disinfecting nasal pillows
disinfecting nasal pillows
I'm having a lot of trouble with recurring staph infections. Since staph germs tend to live in the nose, I'm concerned that I'm reinfecting myself through my nasal pillows. I bought a terrifying cleaning solution online but it doesn't sound like it's safe for pillows. Any suggestions?
Using a Phillips System One Apap with an Airfit P10 for her. Pressure 10-14. Ramp 4.
Re: disinfecting nasal pillows
I don't think most experienced cpapers work too hard to sterilize their masks. When I first got mine, I was instructed to soak in a 50/50 vinegar/water solution once a week. Now I just use soap and water on the actual pillows every day. Some folks wipe with baby wipes. If you have a particular issue with a recurrent bacterial infection, you could use something harsher, but it would need to be fully rinsed off after cleaning and the mask may degrade a bit sooner.
Jim
Jim
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Heated Humidifer, Non-Heated tube, Oscar |
Resmed Airsense 10 Autoset, Swift FX nasal pillows mask
Re: disinfecting nasal pillows
Use the cpap wipes every night . You can find them online or at the medical equipment store. Also suggest you wash your headgear every week. Cleanliness is very important. I also have the so-clean machine which handles the whole thing. You just put the entire mask and hose in the container and its cleaned by ozone.
I recommend it if you can afford it. Its about 300.00 on line. Other wise everything has to be done by hand.
I recommend it if you can afford it. Its about 300.00 on line. Other wise everything has to be done by hand.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Last edited by AMESS on Wed Dec 06, 2017 9:20 pm, edited 1 time in total.
Re: disinfecting nasal pillows
Non scented baby wipes from Walmart, CVS etc do the same thing as CPAP wipes at 1/3rd the price.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: disinfecting nasal pillows
I have no personal experience with this one.
https://www.cpap.com/productpage/Contro ... eaner.html
I think free ground shipping if you use promo code STOCKING17....I don't think it has expired yet.
I just use hot soapy water but if you are really wanting peace of mind about the potential staph issues you might consider this product.
https://www.cpap.com/productpage/Contro ... eaner.html
I think free ground shipping if you use promo code STOCKING17....I don't think it has expired yet.
I just use hot soapy water but if you are really wanting peace of mind about the potential staph issues you might consider this product.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: disinfecting nasal pillows
Is there anything that a 10% chlorine solution does not disinfect???shaina wrote:I'm having a lot of trouble with recurring staph infections. Since staph germs tend to live in the nose, I'm concerned that I'm reinfecting myself through my nasal pillows. I bought a terrifying cleaning solution online but it doesn't sound like it's safe for pillows. Any suggestions?
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/ |
Use data to optimize your xPAP treatment:
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur
Re: disinfecting nasal pillows
Pugsy -- that's the one I bought, and it can't be used on nasal pillows.
I do have wipes -- they're always very cheap so I have a bunch of them, but I don't think they have any sanitizing/disinfecting properties.
Bleach would presumably work, but then I'd have the stink of bleach in my nose. I may just need to bite the bullet and buy some fresh pillows.
I do have wipes -- they're always very cheap so I have a bunch of them, but I don't think they have any sanitizing/disinfecting properties.
Bleach would presumably work, but then I'd have the stink of bleach in my nose. I may just need to bite the bullet and buy some fresh pillows.
Using a Phillips System One Apap with an Airfit P10 for her. Pressure 10-14. Ramp 4.
- chunkyfrog
- Posts: 34394
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: disinfecting nasal pillows
If you are having recurrent staph infections,
SEE YOUR DOCTOR!
SEE YOUR DOCTOR!
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: disinfecting nasal pillows
I've seen numerous doctors for it over the years, including an infectious disease specialist. There is no simple cure.
Using a Phillips System One Apap with an Airfit P10 for her. Pressure 10-14. Ramp 4.
Re: disinfecting nasal pillows
What size nasal pillow do you use...and I assume the P10???
I might have a couple of spares that are brand new you can just have depending on the size you use for cost of postage.
I also have a few new Swift FX nasal pillows.
I didn't realize that it wasn't advisable to use Control III on nasal pillows. I never read the fine print.
I might have a couple of spares that are brand new you can just have depending on the size you use for cost of postage.
I also have a few new Swift FX nasal pillows.
I didn't realize that it wasn't advisable to use Control III on nasal pillows. I never read the fine print.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: disinfecting nasal pillows
Control III is nasty and somewhat dangerous UNTIL DILUTED. READ THE FLIPPING DIRECTIONS.
Before use, you add one ounce of concentrate to 1 gallon of water. Thats something like 100 to 1 dilution.
Once diluted, it's pretty benign. Don't drink it or get it in your eyes. Unfortunately, the labeling on this is somewhat confusing to many people.
Check the manufacturer's web site: http://www.controlthree.com/disinfectantgermicide.html
Where did you get the idea that it's not suitable to disinfect nasal pillows?
Before use, you add one ounce of concentrate to 1 gallon of water. Thats something like 100 to 1 dilution.
Once diluted, it's pretty benign. Don't drink it or get it in your eyes. Unfortunately, the labeling on this is somewhat confusing to many people.
Check the manufacturer's web site: http://www.controlthree.com/disinfectantgermicide.html
--My note-- This is the DILUTED solution.Q. Does repeated use of Control III Disinfectant germicide harm equipment, or degrade the material?
A. No. Control III does not contain drying chemicals that harm rubber, plastic, or metal. Furthermore, the use solution does not contain caustic or corrosive substances which harm equipment or cause irritation to the face if masks are disinfected with it.
Where did you get the idea that it's not suitable to disinfect nasal pillows?
_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Re: disinfecting nasal pillows
All I use is alcohol swaps on all my nasal gear for a nightly clean for the last 7 yrs, never had a problem and yes people will correct me for even suggesting alcohol on silicon products. They are inexpensive and easy available at any pharmacy in N America.
- chunkyfrog
- Posts: 34394
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: disinfecting nasal pillows
I second the idea of alcohol for pillows and other items upwind of my nose.
However, I prefer ETHYL alcohol.
Inexpensive vodka, to be specific--a splash for the slimy tank--and one for my coffee.
Another splash in the funky hose--and another for my coffee.
A final splash for my nasal pillows . . .
DAMN, THIS IS GOOD COFFEE!
However, I prefer ETHYL alcohol.
Inexpensive vodka, to be specific--a splash for the slimy tank--and one for my coffee.
Another splash in the funky hose--and another for my coffee.
A final splash for my nasal pillows . . .
DAMN, THIS IS GOOD COFFEE!
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: disinfecting nasal pillows
Archangle: the instructions on the bottle I have state "not to be used as a terminal strerilant/high-level disinfectant on any surface or instrument that... contacts intact mucous membranes." It's possible I'm reading the instructions wrong -- they are indeed unclear.
Pugsy: That's a very kind suggestion. I use medium size. I should start saving all the extra pillows I get to send on to people, too.
I'm not sure about the efficacy of alcohol wipes for this; I'll look it up.
Pugsy: That's a very kind suggestion. I use medium size. I should start saving all the extra pillows I get to send on to people, too.
I'm not sure about the efficacy of alcohol wipes for this; I'll look it up.
Using a Phillips System One Apap with an Airfit P10 for her. Pressure 10-14. Ramp 4.
Re: disinfecting nasal pillows
See: https://www.resmed.com/us/dam/documents ... er_eng.pdf
This is a longish pdf, so it may take a few seconds to load.
This is a longish pdf, so it may take a few seconds to load.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203 |