Battery Backup for resmed aircurv 10

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Ray4852
Posts: 113
Joined: Tue Nov 16, 2021 7:24 am

Re: Battery Backup for resmed aircurv 10

Post by Ray4852 » Thu Dec 30, 2021 6:59 am

My setup. battle born 100 amp hour lithium battery. I monitored it with a Victron smart shunt. what ever battery you decide to get. make sure you get more than enough power you need. if you under size it. it wont last very long.

mibbim
Posts: 101
Joined: Fri Sep 23, 2016 7:32 am
Location: Rochester, MI

Re: Battery Backup for resmed aircurv 10

Post by mibbim » Fri Dec 31, 2021 8:29 am

does this package includ everything I need? Does it also fit to the aircurv 10?

https://www.cpapdirect.com/parts-and-ac ... ku_id=1122

stevenal
Posts: 193
Joined: Fri Jun 10, 2011 2:00 pm

Re: Battery Backup for resmed aircurv 10

Post by stevenal » Fri Dec 31, 2021 8:39 pm

From what I can tell, the Aircurve and Airsense 10's use the same power supply. If this is correct, yes.

See https://batterypowersolutions.net for more info.
AirSence 10 Autoset, Dreamstation 2 backup, Swift FX, Z1 for travel.

User avatar
loggerhead12
Posts: 585
Joined: Fri Sep 01, 2017 4:00 pm

Re: Battery Backup for resmed aircurv 10

Post by loggerhead12 » Fri Dec 31, 2021 8:42 pm

mibbim wrote:
Fri Dec 31, 2021 8:29 am
does this package includ everything I need? Does it also fit to the aircurv 10?

https://www.cpapdirect.com/parts-and-ac ... ku_id=1122
Yeah, if you're OK spending $330 for 7 hours of battery life.

User avatar
babydinosnoreless
Posts: 2364
Joined: Fri Nov 02, 2018 2:53 pm

Re: Battery Backup for resmed aircurv 10

Post by babydinosnoreless » Fri Dec 31, 2021 9:51 pm

loggerhead12 wrote:
Fri Dec 31, 2021 8:42 pm
mibbim wrote:
Fri Dec 31, 2021 8:29 am
does this package includ everything I need? Does it also fit to the aircurv 10?

https://www.cpapdirect.com/parts-and-ac ... ku_id=1122
Yeah, if you're OK spending $330 for 7 hours of battery life.
How many hours do you think something like this might last ?
https://www.jackery.com/products/jacker ... 4423198807

Could we boondock for a 4 or 5 day trip with it ?

Ray4852
Posts: 113
Joined: Tue Nov 16, 2021 7:24 am

Re: Battery Backup for resmed aircurv 10

Post by Ray4852 » Sat Jan 01, 2022 7:32 am

The Jackery 500 is loaded with a lot of stuff you don't need. the battery they put in it is only rated for 518 amps. this seems a lot for a cpap that will only use about 90 watts per 8 hours of use. cut the 518 amps in half you will only have half the battery you can use safely if you want this unit to last. for this kind of money. you can do much better and get twice the battery power by going with an 100 amp hour AGM battery. theres plenty of good AGM batteries you can buy today. its all up to you how much you want to spend. 100 amp battery is rated for 1200 watts. cut this battery in half you only have 600 watt battery you can use with your cpap. divide 90 into 600 you can use this battery for 6 days. now you have to have something to charge this battery. solar power is great if you have the sun. no sun you need a generator along with a battery charger to top off your battery. when I boondock I go out for 30 days or more. I have a Renogy 100 watt solar suitcase along with a Victron blue smart 20 amp battery charger. my Yamaha generator supply my power I need for the battery charger. what I recommend for a AGM battery. A 100 amp Vmax for 300 dollars. The solar suitcase for 250 dollars. I use the lithium battery from battle born. this battery cost about 850 dollars. you need the battery monitored too. the smart shunt cost 130 dollars. When I use any battery. I use the upper 30 percent of my battery only not 50 percent.

mibbim
Posts: 101
Joined: Fri Sep 23, 2016 7:32 am
Location: Rochester, MI

Re: Battery Backup for resmed aircurv 10

Post by mibbim » Sat Jan 01, 2022 8:45 am

Sorry, I am keeping ask question since I am not familar with the eletrical stuff....so the battery tender is to charge the 35M battery... resmed conveter is the one plug in to the cpap machine? Do I still need that spillter? Or can be just the 35M battery + resmed converter (is the 12-volt accessory adapter cigarette lighter needed for connecting the converter and 35M battery?) + battery tender?

Ray4852
Posts: 113
Joined: Tue Nov 16, 2021 7:24 am

Re: Battery Backup for resmed aircurv 10

Post by Ray4852 » Sat Jan 01, 2022 10:55 am

you need the DC converter. one end plug in to the cpap. the other end goes to your battery. the battery tender or charger, charges your battery. if you plan to use a 35 amp battery. get a 7 amp charger from Walmart or auto store. the plug that connects to your battery comes with clamps. cut off these clamps and have someone crimp on two rig terminals for you. this will give you a better connection to your battery. The market is loaded with cheap battery packs you can buy for 150 dollars from amazon. all you do is plug in and you have power. what I don't like about them. most of them are poor quality. those compact battery packs are nice for people who like to camp on the go with light equipment. this is the most expensive way to go.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Battery Backup for resmed aircurv 10

