Plastic and Ball Pressure Checker

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
scrollinggrandpa
Posts: 19
Joined: Fri Dec 12, 2014 4:59 pm
Location: Hemet, California

Plastic and Ball Pressure Checker

Post by scrollinggrandpa » Tue Dec 12, 2017 3:07 pm

Hi all. I have a resmed S9 machine with water chamber. I used to use a Resperonics machine but it died and I ended up with the S9. I also have one of those plastic tube with steel ball pressure checkers. With the Resperonics machine if my pressure was set at 8 the checker registered 8 as well. Now, with the S9 machine my pressure is supposed to be 11 and if it is set at 11 in the machine the checker says it is at 8. In order to get 11 in the checker I have to set the machine at 14. I also am using a 10 foot standard hose. As a test I changed the hose to a 6 foot standard and there was no change in the checker readings, so there is no hose problem. I guess my question to all of you is: are these checkers accurate or should I throw it in the trash. I can't afford one of those expensive manometers so If any of you know of a good way to check the pressure I would like to hear from you.

Thanks, Merry Christmas and Happy New Year!

_________________
MaskHumidifier
Additional Comments: Using Sleepyhead Software

Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Re: Plastic and Ball Pressure Checker

Post by Woody » Tue Dec 12, 2017 3:39 pm

I built my own manameter using a 28" clear glass vase. After filling it to 24" I then stuck the
hose in it and kept lowering it till the air flow stopped and measured how far this was below
my fill mark. The 2 times I have used it the measurement matched what my 2 machines were
set to.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Plastic and Ball Pressure Checker

Post by Wulfman... » Tue Dec 12, 2017 3:50 pm

It's easy (and cheap) to check the way Woody did it.
That's the way I did it for a long time before I finally bought a gauge manometer.

viewtopic/t157764/search.php?keywords=h ... +manometer

viewtopic/t157764/viewtopic.php?f=1&t=4 ... ter#p37259


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Midnight Strangler
Posts: 248
Joined: Tue Oct 07, 2014 5:39 am

Re: Plastic and Ball Pressure Checker

Post by Midnight Strangler » Tue Dec 12, 2017 4:26 pm

scrollinggrandpa wrote:Now, with the S9 machine my pressure is supposed to be 11 and if it is set at 11 in the machine the checker says it is at 8. In order to get 11 in the checker I have to set the machine at 14.
I have five machines, some of them quite old. My thinking is why do I care what the actual pressure is? I just look at the AHI on the machine display*. If my AHI is low (<2.0), what do I care what the actual pressure is? The setting works!

BTW, is your S9 an Autoset?

Also, here is a curiosity question, who is telling you that you need 8 on one machine and 11 on another? How did they come up with those numbers?


(*Should anything on the display look funny, I will also look at the data in Sleepyhead.)

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Plastic and Ball Pressure Checker

Post by palerider » Tue Dec 12, 2017 5:25 pm

scrollinggrandpa wrote:Hi all. I have a resmed S9 machine with water chamber. I used to use a Resperonics machine but it died and I ended up with the S9. I also have one of those plastic tube with steel ball pressure checkers.
You've got a flow meter, not a "pressure checker".

so for your question:
scrollinggrandpa wrote:I guess my question to all of you is: are these checkers accurate or should I throw it in the trash.
For your purposes, NO, it's not accurate, since it doesn't measure pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13393
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Plastic and Ball Pressure Checker

Post by LSAT » Tue Dec 12, 2017 7:16 pm

If you are near a DME go in and ask (nicely) if they would check it

scrollinggrandpa
Posts: 19
Joined: Fri Dec 12, 2014 4:59 pm
Location: Hemet, California

Re: Plastic and Ball Pressure Checker

Post by scrollinggrandpa » Wed Dec 13, 2017 10:04 am

Thanks to all for your responses. Let me clearify a couple of things in my post: 1. I used the pressure of 8 as an example only. My old machine was set at 11 (Per sleep study) and the meter I have read 11. On my S9 the machine says 11 and the (now I know it is a flow meter) reads 8. My machine does not give me the AHI on the screen, I have to take out the SD card and put it into my computer and use the software to read it.

I wil find a vase like Woody had and give that a shot as well as check the data on the card. If my AHI is 2 or below and the machine says the pressure is 15 (per my sleep study two weeks ago) I will just leave it alone.

_________________
MaskHumidifier
Additional Comments: Using Sleepyhead Software

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Plastic and Ball Pressure Checker

Post by Goofproof » Wed Dec 13, 2017 10:29 am

scrollinggrandpa wrote:Thanks to all for your responses. Let me clearify a couple of things in my post: 1. I used the pressure of 8 as an example only. My old machine was set at 11 (Per sleep study) and the meter I have read 11. On my S9 the machine says 11 and the (now I know it is a flow meter) reads 8. My machine does not give me the AHI on the screen, I have to take out the SD card and put it into my computer and use the software to read it.

