So...can't wait till Friday
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: So...can't wait till Friday
.
Last edited by Barb (Seattle) on Wed Apr 18, 2018 5:55 am, edited 1 time in total.
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: So...can't wait till Friday
Yes CA means central apnea or clear airway. But that is not what I said. I said the events look like OSA to me. That means obstructive sleep apnea. Have you tried IPAP above 20?
[/quote]
just did last night. horrible results once more.......
22/18
stupid stupid stupid I am beginning to think there's something wrong with my brain...it just keeps stopping no matter what settings I use.
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
Re: So...can't wait till Friday
I take it last night's experiment did go so great. What happened?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: So...can't wait till Friday
well, this didn't work. I had to try. I was curious. now, I'll try the settings you wanted last night tonight. One more day
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: So...can't wait till Friday
23 CA apneas. per hour.
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: So...can't wait till Friday
The Hypopneas are probably central related as well mine were and all resolved by my ASV I had similar patterns to you on a bad night (AHI of 30) now a nightly AHI of 0.1 I hope the machine Pugsy is loaning to you works as well.
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
Re: So...can't wait till Friday
Still curious about auto mode at the settings I suggested which apparently have not been tried before.
It might be a disaster..might not but I don't think any worse than last night.
Plan B still in the works.
Here's the deal...the Adapt/ASV will treat both any obstructive apneas and the centrals if/when they pop up.
If centrals reduce for some reason ...the Adapt won't need to do much in terms of centrals and it still does its job on the obstructive stuff.
Win win...IMHO. It works based on the workload it faces.
It might be a disaster..might not but I don't think any worse than last night.
Plan B still in the works.
Here's the deal...the Adapt/ASV will treat both any obstructive apneas and the centrals if/when they pop up.
If centrals reduce for some reason ...the Adapt won't need to do much in terms of centrals and it still does its job on the obstructive stuff.
Win win...IMHO. It works based on the workload it faces.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: So...can't wait till Friday
it does seem like win-win to me too.Pugsy wrote: ↑Wed Apr 18, 2018 6:38 amStill curious about auto mode at the settings I suggested which apparently have not been tried before.
It might be a disaster..might not but I don't think any worse than last night.
Plan B still in the works.
Here's the deal...the Adapt/ASV will treat both any obstructive apneas and the centrals if/when they pop up.
If centrals reduce for some reason ...the Adapt won't need to do much in terms of centrals and it still does its job on the obstructive stuff.
Win win...IMHO. It works based on the workload it faces.
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: So...can't wait till Friday
that's great to hear. hope it works that way for me too!Jas_williams wrote: ↑Wed Apr 18, 2018 6:35 amThe Hypopneas are probably central related as well mine were and all resolved by my ASV I had similar patterns to you on a bad night (AHI of 30) now a nightly AHI of 0.1 I hope the machine Pugsy is loaning to you works as well.
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
Re: So...can't wait till Friday
Have you ever tried *less* pressure support?
And have you had your heart checked out?
And have you had your heart checked out?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: So...can't wait till Friday
I wondered about this myself but I haven't seen any reports from before the bipap...as in the probable first DreamStation if there was a first non bilevel DreamStation after the M series.
I was going to suggest reducing PS after tonight's auto bilevel trial while the ASV is in transit.
Kinda wanted to get an idea where the bilevel wanted to go to first because if it wants to spend all night up at 18...going without much PS might not be much fun.
I was thinking maybe let PS run between 1 and 2 just to see if there was much difference...or off maybe.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- Barb (Seattle)
- Posts: 663
- Joined: Wed Aug 03, 2005 6:41 pm
Re: So...can't wait till Friday
pressure support?
I was on bipap...I'm not sure about the pressure support....I think it has to be on Auto for that?
I haven't had my heart checked out, but that did cross my mind. if the machine that's coming doesn't work, I'll go to my dr and tell
her what's going on.
USED to be on DreamStation..now on S9 VPAP ADAPT
21 years on CPAP
Mirage Quattro Air Small
21 years on CPAP
Mirage Quattro Air Small
Re: So...can't wait till Friday
Nope...you don't have to be on auto to be using Pressure Support.Barb (Seattle) wrote: ↑Wed Apr 18, 2018 12:36 pmI was on bipap...I'm not sure about the pressure support....I think it has to be on Auto for that?
Pressure Support is nothing more than the difference between inhale and exhale.
Example
16 EPAP and 20 IPAP...that's a difference of 4 cm...so PS of 4
14 EPAP and 19 IPAP...that's a difference of 5 cm...so PS of 5
now in auto mode with your machine the PS can either be fixed or vary some but it doesn't vary fast enough to do what the ASV can do when it varies and that's why it can't help your breathing that much.
The reason PR asked if you ever tried no PS...is because in some people it's the PS that triggers centrals. It's not hugely common though but I have seen it happen.
Had a lady that I was helping that had a truckload of centrals with PS of 4 and essentially none with PS of 3.
It's something that I was aware of and was going to discuss with you when we get a chance to chat.
Problem is when people need the PS for comfort and if that PS causes the centrals...damned if you do and damned if you don't.
But like I said...it's not all that common but something I was planning on ruling in or out in your situation just in case.
I first wanted auto bilevel mode at the settings I suggested...then depending on those results and your comfort and what pressures you appeared to be needing for the obstructive stuff I was going to suggest playing with the PS a bit.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- chunkyfrog
- Posts: 34397
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: So...can't wait till Friday
If it is your heart, you do NOT want to wait.
Call the office, and at least talk to a nurse or PA.
Call the office, and at least talk to a nurse or PA.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: So...can't wait till Friday
The last time I saw a person with a chart that looked like yours from the other day,Barb (Seattle) wrote: ↑Wed Apr 18, 2018 12:36 pmI haven't had my heart checked out, but that did cross my mind. if the machine that's coming doesn't work, I'll go to my dr and tell
her what's going on.
was from someone that turned out to have severe heart failure...
It was actually discovered from that very image... his doctor had been fobbing him off saying he thought he just had a hiatal hernia...
Now, I'm *certainly* not saying you do... but I'd hate for that to be the case, and you not know, till it was way too late.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.