. I do not have "stuff" in my humidifier.
I would suspect whatever you are using to clean it. I have even sometimes used bottled water when traveling with no issues.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Water yuckies
Do you dry the tank with a lint-free cloth?
You could be seeing the previous day's towel lint.
Cleaning in the evening, and leaving it wet helps prevent lint.
You could be seeing the previous day's towel lint.
Cleaning in the evening, and leaving it wet helps prevent lint.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Water yuckies
I use the same water and do not have any such deposits. Perhaps it's because it's a new machine.
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Water yuckies
No.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Water yuckies
No...has nothing to do with it being a new machine nor the source of the water.
It's more likely something in the air in the bedroom getting sucked through the filter (assuming on is in place and if there isn't one that might be why the crud is in the water) and landing in the water.
Unless the machine itself is faulty/damaged and spitting out stuff from the motor that is getting into the water chamber and that's highly unlikely.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Water yuckies
Can always count on D.H. to add a few misguided "Yuckies" on his own.
I used to read the newspapers sunday comics, back in the 50's when they were worth reading.
Now I don't, they have probably called them "Yuckies", I do! Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Water yuckies
I have same machine and have used the same water and I'm with pretty much everyone else. I just fill my tank every night.
I too would start at the filter and then the tank. I have only dumped my tank a few times, the last time friday before packing it up before a trip. When I have dumped and rinsed it I just let it air dry.
I too would start at the filter and then the tank. I have only dumped my tank a few times, the last time friday before packing it up before a trip. When I have dumped and rinsed it I just let it air dry.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Simplus Full Face CPAP Mask with Headgear |
-
- Posts: 142
- Joined: Thu Feb 01, 2018 2:53 am
- Location: British Columbia
Re: Water yuckies
I have no idea as to the veracity of her claim, but my respiratory therapist told me to avoid WalMart's distilled water...too many of her customers have reported problems with it. FWIW, I have been using tap water since purchasing my ResMed machine, now two full months into it's life, and I have no residues, precipitation, icky stuff, fibers, or discoloration when I empty my reservoir every third day or so. I have the lowest humidity setting, so my volume sits in the reservoir for up to three days...no growth, nothing visible when I empty what's left.Hel8951 wrote: ↑Sat Apr 14, 2018 5:10 pmI have a question. I have had my AirSense™ 10 for several months now and am having a problem. Wondering if any body else has had this problem and what the solution would be or could possibly be. When I wake up in the morning, I go to pour out my remaining water from the reservoir and notice there is "stuff" in it. It's kind of gross. I have cleaned and dried my humidifier daily, I am purchasing distilled water from Walmart. Everything looks clean when I pour it in. But in the morning there is "stuff" Any ideas what's happening? Tks. for any suggestions.
The idea is to never let the water completely dry out in the reservoir. Always have a few tablespoons of volume left to retain the salts...IF there are salts. If you have staining, try some CLR or vinegar, or lemon juice (fresh). Even apple juice will work in a pinch.
BTW, our local water comes from a glacial catch basin, and it's a steep one with high volumes, so little opportunity for the waters supplied locally to have gathered much mineralization. That's why I have no hesitation about using it as it emerges from our taps. Those in other areas, especially the prairies, would be advised to at least mix in some distilled water if they come close to a dry reservoir each morning.
-
- Posts: 1739
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Water yuckies
You forgot about the pipes that can add deposits. Remember the pipes in Flint, Michigan?mesenteria wrote: ↑Wed Apr 18, 2018 9:53 am
BTW, our local water comes from a glacial catch basin, and it's a steep one with high volumes, so little opportunity for the waters supplied locally to have gathered much mineralization. That's why I have no hesitation about using it as it emerges from our taps. Those in other areas, especially the prairies, would be advised to at least mix in some distilled water if they come close to a dry reservoir each morning.
I know of no issue with Walmart distilled water.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Water yuckies
Walmart distilled water is usually OK.
I bought some at a ghetto grocery that made pink goo in my tank-ONCE.
My home-made stuff is cleaner and cheaper,
but I wish I had the counter space back.
Still better than all that schlepping jugs around.
I bought some at a ghetto grocery that made pink goo in my tank-ONCE.
My home-made stuff is cleaner and cheaper,
but I wish I had the counter space back.
Still better than all that schlepping jugs around.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Water yuckies
The more you take that stuff apart and "clean" it (including drying), the more opportunity there is for contaminates to enter.Hel8951 wrote: ↑Sat Apr 14, 2018 5:10 pmI have a question. I have had my AirSense™ 10 for several months now and am having a problem. Wondering if any body else has had this problem and what the solution would be or could possibly be. When I wake up in the morning, I go to pour out my remaining water from the reservoir and notice there is "stuff" in it. It's kind of gross. I have cleaned and dried my humidifier daily, I am purchasing distilled water from Walmart. Everything looks clean when I pour it in. But in the morning there is "stuff" Any ideas what's happening? Tks. for any suggestions.
As was previously asked........do you have a clean air intake filter in place?
I've used distilled water from Walmart since I started therapy and have never had problems with it.
Try doing your cleaning routine in the evening (if you really think it needs to be done that often) and don't bother drying or hanging it up in unsanitary locations (like the bathroom). Drying is unnecessary and cleaning is often done too frequently. Distilled water shouldn't be an issue.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05