Central Apneas
So I am looking at the machine today. Still trying to get used to the full face mask and finding that 'right pressure to fall asleep to and not wake up' but my machine as well as my sleep test was showing some central apneas.
Re: Central Apneas
If your sleep test had CA, the odds are you are going to keep some. The cpap CA can be a mix of real and getting use to treatment ca. How many are you getting an hour?
you can upload sleepyhead and post the link http://imgur.com/
you can upload sleepyhead and post the link http://imgur.com/
_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 ST-A iVAPS and adapt ASV |
-
- Posts: 10
- Joined: Sat Feb 04, 2017 8:23 pm
Re: Central Apneas
Around 3 per hour according to the machine.ajack wrote:If your sleep test had CA, the odds are you are going to keep some. The cpap CA can be a mix of real and getting use to treatment ca. How many are you getting an hour?
you can upload sleepyhead and post the link http://imgur.com/
Re: Central Apneas
Are you for sure asleep when you see the central flags?
Are you seeing any OA or hyponea flags around the same time?
Are you seeing any OA or hyponea flags around the same time?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Central Apneas
At this stage 3 an hour, if they are 10-20 seconds or so and not 2 minutes aren't an issue. You need an o2 reading to confirm of any are real. As pugsy is saying, there are a lot of reasons for flagged 'clear airway' CA events that aren't real central apnea.
_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: S9 ST-A iVAPS and adapt ASV |
Re: Central Apneas
This.Pugsy wrote:Are you for sure asleep when you see the central flags?
Are you seeing any OA or hyponea flags around the same time?
You may be like me and getting a lot of SWJ (sleep/wake/junk) clusters as you're drifting off to sleep.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Hybrid Full Face CPAP Mask with Nasal Pillows and Headgear |
Analog guy in a digital world.