Resperonics Amara View Mask Blowout

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
AMESS
Posts: 379
Joined: Fri Sep 22, 2017 6:58 am
Location: North Carolina
Contact:

Resperonics Amara View Mask Blowout

Post by AMESS » Wed Oct 18, 2017 9:06 am

Seems I am having an Amara View mask blowout. The machine is ramping up faster than the mask can take it. The Drs office set the pressure. Does anybody know how to get into the machine to change it with going to drs office with the machine? And, has anyone had this problem?? Thanx

User avatar
Pugsy
Posts: 64029
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resperonics Amara View Mask Blowout

Post by Pugsy » Wed Oct 18, 2017 9:13 am

Your machine actually doesn't increase the pressure very fast at all. I used the View with a machine that does increase the pressure really quickly and didn't have blow outs.

What machine are you actually using and at what pressure setting and are you using ramp or not?

And yes, settings can easily be changed but we need to know what machine you are using to be able to tell you how.

When you have time you might want to add your equipment to your profile so we can see at a glance what you are using.
wiki/index.php/Registering_Equipment_in_User_Profile

Are you using the mask fit option feature? If not, you should because you need to fit the mask at your 14 cm pressure and not what it starts with when you first turn it on unless you are using fixed cpap mode with no ramp.
Fitting at 4 to 6 cm when machine is first turned on isn't going to work so great when the machine finally gets to 14.
You have to do the fitting at 14.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
AMESS
Posts: 379
Joined: Fri Sep 22, 2017 6:58 am
Location: North Carolina
Contact:

Re: Resperonics Amara View Mask Blowout

Post by AMESS » Wed Oct 18, 2017 10:09 am

Using res med Auto air sense 10

User avatar
Pugsy
Posts: 64029
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resperonics Amara View Mask Blowout

Post by Pugsy » Wed Oct 18, 2017 10:22 am

And what are your pressure settings? What pressure do you fit your mask at the beginning of the night?
And if you don't know how to find out...read this manual.
https://sleep.tnet.com/home/files/resme ... -guide.pdf

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
AMESS
Posts: 379
Joined: Fri Sep 22, 2017 6:58 am
Location: North Carolina
Contact:

Re: Resperonics Amara View Mask Blowout

Post by AMESS » Wed Oct 18, 2017 11:38 am

machine is set to go from 5-20--but hangs around 14

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: Resperonics Amara View Mask Blowout

Post by TASmart » Wed Oct 18, 2017 11:45 am

You need to do two things IMO. First is get sleepyhead so you can see the pressure profile and set the operating range to the correct narrower range. After that then using the check mask leak feature before you start the night's therapy which lets the machine run at the max selected pressure so you can fit your mask appropriately.

And one more thing. If you go to the user control panel, you can select keep me signed on and never have to mess with the signed in /not signed in mess anymore.
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
Pugsy
Posts: 64029
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resperonics Amara View Mask Blowout

Post by Pugsy » Wed Oct 18, 2017 11:48 am

Okay, you start the night at 5 cm and probably fit the mask at 5 cm but the fit doesn't work so great when you hit 14.
You need to anticipate the 14 cm by fitting at 14 cm at the beginning of the night.
This is where the mask fit function on your machine can help you out. It will deliver higher pressures while you are fitting the mask so that hopefully it won't be so leaky at 14 when you hit the higher pressures. Once the mask fit option is completed it will revert back to the 5 cm starting pressure so you can go to sleep.

Your machine doesn't change pressures rapidly..it can't go from 5 to 14 in a blink of an eye. It actually takes several minutes to get there assuming it doesn't stop along the way.
The leaks at 14 are simply because you fitted for 5 cm and what works fine at 5 doesn't work so great at 14.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Resperonics Amara View Mask Blowout

Post by Goofproof » Wed Oct 18, 2017 12:10 pm

TASmart wrote:You need to do two things IMO. First is get sleepyhead so you can see the pressure profile and set the operating range to the correct narrower range. After that then using the check mask leak feature before you start the night's therapy which lets the machine run at the max selected pressure so you can fit your mask appropriately.

And one more thing. If you go to the user control panel, you can select keep me signed on and never have to mess with the signed in /not signed in mess anymore.
Never count on never. If you clear your cookies, at even without, you will have to re-login, but it work.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Pugsy
Posts: 64029
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resperonics Amara View Mask Blowout

Post by Pugsy » Wed Oct 18, 2017 12:25 pm

Goofproof wrote: Never count on never. If you clear your cookies, at even without, you will have to re-login, but it work.
Yep, sometimes the forum cookie goes awol and we did nothing to cause it and we have to log in again but it doesn't happen very often.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
TASmart
Posts: 1071
Joined: Wed Jan 04, 2017 7:23 pm
Location: Eugene, OR

Re: Resperonics Amara View Mask Blowout

Post by TASmart » Wed Oct 18, 2017 12:47 pm

OK very rarely. Are ya happy now?
All posts reflect my own opinion based on my experience and reading.
Your mileage may vary
Past performance is no guarantee of future results
Consult with your own physician as people very

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Resperonics Amara View Mask Blowout

Post by Goofproof » Wed Oct 18, 2017 1:00 pm

TASmart wrote:OK very rarely. Are ya happy now?
Never and Can't aren't allowed in my World! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Okie bipap
Posts: 3553
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Resperonics Amara View Mask Blowout

Post by Okie bipap » Wed Oct 18, 2017 2:50 pm

I used that mask when my pressie was running around 20 cm all night long. Once I got it adjusted, it worked fine. At that time, my pressure was set to 20 - 25. If you use the max fit function on your machine, it will give you three minutes to get your mask adjusted. I would adjust it sitting on the side of the bed, then lay down and check the fit. I would just let the machine continue to run in the mask fit mode until it Timed out and started running normally. I still do the same thing with my current mask.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

A mess

Re: Resperonics Amara View Mask Blowout

Post by A mess » Wed Oct 18, 2017 3:34 pm

Okay TY all will see how it goes tonight

A MESS

Re: Resperonics Amara View Mask Blowout

Post by A MESS » Wed Oct 18, 2017 3:35 pm

Pugs-yes already have sleepy head

User avatar
Hannibal 2
Posts: 485
Joined: Sun Nov 20, 2016 4:46 am
Location: Derbyshire UK

Re: Resperonics Amara View Mask Blowout

Post by Hannibal 2 » Sun Oct 22, 2017 7:41 am

AMESS wrote:machine is set to go from 5-20--but hangs around 14
You'll probably find it less likely to leak once you can work out your optimum pressures. I have been using this mask for 5 nights now and have had some minor leak issues, not loud and farty but enough to disturb my sleep, checked sleepyhead and nothing majorly over the red leak line. I kept making some subtle strap and positioning adjustments until I got it right, you have to patient with any mask to get it to fit right, and that can take several nights. Running the leak test and finding your optimum pressures may help but sometimes it just may not be the right mask for you.

This may help: https://www.youtube.com/watch?v=Mp02l4r6Je8

_________________
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Air 10 ClimateLineAir, SleepyHead Software
"Welcome my son, welcome to the machine!
Where have you been? It's alright we know where you've been!"
(You've been in the pipeline...)