help! still tired!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Dgriffin333
Posts: 44
Joined: Sun Oct 29, 2017 10:23 pm
Location: Utah

help! still tired!

Post by Dgriffin333 » Wed Nov 29, 2017 12:20 pm

Image

Image

hi newbie here. i have been getting about 8 hours of sleep on apap for the last 7 days and still feel very tired. my AHI has been averaging about 4 for the 7 days. however my starting AHI before apap from the sleep study was 14 so wondering are these results good enough to start feeling better? and if so how long will it take to see improvement? or is there anything from these results that stand out? any help is much appreciated thanks.

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: help! still tired!

Post by Julie » Wed Nov 29, 2017 12:23 pm

What are your pressure settings? You've kept them from showing up under the pie chart results (along with others).

Dgriffin333
Posts: 44
Joined: Sun Oct 29, 2017 10:23 pm
Location: Utah

Re: help! still tired!

Post by Dgriffin333 » Wed Nov 29, 2017 12:25 pm

min 6 max 10

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: help! still tired!

Post by Wulfman... » Wed Nov 29, 2017 1:20 pm

Personally, I could not sleep through those pressure fluctuations/changes (and would feel like crap).
And, your minimum is too low based on the amount of events you're still having.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: help! still tired!

Post by Julie » Wed Nov 29, 2017 1:54 pm

Bump your min. pressure to e.g. 8 or 9 for a few nights to see how things go and raise the upper one to 15 or higher.

FYI Den has a bee in his bonnet re not liking auto, but the majority of us do like it, so keep that in mind.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: help! still tired!

Post by Wulfman... » Wed Nov 29, 2017 2:10 pm

Julie wrote:Bump your min. pressure to e.g. 8 or 9 for a few nights to see how things go and raise the upper one to 15 or higher.

FYI Den has a bee in his bonnet re not liking auto, but the majority of us do like it, so keep that in mind.
It comes from personal experience.
When someone says "I've been on APAP for XX days/weeks/months and still feel unrested" and they post reports like that, I post my recommendations as a "SUGGESTION". It doesn't cost any more to TRY some different settings.
"Doing the same thing over and over and expecting different results is the definition of insanity"

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Dgriffin333
Posts: 44
Joined: Sun Oct 29, 2017 10:23 pm
Location: Utah

Re: help! still tired!

Post by Dgriffin333 » Wed Nov 29, 2017 2:15 pm

Also for a little more information I had my pressure settings at 7 to 15 and was not able to sleep and bumped my upper pressure down to 10 recommended from my sleep doc and was able to sleep through the night so it seems that I can’t handle the higher pressure without waking up every hour

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: help! still tired!

Post by Wulfman... » Wed Nov 29, 2017 2:18 pm

Dgriffin333 wrote:Also for a little more information I had my pressure settings at 7 to 15 and was not able to sleep and bumped my upper pressure down to 10 recommended from my sleep doc and was able to sleep through the night so it seems that I can’t handle the higher pressure without waking up every hour
For the most part, the important setting is the minimum pressure......to eliminate almost all of the events......similar to setting a straight CPAP machine.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: help! still tired!

Post by DreamStalker » Wed Nov 29, 2017 2:24 pm

Wulfman... wrote:
Julie wrote:Bump your min. pressure to e.g. 8 or 9 for a few nights to see how things go and raise the upper one to 15 or higher.

FYI Den has a bee in his bonnet re not liking auto, but the majority of us do like it, so keep that in mind.
It comes from personal experience.
When someone says "I've been on APAP for XX days/weeks/months and still feel unrested" and they post reports like that, I post my recommendations as a "SUGGESTION". It doesn't cost any more to TRY some different settings.
"Doing the same thing over and over and expecting different results is the definition of insanity"

Den

.
Common sense would suggest that not everyone has any.

I started off on APAP and found my "optimal" pressure, then took Den's suggestion and switched to CPAP and have never NOT had a good night's sleep since.
Last edited by DreamStalker on Wed Nov 29, 2017 2:36 pm, edited 1 time in total.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: help! still tired!

Post by Goofproof » Wed Nov 29, 2017 2:29 pm

Wulfman... wrote:
Julie wrote:Bump your min. pressure to e.g. 8 or 9 for a few nights to see how things go and raise the upper one to 15 or higher.