I wil find a vase like Woody had and give that a shot as well as check the data on the card. If my AHI is 2 or below and the machine says the pressure is 15 (per my sleep study two weeks ago) I will just leave it alone.
Flow on a XPAP means little, it's the pressure that holds the airway open, But the flow has to be able to supply the needed pressure and the flow to keep that pressure up.

The real tester would measure the pressure while pumping air at a set leak rate, bad news most testers are unable to do that. I use to know what sized hole was required to provide the leak, but the brain cells I stored that info in have died... Using the software can show you if your machine can keep up with needed flow. It's not really something that happens hardly ever, as long as the XPAP is working. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Plastic and Ball Pressure Checker

Post by chunkyfrog » Wed Dec 13, 2017 10:34 am

scrollinggrandpa wrote: . . . My machine does not give me the AHI on the screen, I have to take out the SD card and put it into my computer and use the software to read it. . . .
If your S9 gives AHI, you can set the display to show it.
I gave my old S9 away, so I don't recall the exact directions, but somebody will.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65257
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Plastic and Ball Pressure Checker

Post by Pugsy » Wed Dec 13, 2017 10:53 am

chunkyfrog wrote:If your S9 gives AHI, you can set the display to show it.
I gave my old S9 away, so I don't recall the exact directions, but somebody will
I don't remember exactly either but the manual explains how to do it.
http://www.apneuvereniging.nl/forum/pdf ... manual.pdf
Have to go into the clinical setup menu (manual also explains how to do that)...
Sleep Quality has to be set to "ON" and then the AHI and other data is available in the Sleep Report.
Somewhere around page 15 of the manual above.
The default from the factory is for Sleep Quality to be "OFF" and that is what prevents the data from being available on the LCD screen.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

scrollinggrandpa
Posts: 19
Joined: Fri Dec 12, 2014 4:59 pm
Location: Hemet, California

Re: Plastic and Ball Pressure Checker

Post by scrollinggrandpa » Wed Dec 13, 2017 10:59 am

Thanks again for all your information. I just pulled the SD card out of my machine and inported the data into Sleepyhead. For the past week, except for one night, my AHI has been .58 to 1.75
I downloaded the clinical manual and will check it our later today. Again, thanks to all for your help. May all of you have a Merry Christmas, Happy New Year and sleep well.

_________________
MaskHumidifier
Additional Comments: Using Sleepyhead Software

User avatar
ChicagoGranny
Posts: 15467
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Plastic and Ball Pressure Checker

Post by ChicagoGranny » Wed Dec 13, 2017 11:58 am

scrollinggrandpa wrote: ... AHI on the screen ...
Make sure sleep quality is turned on in clinical settings. To access clinical settings, hold down double check mark button and knob button simultaneously for 3 seconds.

Once sleep quality is turned on in clinical settings AHI will be displayed at the bottom when you press the "I" button. To get breakdown and see averages for time frames, go into clinical info - to get there you need to hold down "I" button and double check mark button simultaneously for three seconds. Once in, turn knob one click ccw to get to sleep report then press knob down to enter sleep report. Turn knob to scroll to the bottom. AHI breakdown is right under leak report. Click on "Period" and turn knob to change the time frame for displayed averages.
The data resets at noon. So if you look at the one-day report after noon, there will be no data. (Unless you took an afternoon nap with CPAP. )

scrollinggrandpa
Posts: 19
Joined: Fri Dec 12, 2014 4:59 pm
Location: Hemet, California

Re: Plastic and Ball Pressure Checker

Post by scrollinggrandpa » Wed Dec 13, 2017 3:02 pm

Thanks for the info ChicagoGranny! Your input will be very useful.

_________________
MaskHumidifier
Additional Comments: Using Sleepyhead Software

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: Plastic and Ball Pressure Checker

Post by grayghost4 » Wed Dec 13, 2017 3:40 pm

Quote : using a 28" clear glass vase.

you don't need anything that tall : 11cm. is less that 5 in.


https://www.google.com/search?q=11+cm+t ... 8&oe=utf-8

Image
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

scrollinggrandpa
Posts: 19
Joined: Fri Dec 12, 2014 4:59 pm
Location: Hemet, California

Re: Plastic and Ball Pressure Checker

Post by scrollinggrandpa » Thu Dec 14, 2017 11:01 am

Thanks Grayghost4 for the info and photo. This morning I looked on the screen showing the stastics and even though my machine is set for a pressure of 15 I only used 12.5 overnight and my AHI was 0.9. I am quite happy with these results.

_________________
MaskHumidifier
Additional Comments: Using Sleepyhead Software