2nd scan for review

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
grybeard77
Posts: 292
Joined: Fri Oct 06, 2006 1:27 pm
Location: Perris, Calif.

2nd scan for review

Post by grybeard77 » Sun Sep 17, 2017 12:43 pm

Image

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: I use Sleepyhead software, it's the best!

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 2nd scan for review

Post by Pugsy » Sun Sep 17, 2017 12:55 pm

Did you sleep through those leaks?
Did you do anything to help you decided if those are mask movement or mouth breathing leaks?

The report is marginally improved in terms of leak over the other report.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
LSAT
Posts: 13232
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: 2nd scan for review

Post by LSAT » Sun Sep 17, 2017 2:00 pm

I'd guess it's mouth breathing.

User avatar
grybeard77
Posts: 292
Joined: Fri Oct 06, 2006 1:27 pm
Location: Perris, Calif.

Re: 2nd scan for review

Post by grybeard77 » Sun Sep 17, 2017 2:10 pm

Pugsy wrote:Did you sleep through those leaks?
Did you do anything to help you decided if those are mask movement or mouth breathing leaks?

The report is marginally improved in terms of leak over the other report.
Yes, I sleep through them all and even using the C collar and gel stuff, I wake up about ever 4 hours to pee and mouth is usually dry.

I can't really say I move or check my mask very often so guess 95% of leaks must be mouth breathing. Would the next size larger
C collar help? Can't tape until I cut beard and mustache off and I've had them a long time...LOL

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: I use Sleepyhead software, it's the best!

User avatar
Pugsy
Posts: 64027
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 2nd scan for review

Post by Pugsy » Sun Sep 17, 2017 2:22 pm

I don't know about how cervical collars help or sizes and results. No experience there.

You are within what ResMed would give you a Mr Smiley for in terms of mask fit. ResMed won't give you Mr Frowny until you hit 30% or more.
Given the sporadic tendency and the fact that a lot of the leaks don't go over 35 L/min...your AHI is probably fairly accurate.
It's not like it is likely missing a ton of events during those leaks.

Since you are sleeping well...up to you how much you want to worry about the leaks and what you wish to try to reduce them.

Roughly 10 hours of sleep...so 600 minutes and 19% over 24 L/min which is 114 minutes but a lot of it not over 24 really bad.
Not quite 2 hours. It's a bit more than I would want to see myself but your longest block was from 4:30 to 5:00..so 30 minutes there. Not all of that 30 minutes was above 35 where the machine could really have a problem flagging something.
Up to 35..it still will do a decent enough job and you would most likely see some unknown apneas if something were happening.

I know people will say "get a full face mask" but I assume you have already been down that road and probably decided it wasn't for you. It's not like they don't have their own set of leak problems.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
grybeard77
Posts: 292
Joined: Fri Oct 06, 2006 1:27 pm
Location: Perris, Calif.

Re: 2nd scan for review

Post by grybeard77 » Sun Sep 17, 2017 5:28 pm

Pugsy wrote:I don't know about how cervical collars help or sizes and results. No experience there.

You are within what ResMed would give you a Mr Smiley for in terms of mask fit. ResMed won't give you Mr Frowny until you hit 30% or more.
Given the sporadic tendency and the fact that a lot of the leaks don't go over 35 L/min...your AHI is probably fairly accurate.
It's not like it is likely missing a ton of events during those leaks.

Since you are sleeping well...up to you how much you want to worry about the leaks and what you wish to try to reduce them.

Roughly 10 hours of sleep...so 600 minutes and 19% over 24 L/min which is 114 minutes but a lot of it not over 24 really bad.
Not quite 2 hours. It's a bit more than I would want to see myself but your longest block was from 4:30 to 5:00..so 30 minutes there. Not all of that 30 minutes was above 35 where the machine could really have a problem flagging something.
Up to 35..it still will do a decent enough job and you would most likely see some unknown apneas if something were happening.

I know people will say "get a full face mask" but I assume you have already been down that road and probably decided it wasn't for you. It's not like they don't have their own set of leak problems.
Actually , I do have a F10 (never used) in my supply box of stuff and I will try it out over the next couple of nights and I also have a next larger size C collar to try if the F10 doesn't work out. Might have to end up using both.

Stand by for results, as they occur....or not!!!

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: I use Sleepyhead software, it's the best!