FYI Den has a bee in his bonnet re not liking auto, but the majority of us do like it, so keep that in mind.
It comes from personal experience.
When someone says "I've been on APAP for XX days/weeks/months and still feel unrested" and they post reports like that, I post my recommendations as a "SUGGESTION". It doesn't cost any more to TRY some different settings.
"Doing the same thing over and over and expecting different results is the definition of insanity"

Den

.
It's only one small Bee, and we both like Honey, I have two APAPS and one CPAP, all "Tanks". The CPAP is my first 12 years old, I will go to APAP when it dies But CPAP also provides great results, as long as you have it set correctly, ad if pressure changes bother you, with CPAP you don't have them, APAPS can also be set to CPAP Mode, win, win! Keep that Honey flowing Den! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: help! still tired!

Post by Wulfman... » Wed Nov 29, 2017 2:30 pm

DreamStalker wrote:
Wulfman... wrote:
Julie wrote:Bump your min. pressure to e.g. 8 or 9 for a few nights to see how things go and raise the upper one to 15 or higher.

FYI Den has a bee in his bonnet re not liking auto, but the majority of us do like it, so keep that in mind.
It comes from personal experience.
When someone says "I've been on APAP for XX days/weeks/months and still feel unrested" and they post reports like that, I post my recommendations as a "SUGGESTION". It doesn't cost any more to TRY some different settings.
"Doing the same thing over and over and expecting different results is the definition of insanity"

Den

.
Common sense would suggest that not everyone has any.

I started off on APAP and found my "optimal" pressure, then took Den's suggestion and switched to CPAP and have never had a good night's sleep since.
I missed the "wink" when I first read that.......thinking you actually meant it or mis-typed it.
It's OK. Over the years I've had lots and lots of "converts" once they try it.

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: help! still tired!

Post by DreamStalker » Wed Nov 29, 2017 2:35 pm

Wulfman... wrote:
DreamStalker wrote:
Wulfman... wrote:
Julie wrote:Bump your min. pressure to e.g. 8 or 9 for a few nights to see how things go and raise the upper one to 15 or higher.

FYI Den has a bee in his bonnet re not liking auto, but the majority of us do like it, so keep that in mind.
It comes from personal experience.
When someone says "I've been on APAP for XX days/weeks/months and still feel unrested" and they post reports like that, I post my recommendations as a "SUGGESTION". It doesn't cost any more to TRY some different settings.
"Doing the same thing over and over and expecting different results is the definition of insanity"

Den

.
Common sense would suggest that not everyone has any.

I started off on APAP and found my "optimal" pressure, then took Den's suggestion and switched to CPAP and have never had a good night's sleep since.
I missed the "wink" when I first read that.......thinking you actually meant it or mis-typed it.
It's OK. Over the years I've had lots and lots of "converts" once they try it.

Den

.

Ummm ... that didn't come out right.

I meant to say I HAVE HAD a good night's sleep ever since.



I updated post.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

Dgriffin333
Posts: 44
Joined: Sun Oct 29, 2017 10:23 pm
Location: Utah

Re: help! still tired!

Post by Dgriffin333 » Wed Nov 29, 2017 2:52 pm

So seems like everyone is suggesting increasing min pressure a bit ?

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: help! still tired!

Post by Wulfman... » Wed Nov 29, 2017 3:00 pm

Dgriffin333 wrote:So seems like everyone is suggesting increasing min pressure a bit ?
Well, if you wish to stick with a range of pressures, try raising the minimum by 1 cm. for several nights at a time (3 or more).

If you wish to try a straight pressure, try raising the minimum to 10 cm. and see what that does. In effect, when the minimum and the maximum are set to the same setting, you have straight pressure but other events will also be recorded (like Flow Limitations).


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: help! still tired!

Post by Julie » Wed Nov 29, 2017 3:25 pm

I never suggested NOT trying anything, only alerted the newbie to Den's particular 'prejudice' re therapy mode. THe fact remains that the majority use Apap and find it helpful and I suggest you continue with it until such time as it's not and you've proven that is the problem rather than something else. A min. setting of 6 is low by any standards and the usual advice apart from other possible issues is to try it at a bit higher for a short time. Certainly anyone's free to do what they want at any time.