Post by palerider » Sat Jan 01, 2022 12:13 pm

Ray4852 wrote:
Sat Jan 01, 2022 7:32 am
The Jackery 500 is loaded with a lot of stuff you don't need. the battery they put in it is only rated for 518 amps. this seems a lot for a cpap that will only use about 90 watts per 8 hours of use. cut the 518 amps in half you will only have half the battery you can use safely if you want this unit to last. for this kind of money. you can do much better and get twice the battery power by going with an 100 amp hour AGM battery. theres plenty of good AGM batteries you can buy today. its all up to you how much you want to spend. 100 amp battery is rated for 1200 watts. cut this battery in half you only have 600 watt battery you can use with your cpap. divide 90 into 600 you can use this battery for 6 days. now you have to have something to charge this battery. solar power is great if you have the sun. no sun you need a generator along with a battery charger to top off your battery. when I boondock I go out for 30 days or more. I have a Renogy 100 watt solar suitcase along with a Victron blue smart 20 amp battery charger. my Yamaha generator supply my power I need for the battery charger. what I recommend for a AGM battery. A 100 amp Vmax for 300 dollars. The solar suitcase for 250 dollars. I use the lithium battery from battle born. this battery cost about 850 dollars. you need the battery monitored too. the smart shunt cost 130 dollars. When I use any battery. I use the upper 30 percent of my battery only not 50 percent.
You're using a lot of terms that that are going to confuse people looking for information, because you're using them wrong.

"amps" is an instantaneous current flow, it has NOTHING to do with the storage capacity or the amount of power that a battery can supply.

"watts" is likewise an instantaneous measure of power flow, and having nothing to do with capacity, or total power. You don't get charged for "watts" on your electric bill, you get charged per thousand watthours (KWh).

Capacity is defined as watthours, ie, x watts for y hours.

"amphours" is another partial definition of capacity, but you have to know how many volts that specification is made at. It's fairly simple for a lead acid battery which has a nominal voltage of 12 volts, for instance, a 35amp hour scooter battery, times 12 volts gives you 420 watthours. to compare capacity, you *MUST* convert everything to watthours.

The jackery 500 claims to offer 518 watthours. (not 518 "amps" as you allege).

A CPAP uses around 80-100 watthours,

a 100AH AGM battery has a capacity of 1200wattHOURS

The discharge life cycle between the jackery 500 with lithium ion batteries, a 100AH AGM battery and your battleborn with lifepo4 cells are all completely different.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

stevenal
Posts: 193
Joined: Fri Jun 10, 2011 2:00 pm

Re: Battery Backup for resmed aircurv 10

Post by stevenal » Sat Jan 01, 2022 3:31 pm

palerider,

I was with you until:
palerider wrote:
Sat Jan 01, 2022 12:13 pm

A CPAP uses around 80-100 watthours,
A statement like that needs a time period. Per night perhaps?
AirSence 10 Autoset, Dreamstation 2 backup, Swift FX, Z1 for travel.

Aretino
Posts: 109
Joined: Sun Mar 15, 2020 2:07 pm

Re: Battery Backup for resmed aircurv 10

Post by Aretino » Sat Jan 01, 2022 5:05 pm

My Resmed Aircurve 10 is a 24vdc machine. I use two Duracell 12-35J batteries connected in series to achieve 24vdc.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Started 2005 with a Remstar machine. Then a 2020 Dreamstation 1. Now a Resmed Aircurve ST

stevenal
Posts: 193
Joined: Fri Jun 10, 2011 2:00 pm

Re: Battery Backup for resmed aircurv 10

Post by stevenal » Sat Jan 01, 2022 7:51 pm

If you use the Resmed DC adapter, I don't think 12 versus 24 V input matters. Two batteries in series or parallel will double the Wh of a single. If bypassing the Resmed adapter with 24 V, the 3.33 V on the center pin requirement must be dealt with.
AirSence 10 Autoset, Dreamstation 2 backup, Swift FX, Z1 for travel.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Battery Backup for resmed aircurv 10

Post by palerider » Sat Jan 01, 2022 8:49 pm

stevenal wrote:
Sat Jan 01, 2022 3:31 pm
palerider,

I was with you until:
palerider wrote:
Sat Jan 01, 2022 12:13 pm

A CPAP uses around 80-100 watthours,
A statement like that needs a time period. Per night perhaps?
No, it doesn't need a time period, because it's a measure of power used over time already. It's right there in the word.

Now, if I'd say "a CPAP uses 10-12 WATTS", then yes, that would need a duration of how long it pulls those 10-12 watts, then you would multiply the hours (assuming 8) by the 10-12, and get 80-96 watthours.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Battery Backup for resmed aircurv 10

Post by palerider » Sat Jan 01, 2022 8:50 pm

Aretino wrote:
Sat Jan 01, 2022 5:05 pm
My Resmed Aircurve 10 is a 24vdc machine. I use two Duracell 12-35J batteries connected in series to achieve 24vdc.
The Resmed DC adapter works on 12-24 volts, it doesn't matter what the machine uses.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
loggerhead12
Posts: 585
Joined: Fri Sep 01, 2017 4:00 pm

Re: Battery Backup for resmed aircurv 10

Post by loggerhead12 » Sat Jan 01, 2022 9:41 pm

mibbim wrote:
Sat Jan 01, 2022 8:45 am
Sorry, I am keeping ask question since I am not familar with the eletrical stuff....so the battery tender is to charge the 35M battery... resmed conveter is the one plug in to the cpap machine? Do I still need that spillter? Or can be just the 35M battery + resmed converter (is the 12-volt accessory adapter cigarette lighter needed for connecting the converter and 35M battery?) + battery tender?
You need the 4 pieces I told you in the above post.

1. the battery
2. the battery charger (battery tender)
3. accessory plug/cigarette lighter adapter
4. Resmed converter

You don't need the splitter. You can manually plug the battery into the battery tender to charger it, then plug the battery into the accessory plug to run the CPAP. The splitter just makes is so you don't have to keep plugging and unplugging. It costs $8.