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: 2nd scan for review

Post by zonker » Sun Sep 17, 2017 5:45 pm

my apologies if i've posted this link to you before. can't remember who i tell WHAT anymore. the joys of old age. <sigh>

a forum member has a helpful way of keeping the mouth closed-

viewtopic.php?f=1&t=112758&st=0&sk=t&sd ... i#p1086296

again, sorry if i'm repeating myself.....
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
grybeard77
Posts: 292
Joined: Fri Oct 06, 2006 1:27 pm
Location: Perris, Calif.

Re: 2nd scan for review

Post by grybeard77 » Sun Sep 17, 2017 7:03 pm

zonker wrote:my apologies if i've posted this link to you before. can't remember who i tell WHAT anymore. the joys of old age. <sigh>

a forum member has a helpful way of keeping the mouth closed-

viewtopic.php?f=1&t=112758&st=0&sk=t&sd ... i#p1086296

again, sorry if i'm repeating myself.....
Thanks for the link, I have also used the Scunci head band with my cervical collar on many nights, it helped me some, but wasn't the cure all I keep looking for. I could still manage to open my lips enough to score big leaks. The Scunci band could be all some people need to stop leaks, I'm just not one of them.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: I use Sleepyhead software, it's the best!

TedVPAP
Posts: 974
Joined: Sat Jun 12, 2010 10:29 am

Re: 2nd scan for review

Post by TedVPAP » Sun Sep 17, 2017 8:38 pm

I use the Ultimate Chinstrap to control mouth leaking.
It is not clear whether your leaking is causing any issues or not.
If it were me, I would want to experiment as there MAY be a benefit.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: AutoPAP 16-20, Ultimate Chin Strap http://sleepapneasolutionsinc.com/
Use data to optimize your xPAP treatment:
how to see your data https://sleep.tnet.com/resources/sleepyhead
how to present your data https://sleep.tnet.com/resources/sleepyhead/shorganize
how to post your data https://sleep.tnet.com/reference/tips/imgur

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: 2nd scan for review

Post by zonker » Mon Sep 18, 2017 11:50 am

grybeard77 wrote:
zonker wrote:my apologies if i've posted this link to you before. can't remember who i tell WHAT anymore. the joys of old age. <sigh>

a forum member has a helpful way of keeping the mouth closed-

viewtopic.php?f=1&t=112758&st=0&sk=t&sd ... i#p1086296

again, sorry if i'm repeating myself.....
Thanks for the link, I have also used the Scunci head band with my cervical collar on many nights, it helped me some, but wasn't the cure all I keep looking for. I could still manage to open my lips enough to score big leaks. The Scunci band could be all some people need to stop leaks, I'm just not one of them.
i can certainly understand that. i did a bit of modification that helped keep my lips closed. i bought a bunch of "Goody Ouchless comfort headwraps". hard for me to describe, but these are basically rubber bands with a tight cloth covering. i use one of these over the scunci right below where the scunci has that sewn in band. it pulled it taught enough that my lips weren't parting.

also, i use two scunci bands, one with the sewn in band down and the other with the sewn in band up. this gives a gentle but firm press on my lips, pushing them slightly into my teeth.

it did the trick for me and might be worth a shot for you.

if not, i hope you find something that works for you....
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
grybeard77
Posts: 292
Joined: Fri Oct 06, 2006 1:27 pm
Location: Perris, Calif.

Re: 2nd scan for review

Post by grybeard77 » Mon Sep 18, 2017 3:24 pm

I searched and found Walmart's various "Goody Ouches Comfort Headbands" and many other varieties as well. Don't see any described as cloth covered rubber bands. Also don't under stand the band thing for the Souchi head band, much less the one up and one down you describe.
Could you PM me more information on how you did this?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: I use Sleepyhead software, it's the best!

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: 2nd scan for review

Post by zonker » Mon Sep 18, 2017 7:26 pm

grybeard77 wrote:I searched and found Walmart's various "Goody Ouches Comfort Headbands" and many other varieties as well. Don't see any described as cloth covered rubber bands. Also don't under stand the band thing for the Souchi head band, much less the one up and one down you describe.
Could you PM me more information on how you did this?
sorry, i'm sure i'm not making myself clear. give me a couple of days(other things going on) and i'll send a pm with what i hope will be a better explanation.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
grybeard77
Posts: 292
Joined: Fri Oct 06, 2006 1:27 pm
Location: Perris, Calif.

Re: 2nd scan for review

Post by grybeard77 » Mon Sep 18, 2017 7:30 pm

Zonker , please see new PM I just sent you. Thanks in advance and take your time....

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: I use Sleepyhead software, it's the best!