Resmed filter change schedule confusion

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
davidr
Posts: 5
Joined: Thu Feb 19, 2009 9:39 am

Resmed filter change schedule confusion

Post by davidr » Tue Jan 09, 2018 8:10 pm

I just got a new ResMed Airsense 10. I'm trying to figure out how often to replace the filter.
The DME says every 2 weeks.
The ResMed quick start guide says every month.
The ResMed user manual says at least every 6 months.
That's a big range: anywhere from 2 weeks to 6 months.
I'm looking for any guidance anyone can offer about a good replacement schedule.

A related question: I found cheap filters on Amazon from Sunset Healthcare. Are those good? Some of the reviews say are are thinner than the ones that come with the machine. Or do you have better suggestions about where and what brand to buy?

Thanks,
-David

springman946
Posts: 156
Joined: Mon Sep 02, 2013 12:40 pm

Re: Resmed filter change schedule confusion

Post by springman946 » Tue Jan 09, 2018 9:29 pm

you should replace the filter every 6 months or when you start to see it turn black

_________________
MachineMaskHumidifier
Additional Comments: my machine is not listed so i going to list my machine manually I am using a Resmed Air Curve 10 ASV

User avatar
Okie bipap
Posts: 3553
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Resmed filter change schedule confusion

Post by Okie bipap » Tue Jan 09, 2018 10:23 pm

I check the filter once a month and change it when it starts to look dirty. During the high pollen season, I change it every two weeks because of my allergies.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed filter change schedule confusion

Post by palerider » Tue Jan 09, 2018 11:22 pm

davidr wrote:I just got a new ResMed Airsense 10. I'm trying to figure out how often to replace the filter.
The DME says every 2 weeks.
They want to sell you more and more shit.
davidr wrote:The ResMed quick start guide says every month.
Newbie fluff.
davidr wrote:The ResMed user manual says at least every 6 months or as needed.
The technical guide... trust this.
davidr wrote:That's a big range: anywhere from 2 weeks to 6 months.
I'm looking for any guidance anyone can offer about a good replacement schedule.
any other questions?
davidr wrote:A related question: I found cheap filters on Amazon from Sunset Healthcare. Are those good? Some of the reviews say are are thinner than the ones that come with the machine. Or do you have better suggestions about where and what brand to buy?

Thanks,
-David
I use whatever's cheaper from ebay or amazon, have for years.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: Resmed filter change schedule confusion

Post by Muse-Inc » Thu Jan 11, 2018 5:28 am

Change every month or more often during allergy season for me Feb-June. Then, Aug/Sept, Nov/Dec. I bought a LOT from sleep doc last visit 2011 and still have some.

_________________
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear
Humidifier: HumidAire H4i™ Heated Humidifier
Additional Comments: Started Aug '07 CPAP Escape at 11 then 9, S9 APAP range 9.8-16, now 10-17
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

Rob K
Posts: 642
Joined: Sun Feb 05, 2017 5:22 pm

Re: Resmed filter change schedule confusion

Post by Rob K » Thu Jan 11, 2018 7:51 pm

A wide range means change it when it's dirty. Could be 2 weeks to six months depending on hours of use and your environment. You'll have to decide when it looks dirty enough to change it. At home I just use the gray filter and wash it once a week. If I'm camping or sleeping somewhere dirty I'll put in a white filter and use that until it looks like it needs to be thrown away.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Additional Comments: Resmed F10 Mask for colds. When camping on battery power I use P10 mask and PR 560p machine.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed filter change schedule confusion

Post by palerider » Thu Jan 11, 2018 8:52 pm

Rob K wrote:A wide range means change it when it's dirty. Could be 2 weeks to six months depending on hours of use and your environment. You'll have to decide when it looks dirty enough to change it. At home I just use the gray filter and wash it once a week. If I'm camping or sleeping somewhere dirty I'll put in a white filter and use that until it looks like it needs to be thrown away.
One thing to note is that the resmed filters don't have to be changed as often as respironics, because they're something like 2.3 times the size... larger filter, longer time between changes.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

FlyingMoose
Posts: 129
Joined: Tue Jan 06, 2015 11:25 am

Re: Resmed filter change schedule confusion

Post by FlyingMoose » Thu Jan 11, 2018 9:35 pm

I use the hypoallergenic filters from cpap.com, I want to keep as much stuff out of my lungs and my machine as possible. It's not worth it to save a couple bucks a year and gum up an almost $900 machine (my dad had one fail most likely due to baby powder).

Also if you use a P10 nasal pillow mask, it will last longer if you have a good filter, so that the mesh doesn't get clogged. I've had one get ruined in a month in a dusty environment, even with the hypoallergenic filter. Again, it's not worth saving a few bucks on a filter and then having to buy a $100 mask.

If you hold the filter up to the light, you'll see little dots on it formed by the holes in the filter door. Also one small corner isn't really used so this is a good basis for comparison. If the light is being blocked, the air will be blocked also. Note that a filter actually improves over time as it gets more plugged up, but it becomes harder for the CPAP to pull air through it and will wear out the fan faster. So it's a balance, best judged by looking through the filter.

If you take it out to look, make sure you insert it back with the same side out, or all the stuff caught by the filter will then be backwashed into the machine.

I have a hepa air filter in my bedroom for allergies, and so the filter in the CPAP is still pretty much brand new after a whole year. But I've gone on vacation to a dusty area and had the filter be used up after 3 weeks. So it really depends on how clean the air is going in.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Autoset for Her 8.4-12.4

User avatar
chunkyfrog
Posts: 34394
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Resmed filter change schedule confusion

Post by chunkyfrog » Thu Jan 11, 2018 9:43 pm

The hypoallergenic filters fill up quicker.
Medicare will let me have two a month, but I can go three weeks
before the filter gets gray and grubby enough that I am OK with using a new one.
My allergies let me know when it is time for a new filter.
I live in a very old, dusty house; so YMMV.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Resmed filter change schedule confusion

Post by Goofproof » Thu Jan 11, 2018 11:08 pm

Just info: As filters catch more particles they become more efficient, removing smaller and smaller particles. This is a good thing, Until the filter gets too dirty and can no longer meet the flow requirements. Tonight mine gets changed. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Pugsy
Posts: 64023
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed filter change schedule confusion

Post by Pugsy » Fri Jan 12, 2018 8:55 am

In terms of ResMed filter quality...name brand vs aftermarket brands like Sunset.
I recently ordered an aftermarket hypoallergenic filter from Amazon seller (the same seller I was getting a replacement water chamber from when I dropped mine and broke it). I think what I got was Sunset brand but I would need to go dig them out to be sure.

There is a noticeable different in the density or thickness of the aftermarket brand when compared to a name brand ResMed hypoallergenic filter.
It takes 2 of the aftermarket filters to feel the same thickness in my fingers....I played with them...
One brand in one hand and the other brand in the other hand.

Now is this difference in density a critical issue? That I don't know the answer to. The less dense aftermarket filter is still denser than the mesh like filter that came with the machine.

If someone doesn't understand what I mean by "mesh like"...here's a picture that shows 3 filters.
One is the hypoallergenic filter on the left and the other 2 are 2 different brands of mesh filters.
ResMed routinely supplies the mesh like filter...so apparently they think it is sufficient for whatever filtration needs the machine might need.
And it is this filter that they are referring to in the manual when they say 6 months or sooner if it get too dirty looking.
Those other more dense filters like the white fine filter or the white fine hypoallergenic filters are filters that we as users might want to use for peace of mind. Peace of mind can be very important to some people. From a technical only standpoint ResMed seems to think that the mesh filters are sufficient...for my own personal preferences and peach of mind I prefer the fine filters and for me either the regular fine or hypoallegenic filters I am good with. After actually having the aftermarket fine filter in my hand and feeling the difference in density...it's a tough call but I get more peace of mind from the ResMed brand filter. Is it a critical need...probably not because even the cheap after market fine filter is denser than the ResMed supplied mesh filter...but I am using 2 of the aftermarket fine filters so as to give me about what I was getting with the ResMed brand hypoallergenic filter. It is probably over kill...but again for my own peace of mind it is just something I prefer.

I only got 4 of the cheap filters so won't take me long to get them dirty and end up in the trash....and after that I am going to stick with name brand hypoallergenic filters mainly for the peace of mind thing. Yes, they cost a little more and a bit harder to find but I think I have found a different source (I used to have a guy on EBay that I could get name brand filters from but he's gone) and probably will just buy in bulk from the source I found.
If anyone wants to go in with the cost on these name brand filters..send me a PM and we can talk about it. Near as I can tell about $2 a filter when in bulk of 50 package.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.